Additional humidification help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Additional humidification help

Post by Legrunt » Thu Aug 05, 2021 12:07 am

Using a dreamstation go. (with the foam removed)
when I use the heated humidifier at level 4, the amount of water is just about right, but when I sleep for longer than usual (eg during weekend) the water runs out and I get dry throat.
If I set the humidifier at level 3, the water lasts long enough, but it feels much better at 4 compared to 3.

I was wondering if I could use an HME filter in conjunction with the heated humidifier at level 3?

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

User avatar
LSAT
Posts: 13393
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Additional humidification help

Post by LSAT » Thu Aug 05, 2021 4:35 am

Using it at home, you could add a small humidifier to the room. Doesn't it get very noisy without the foam?

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Additional humidification help

Post by Pugsy » Thu Aug 05, 2021 9:05 am

Depending on the mask used ....the HME disc might help...it has to be in place real close to your face to help much and some masks have a short hose with the mask that makes it hard to get the disc close to the face.
It's a relatively cheap experiment to try though.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15467
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Additional humidification help

Post by ChicagoGranny » Thu Aug 05, 2021 10:09 am

Legrunt wrote:
Thu Aug 05, 2021 12:07 am
the water runs out and I get dry throat
What are your leak level and pressure? Which type of mask do you use?

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Additional humidification help

Post by Legrunt » Thu Aug 05, 2021 7:56 pm

LSAT wrote:
Thu Aug 05, 2021 4:35 am
Using it at home, you could add a small humidifier to the room. Doesn't it get very noisy without the foam?
The increase in noise is very slight. Makes me wonder why they bothered with the foam in the first place.
Pugsy wrote:
Thu Aug 05, 2021 9:05 am
Depending on the mask used ....the HME disc might help...it has to be in place real close to your face to help much and some masks have a short hose with the mask that makes it hard to get the disc close to the face.
It's a relatively cheap experiment to try though.
I'm using a resmed airtouch f20. They have recently made a humidx hme for the f20 where it sits inside the mask itself.
20210806_095027.jpg
20210806_095027.jpg (120.44 KiB) Viewed 2449 times
20210806_095039.jpg
20210806_095039.jpg (115.93 KiB) Viewed 2449 times
ChicagoGranny wrote:
Thu Aug 05, 2021 10:09 am
What are your leak level and pressure? Which type of mask do you use?
Not sure about leak level, but I get a 100% fit every night. No issue with the mask. Full face resmed f20.

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Last edited by Legrunt on Thu Aug 05, 2021 8:25 pm, edited 1 time in total.
Derek

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Additional humidification help

Post by Janknitz » Thu Aug 05, 2021 8:04 pm

Does your mouth open? Could that be why you use so much water in the FF mask? Or do you live in a very dry climate?
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Additional humidification help

Post by Legrunt » Thu Aug 05, 2021 8:27 pm

Janknitz wrote:
Thu Aug 05, 2021 8:04 pm
Does your mouth open? Could that be why you use so much water in the FF mask? Or do you live in a very dry climate?
Mouth only opens if I use a nasal mask or nasal pillows. That's why I need to use FF. But once I use FF, mouth remains closed.
I live in a very hot area (Malaysia) so we sleep with aircons on (which dries the air out)

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Additional humidification help

Post by Janknitz » Thu Aug 05, 2021 8:42 pm

Legrunt wrote:
Thu Aug 05, 2021 8:27 pm
Janknitz wrote:
Thu Aug 05, 2021 8:04 pm
Does your mouth open? Could that be why you use so much water in the FF mask? Or do you live in a very dry climate?
Mouth only opens if I use a nasal mask or nasal pillows. That's why I need to use FF. But once I use FF, mouth remains closed.
I live in a very hot area (Malaysia) so we sleep with aircons on (which dries the air out)
That makes sense.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
ChicagoGranny
Posts: 15467
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Additional humidification help

Post by ChicagoGranny » Fri Aug 06, 2021 11:40 am

Legrunt wrote:
Thu Aug 05, 2021 8:27 pm
I live in a very hot area (Malaysia) so we sleep with aircons on (which dries the air out)
Try to place your CPAP so that AC is not blowing directly on it.

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Additional humidification help

Post by Legrunt » Sun Aug 08, 2021 8:10 pm

ChicagoGranny wrote:
Fri Aug 06, 2021 11:40 am
Legrunt wrote:
Thu Aug 05, 2021 8:27 pm
I live in a very hot area (Malaysia) so we sleep with aircons on (which dries the air out)
Try to place your CPAP so that AC is not blowing directly on it.
My machine is actually inside my bedside cabinet. I've drilled a hole in the side for the hose to exit. This way, it is dead quiet.
Cabinet is definitely NOT airtight, so there is enough airflow into it for the machine to ingest for pressure.

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Additional humidification help

Post by Legrunt » Sun Aug 08, 2021 8:24 pm

Just to update everyone here...
I've tried it with just the humidX F20 disc without the humidifier chamber. Too dry.
I tried it with the humidX F20 disc along with another inline HME (but still without the humidification chamber, trying to see if I can do away with it). A bit better, but still too dry. (the inline HME sits post vent, so inbetween breaths, the HME is losing moisture to the vent)
finally I tried it with a lower level of humidity set on the humidifier chamber with the humidX F20 disc... This finally works well enough. Before the water warms up, there is significant humidification provided by the F20 disc. Once the water warms up, there is enough humidification for my comfort and the water doesn't run out after 8 hours. I can feel the level of "wetness" in the air... like at the start of the breath, there is a LOT of humidity, but as it tapers off, I can feel it drying out slightly, indicating that the disc has released all its moisture, but there is still a good amount being delivered by the water in the chamber. But nowhere does it feel completely dry.

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

CPAP2017
Posts: 58
Joined: Wed Mar 11, 2020 12:28 pm

Re: Additional humidification help

Post by CPAP2017 » Sat Nov 20, 2021 7:39 am

This may be a difficult solution, but I like tinkering and created a piggyback water chamber that flows into the main chamber as it drains. The water used to last through the night but when I grew a beard it no longer did due a slightly increased airflow for the extra leakage. It was either lose the beard or find a way to make more water available. Works great. Ho Ho Ho.
Attachments
728F869C-4C44-4D67-BFB7-CE4B5838C25E.jpeg
728F869C-4C44-4D67-BFB7-CE4B5838C25E.jpeg (581.5 KiB) Viewed 2199 times