Inline Hose Filter - Any Reduction of Air Flow?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Grumpy48
Posts: 463
Joined: Tue May 12, 2020 7:57 am

Inline Hose Filter - Any Reduction of Air Flow?

Post by Grumpy48 » Tue Jun 22, 2021 11:44 am

Contemplating adding an inline hose filter to my DreamStation. A few I've looked at on Amazon a smattering of the reviews mention the air flow was reduced and the CPAP didn't work correctly to provide therapy, otherwise most reviews were good and the filter did what it was supposed to.
Other than a filter getting plugged up with crap is there any likelihood a fresh inline hose filter could alter air flow enough to change the therapy delivered?

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by jimbud » Tue Jun 22, 2021 1:28 pm

Manometer

A manometer is an instrument for measuring low-range positive, negative, and differential air and gas pressure. It has no moving parts and takes measurements using a tube containing water or another liquid. Manometers are commonly used for gauging airflow through air filters and measuring small pressure differences in laboratory applications.

Amazon has them.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by chunkyfrog » Tue Jun 22, 2021 1:55 pm

Or you can build your own.
Instructions on the internet.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Grumpy48
Posts: 463
Joined: Tue May 12, 2020 7:57 am

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Grumpy48 » Wed Jun 23, 2021 7:55 am

Presuming I either get or build a manometer, with the measured results I would get, how do I translate that information to something that would determine if the functioning of my DreamStation would affected and thus affect the therapy. What would be the resulting pressure differential numbers that could affect delivered therapy.

My question was more of a polling of users that may have had issues of adding a hose filter that may have caused a problem with their CPAP therapy.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Pugsy » Wed Jun 23, 2021 8:25 am

Grumpy48 wrote: ↑
Wed Jun 23, 2021 7:55 am
My question was more of a polling of users that may have had issues of adding a hose filter that may have caused a problem with their CPAP therapy.
You are asking a question from a very limited pool of forum members. Not many use the additional in line bacterial filter ....at least here on the forum.
I personally know of one forum member that comes to mind. That's it. He uses it for additional allergy/pollen filtering. There are probably a handful more that I don't know about.
If someone uses that in line filter...they can't be using the heated hose.
So hence probably not many have even tried using that filter. They didn't want to give up the heated hose much less worry about maybe slightly less pressure coming out at the mask end.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Dog Slobber » Wed Jun 23, 2021 8:47 am

Given that when one uses the inline filter on a ResMed 10, a change needs to be made in the settings. This suggests that there is some impact by an inline filter.

Adding an AB filter, requiring a ResMed settings change may be unique to ResMeds, and not impact Respironics.
Battery Backup: EcoFlow Delta 2

Grumpy48
Posts: 463
Joined: Tue May 12, 2020 7:57 am

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Grumpy48 » Wed Jun 23, 2021 8:58 am

Yup, I know the heated hose wouldn't be an option, but could likely forego with that at least for the warmer summer months when hose rainout should be less of a problem. The hose temp setting is at minimum presently and no problems, so should be able to go to a non-heated hose.

I've feeling users of Philips machines may likely consider a filter as an option to capture possible foam particulates, rather than just stop using their machine per Philips. Granted a filter may do nothing for whatever minimal offgassing there may be from the foam, but my own feeling is the degree of VOC's that could be inhaled is not likely to cause a health issue and per Philips own data from the recall would seem to be the case.

I do find it odd/amusing that the hose filter is referred to as a 'bacteria filter'. Makes me wonder if folks use swamp water in their humidifiers :wink:

GrumpyHere
Posts: 523
Joined: Sat Feb 13, 2021 2:40 am

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by GrumpyHere » Wed Jun 23, 2021 9:29 am

Grumpy48 wrote: ↑
Wed Jun 23, 2021 8:58 am
Yup, I know the heated hose wouldn't be an option
On masks where hose attachment is above your head, you likely could added it to the mask end and still use a heated hose.
ResMed Lumis Tx

Grumpy48
Posts: 463
Joined: Tue May 12, 2020 7:57 am

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Grumpy48 » Wed Jun 23, 2021 9:52 am

GrumpyHere wrote: ↑
Wed Jun 23, 2021 9:29 am
Grumpy48 wrote: ↑
Wed Jun 23, 2021 8:58 am
Yup, I know the heated hose wouldn't be an option
On masks where hose attachment is above your head, you likely could added it to the mask end and still use a heated hose.
Most of the filters I looked at have a flow direction arrow. Connecting the filter at the mask end would necessitate the flow direction to be reversed in the filter because of connection size on hose and mask. I have no idea if the flow direction makes really a difference. My F&P Vitera mask has a connection below the nose.

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by ILoveFlowers » Wed Jun 23, 2021 10:39 am

I tried one several years ago. It's been a while so I forget the details, but I didn't use it long because I did feel like it affected the therapy. I can't remember specifically if the machine felt sluggish or the pressure seemed affected but it was something like that. Maybe I could have messed around with increasing pressure or something but I just let it go. It might be worth it to try again now with the recall and try harder to make it work.
AirSense 10 AutoSet For Her
Airfit P10 mask

Desert mans wife
Posts: 19
Joined: Thu May 06, 2021 5:07 pm

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Desert mans wife » Wed Jun 23, 2021 11:26 am

I used one a couple nights ago with my Dreamstation. Seemed to be a slight uptick with my AHI, but I'm still a newbie (1 month in - and had the bad luck to get the Dreamstation), and it seems like everything affects it. So I don't really know if the filter made any difference with my therapy. I didn't notice any changing with my breathing, the pressure didn't seem any different. I have the generic bacteria filters from CPAP.com.

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by JLROhio » Wed Jun 23, 2021 11:32 am

Dog Slobber wrote: ↑
Wed Jun 23, 2021 8:47 am
Given that when one uses the inline filter on a ResMed 10, a change needs to be made in the settings. This suggests that there is some impact by an inline filter.

Adding an AB filter, requiring a ResMed settings change may be unique to ResMeds, and not impact Respironics.
^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
This is a VERY important point - I'm glad someone mentioned it!

I'd like to know...from those who have used an AB (anti-bacterial ??) filter, AND have made the changes in their ResMed menu to include the fact that you have an AB filter...did it change anything that you noticed? I'm wondering what the AB Filter button on the ResMed's actually changes? Extra Pressure to compensate??
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Dog Slobber » Wed Jun 23, 2021 11:56 am

JLROhio wrote: ↑
Wed Jun 23, 2021 11:32 am
Dog Slobber wrote: ↑
Wed Jun 23, 2021 8:47 am

Adding an AB filter, requiring a ResMed settings change may be unique to ResMeds, and not impact Respironics.
^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
This is a VERY important point - I'm glad someone mentioned it!

I'd like to know...from those who have used an AB (anti-bacterial ??) filter,
Yes, AB stands for AntiBacterial, at least in the Air 10 documentation, and in the setup.

Surprisingly, in the Air 11 setup, it has been changed to B/V filter. I assume that is Bacteria/Virus.
I'm wondering what the AB Filter button on the ResMed's actually changes? Extra Pressure to compensate??
Could be anything, pressure compensation, flow limitation compensation, OA/CA/H detection or differentiation, leak rate calculation.
Battery Backup: EcoFlow Delta 2

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by squid13 » Wed Jun 23, 2021 12:12 pm

I've been using the AB filter with a heated hose for years. I use it on the end of the hose and then attach my mask hose to the other end of the filter. Just be sure to make the change in the machine.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

Grumpy48
Posts: 463
Joined: Tue May 12, 2020 7:57 am

Re: Inline Hose Filter - Any Reduction of Air Flow?

Post by Grumpy48 » Wed Jun 23, 2021 12:26 pm

Thank for the replies from those that have tried a filter.

The Dreamstation does not have a setting I could find like that of the Resmed for use of a AB or BV filter, but does have a setting to compensate for different mask types to 'fine tune the level of air pressure compensation'. Perhaps if a compensation would be needed for an AB filter, perhaps this setting might allow some trial and error tweaking to compensate for the air flow when a filter is used. I don't have a Philips mask, thus no recommended mask setting and just use the machine default. Have not tried playing with it to see how it would affect therapy.

'Adjusting mask and tube type on the DreamStation | Philips | Sleep therapy system'
https://www.youtube.com/watch?v=-rNT5y4Yi_c