Best battery and inverter to use w/ cpap when dry camping
-
Longbeard1
- Posts: 4
- Joined: Fri Jul 09, 2021 2:52 pm
Best battery and inverter to use w/ cpap when dry camping
It is always a challenge for us to dry camp and run the cpap and maintain a charge on our battery. She has a Resmed Airsense 10 machine with a 750W Everstart Inverter and a series 27 RV/Marine 675CCA battery. We bring the battery and inverter next to the bed so there is no long cords. With battery at full charge 13.1V it will run down to 11.7V after the first night and the alarm will go off during the next night of use. Why can't we get more than two nights out of a fully charged battery? The humidifier is off, has very little air leaks and is set at about 12. Would we be better off with a 100W inverter and AGM 6V battery? Or maybe two 6V batteries? We charge them during the day with our Honda 2000 generator but can't keep up. Help
Re: Best battery and inverter to use w/ cpap when dry camping
Inverters are very inefficient, get the proper 12v cable for the CPAP and you'll likely double your runtime.Longbeard1 wrote: ↑Fri Jul 09, 2021 3:11 pmIt is always a challenge for us to dry camp and run the cpap and maintain a charge on our battery. She has a Resmed Airsense 10 machine with a 750W Everstart Inverter and a series 27 RV/Marine 675CCA battery. We bring the battery and inverter next to the bed so there is no long cords. With battery at full charge 13.1V it will run down to 11.7V after the first night and the alarm will go off during the next night of use. Why can't we get more than two nights out of a fully charged battery? The humidifier is off, has very little air leaks and is set at about 12. Would we be better off with a 100W inverter and AGM 6V battery? Or maybe two 6V batteries? We charge them during the day with our Honda 2000 generator but can't keep up. Help
Also, you've got the wrong battery, you want a deep cycle battery, not a starting battery (as evidenced by the 'cca' rating. Avoid anything "marine" because those are heavier because they're built to handle the pounding on boats.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Best battery and inverter to use w/ cpap when dry camping
Couldn't agree more.Inverters are very inefficient, get the proper 12v cable for the CPAP
I would recommend a *Deep Cycle* battery - choose what ever size you feel comfy carrying. I have a 100AH iirc it weighs about 62#.
Hopefully any heated hose is also OFF.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Best battery and inverter to use w/ cpap when dry camping
Also, turn the heated humidifier to zero.
You can still use passover humidity, but heating water sucks electricity--big time.
You can still use passover humidity, but heating water sucks electricity--big time.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
-
Longbeard1
- Posts: 4
- Joined: Fri Jul 09, 2021 2:52 pm
Re: Best battery and inverter to use w/ cpap when dry camping
Thank you for the responses. What do you mean when you recommend a proper 12V CPAP cable? Right now the inverter cables(about) a foot long are attached to the battery. Then the CPAP is plugged into the inverter.
Re: Best battery and inverter to use w/ cpap when dry camping
You want Resmed part # 37297. Check Cpap.com, amazon, or any other provider:
https://support.resmed.com/en-us/access ... converter/

https://support.resmed.com/en-us/access ... converter/

Re: Best battery and inverter to use w/ cpap when dry camping
You can (and should) power the cpap directly from the battery, WITHOUT an inverter.Longbeard1 wrote: ↑Mon Jul 12, 2021 5:07 pmThank you for the responses. What do you mean when you recommend a proper 12V CPAP cable? Right now the inverter cables(about) a foot long are attached to the battery. Then the CPAP is plugged into the inverter.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Longbeard1
- Posts: 4
- Joined: Fri Jul 09, 2021 2:52 pm
Re: Best battery and inverter to use w/ cpap when dry camping
Great info. Thanks. Does anyone have a preference for 6V versus 12V deep cycle batteries? If I buy a 6V battery can I plug it into the 90W converter or do I need something else?
Re: Best battery and inverter to use w/ cpap when dry camping
You'd need two, in series, to get 12 volts. volts matter.Longbeard1 wrote: ↑Tue Jul 13, 2021 10:59 amGreat info. Thanks. Does anyone have a preference for 6V versus 12V deep cycle batteries? If I buy a 6V battery can I plug it into the 90W converter or do I need something else?