Humidifier heater broken, aircurve V10 auto ?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Dusk2Dawn
Posts: 4
Joined: Fri Apr 30, 2021 3:18 pm

Humidifier heater broken, aircurve V10 auto ?

Post by Dusk2Dawn » Fri May 28, 2021 1:28 pm

I think the humidifier heating element may have gone out on my Resmed Aircurve V10 Auto , unless I have somehow turned it off. Just wondering if anyone else have their heater go out on their unit ( about 5-7 yrs old ) when all else works fine ?

Under "My Options " I have set the humidifer to level 2, then 3, then 4, etc.... and it shows that level on the screen [ My Options... humidity level ] , and yet the water never seems to get warm or seems to evaporate much [ I believe the bottom of the chamber felt warm to the touch before at these levels , but not now ]. Besides at level 4, the water would be evaporated quite a bit by morning, not so now.

I will try a higher level tonight, like 6 or 7, just to see if it gets warm. Is there any other setting I'm missing for the heated humidifier, or any diagnostic that I can run to check it it's working ?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Humidifier heater broken, aircurve V10 auto ?

Post by Julie » Fri May 28, 2021 1:32 pm

Get a replacement from your dealer, the same way you would for anything else you buy (or rent) that's defective.

User avatar
BlueDragon
Posts: 567
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Humidifier heater broken, aircurve V10 auto ?

Post by BlueDragon » Fri May 28, 2021 2:26 pm

Dusk2Dawn wrote: ↑
Fri May 28, 2021 1:28 pm
I think the humidifier heating element may have gone out on my Resmed Aircurve V10 Auto , unless I have somehow turned it off. Just wondering if anyone else have their heater go out on their unit ( about 5-7 yrs old ) when all else works fine?
I have seen other reports of this happening with failure being repaired by replacement of a thermal fuse. Search YouTube for "resmed heater repair" and you'll find at least one video (unfortunately, it is relatively low quality). I've seen other reports of users not being able to repair the problem.

At 5-7 years old, your machine is probably out of warranty. If you have insurance, they may replace the machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Humidifier heater broken, aircurve V10 auto ?

Post by Dog Slobber » Fri May 28, 2021 2:43 pm

Here's a good quality video that shows where the fuse is, and how easy it is to swap the heater plate completely. Humidifier parts are compatible between all Air10 devices.

The humidifier portion is at 2:20


https://youtu.be/M9jdc9_WmwM?t=120
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Humidifier heater broken, aircurve V10 auto ?

Post by Pugsy » Fri May 28, 2021 4:19 pm

What is the ambient humidity in your house?
If it is very high then the machine may not be heating and using much water because the ambient humidity is already fairly high and it just doesn't need to add much moisture to the air.

It is also possible that the heater element has died.
Do a little test...turn everything on and use the machine while awake and check the bottom of the water chamber for warmth after 15 minutes. Don't wait too long though. It goes through cycles of heating being on and off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Humidifier heater broken, aircurve V10 auto ?

Post by squid13 » Fri May 28, 2021 6:24 pm

If that fuse is blown, any ideas where you can get a replacement or a new plate?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

RNeil
Posts: 79
Joined: Tue Jul 31, 2018 7:58 am
Contact:

Re: Humidifier heater broken, aircurve V10 auto ?

Post by RNeil » Fri May 28, 2021 8:15 pm

Replacing the fuse will be harder than it sounds. It was not designed to be easily replaced. I am sure that Digikey would have a replacement, but they won't identify it as a Resmed part.