Cannot control pressure flow

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CpapCraig
Posts: 1
Joined: Mon Jul 05, 2021 10:25 pm

Cannot control pressure flow

Post by CpapCraig » Mon Jul 05, 2021 10:30 pm

I have a Philips Respironics System One A Flex APaP. I recently removed the cancer foam. It seems to work fine, EXCEPT the air pressure is insane. It reads 4.0, but is blowing like 20.0. I'm wondering if there's something I missed during reassembly that could have caused this, or if there's a way I can reboot the system to fix it? Changing the pressure setting via Setup has no effect. Thanks in advance for any assistance.

User avatar
Pugsy
Posts: 65390
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Cannot control pressure flow

Post by Pugsy » Mon Jul 05, 2021 10:34 pm

Get a manometer and verify pressure first of all.
You can buy them or even make one.
There are some DIY instructions here on the forum to make one but I am sorry...don't have that link handy.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Cannot control pressure flow

Post by Wulfman... » Mon Jul 05, 2021 11:24 pm

Here ya go.

search.php?keywords=%22homemade+manometer%22

By the way........What is/are your pressure setting(s)?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Cannot control pressure flow

Post by palerider » Mon Jul 05, 2021 11:37 pm

CpapCraig wrote:
Mon Jul 05, 2021 10:30 pm
I have a Philips Respironics System One A Flex APaP. I recently removed the cancer foam. It seems to work fine, EXCEPT the air pressure is insane. It reads 4.0, but is blowing like 20.0. I'm wondering if there's something I missed during reassembly that could have caused this, or if there's a way I can reboot the system to fix it? Changing the pressure setting via Setup has no effect. Thanks in advance for any assistance.
If you didn't get the little rubber plugs between the board and the plastic module back in place properly, and seated on their appropriate sensors, then the machine has no idea what pressure it's generating.

_________________