Resmed 10 and SoClean Issue! See Pic!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Medic856
Posts: 8
Joined: Sun Jul 18, 2021 7:43 am

Resmed 10 and SoClean Issue! See Pic!

Post by Medic856 » Sun Jul 18, 2021 8:04 am

I have had the Resmed 10 machine for a little over a year. Upon recommendations from friends, I also purchased the SoClean 2. It has been fine until the last few weeks. I started noticing black specks in the water chamber in the morning. I'm pretty good at cleaning the parts, but I cleaned everything again anyway. Hose, mask, reservoir. Changed SoClean filter. Changed black silicon hose. Changed Resmed filter. Still continued to find specks every day. Has to be inside the machine, everything else is spotless. YouTube to see how to disassemble. Find the foam seal in pieces! Now I know where the black pieces are coming from. I'm ready to sue Resmed, then I see the warranty doesn't cover that because of the Soclean. Apparently the damage is caused by Ozone. So they say. Does anyone know if there will be any issues using the machine without the foam? What about a replacement foam piece? Thanks in advance!
Attachments
Resmed.jpg
Resmed.jpg (78.62 KiB) Viewed 8560 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Julie » Sun Jul 18, 2021 8:51 am

Put SoClean in the forum search bar - you'll get lots of answers!

SoClean is a scam, a waste of money and will invalidate your machine warranty. Get your money back if you can!

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Goofproof » Sun Jul 18, 2021 9:24 am

You need to find better Friends. :?
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
LSAT
Posts: 13413
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed 10 and SoClean Issue! See Pic!

Post by LSAT » Sun Jul 18, 2021 9:31 am

ozone.jpg

Medic856
Posts: 8
Joined: Sun Jul 18, 2021 7:43 am

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Medic856 » Sun Jul 18, 2021 10:03 am

It appears that the Phillips recall is for the same reason as this problem
. So why no recall for the Resmed?

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Dog Slobber » Sun Jul 18, 2021 10:13 am

Medic856 wrote: ↑
Sun Jul 18, 2021 10:03 am
It appears that the Phillips recall is for the same reason as this problem
. So why no recall for the Resmed?
No, it is not the same problem.

Philips is *not* being recalled because the foam is incompatible with ozone (all though it is). Philips is being recalled because its foam off-gases VOCs (without degrading), and it degrades even without exposure to ozone.

Ozone is incompatible with many materials, it's not product the manufacturers (ResMed) responsibility because people don't follow instructions and do whatever the hell they want.

If you dump a bunch of bleach in your laundry, is it Levi's responsibility to get you a new pair of jeans?
Last edited by Dog Slobber on Sun Jul 18, 2021 10:28 am, edited 2 times in total.
Battery Backup: EcoFlow Delta 2

User avatar
ups4
Posts: 99
Joined: Fri Jun 06, 2014 7:30 pm
Location: Sunrise, FL

Re: Resmed 10 and SoClean Issue! See Pic!

Post by ups4 » Sun Jul 18, 2021 10:16 am

Medic856 wrote: ↑
Sun Jul 18, 2021 10:03 am
It appears that the Phillips recall is for the same reason as this problem
. So why no recall for the Resmed?
ResMed has issued a statement and I got an email from them saying the type of foam they use is different from the type Philips uses which is causing the recall.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack
Additional Comments: DX 11/2013 Also DX with mild COPD----Pressure 9 cm-17 cm
My get-up-and-go musta got up and went

Medic856
Posts: 8
Joined: Sun Jul 18, 2021 7:43 am

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Medic856 » Sun Jul 18, 2021 10:41 am

ups4 wrote: ↑
Sun Jul 18, 2021 10:16 am
Medic856 wrote: ↑
Sun Jul 18, 2021 10:03 am
It appears that the Phillips recall is for the same reason as this problem
. So why no recall for the Resmed?
ResMed has issued a statement and I got an email from them saying the type of foam they use is different from the type Philips uses which is causing the recall.
Maybe so. But it looks the same.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Resmed 10 and SoClean Issue! See Pic!

Post by squid13 » Sun Jul 18, 2021 11:27 am

Don't blame ResMed cause you used a SoClean machine and ruined your cpap machine.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Resmed 10 and SoClean Issue! See Pic!

Post by ILoveFlowers » Sun Jul 18, 2021 11:29 am

Why do you think Resmed should be sued and not SoClean, which is the real problem here? There is no reason to be using SoClean. Doing so puts a lot of wear and tear on machines and it is completely unnecessary. Why does anyone use it?
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Resmed 10 and SoClean Issue! See Pic!

Post by squid13 » Sun Jul 18, 2021 11:47 am

You might want to read this. If your machine falls in this time frame SoClean will pay for the repairs. https://hartmedical.org/blog/post/socle ... rification

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Resmed 10 and SoClean Issue! See Pic!

Post by chunkyfrog » Sun Jul 18, 2021 1:18 pm

Tell all your friends to avoid Respironics and Noclean.
You don't have to buy a Resmed to get away from "foamgate",
but it's probably the best alternative--if you can $wing it.
Other brands, like Devillbiss, will be grateful for your business,
and you don't have to wait a year or more, or for the FDA, or lawyers, etc.
This is your health--put it on the top of the "to-do" list.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 10 and SoClean Issue! See Pic!

Post by palerider » Sun Jul 18, 2021 11:18 pm

Medic856 wrote: ↑
Sun Jul 18, 2021 10:41 am
ups4 wrote: ↑
Sun Jul 18, 2021 10:16 am
Medic856 wrote: ↑
Sun Jul 18, 2021 10:03 am
It appears that the Phillips recall is for the same reason as this problem
. So why no recall for the Resmed?
ResMed has issued a statement and I got an email from them saying the type of foam they use is different from the type Philips uses which is causing the recall.
Maybe so. But it looks the same.
Idiocy, pure and simple.

"that's not gold, it's not worth anything"
"but it looks the same"
"GET OUT"

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Carleton
Posts: 21
Joined: Thu Jul 15, 2021 11:55 am

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Carleton » Mon Jul 19, 2021 12:35 am

Neat post! I had just finished watching this video that shows the effects of ozone on the foam in Dreamstation 1 & 2, and Airsense 10.

https://www.youtube.com/watch?v=DbHcZUFRDqg

The Resmed foam appeared to more or less hold up in his experiment, but it looks by your experience that may not be the case over the longterm.

I'll just keep cleaning with soap and water. :D

Coffee Man
Posts: 28
Joined: Sun Mar 11, 2018 1:14 pm

Re: Resmed 10 and SoClean Issue! See Pic!

Post by Coffee Man » Mon Jul 19, 2021 4:33 am

Medic856 wrote: ↑
Sun Jul 18, 2021 8:04 am
Does anyone know if there will be any issues using the machine without the foam?
I removed the foam from my ResMed AirSense 10 and have been using it without any issue. No difference in pressure or AHI compared to using the machine with the foam. That’s been my experience.