Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Decameron
Posts: 4
Joined: Sun Apr 25, 2021 9:24 am

Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Decameron » Sun Apr 25, 2021 10:19 am

Hi all

I have just found and joined the site and also downloaded Oscar and my most recent data. I had a ResMed S9 AutoSet for about 5 years and was very happy with it getting my average AHI down to just under 2. As the machine was getting old my provider gave me a new machine - a Dreamstation Auto CPAP. This was about six weeks ago. I have struggled with the new machine and have been waking with headaches and not feeling as refreshed as I used to with the ResMed. I am currently getting average AHI figures of about 5. I have met with my medical person and they can't seem to provide any ideas about how to improve matters other than to give it more time. Its set on opti-start and ramp time is 0. I generally fall asleep in a couple of minutes. I have heated tube humidification and humidifier set around 3. Flex is turned off. I can't understand why I am finding it so different to my old ResMed, it feels like I am not getting as much pressure with the new machine and that its harder breathing in. I have a mirage FX mask. Could it be due to the different algorithms being used by Philips and ResMed? My current provider now only uses Philips machines and I don't know if I should persevere with the hope that things improve or if I should move to a provider who can supply me with a ResMed machine. Any assistance and guidance warmly appreciated. Snip of a daily result follows. Peter
Screenshot 2021-04-25 at 17.07.01.png
Screenshot 2021-04-25 at 17.07.01.png (934.24 KiB) Viewed 2063 times

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Pugsy » Sun Apr 25, 2021 10:38 am

Turn off Opti Start for one thing and set the machine to regular auto mode.

Is EZ Start turned on? If so....turn it off.

What were your settings on your old machine? Do you know? Do you still have it?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Decameron
Posts: 4
Joined: Sun Apr 25, 2021 9:24 am

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Decameron » Tue Apr 27, 2021 12:09 pm

Thanks , I will turn off Opti Start and try regular auto mode. EZ start is off. The hospital took back my old Res Med when they gave me the new Dreamstation so I no longer have it and I don't know what settings it was on. Because I never had problems with the ResMed I never paid much attention to the settings I just left it to get on with its work. I could ask at my next hospital visit to see if they have details of the settings on my old machine.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by squid13 » Tue Apr 27, 2021 1:15 pm

You need to keep your old machine as a backup. If you paid for that machine they had no right to take it cause it's your's.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Pugsy » Tue Apr 27, 2021 2:03 pm

Where are you located?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Decameron
Posts: 4
Joined: Sun Apr 25, 2021 9:24 am

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Decameron » Thu Apr 29, 2021 2:17 pm

I am based in the uk and get my equipment through the NHS at no charge. I saw a technician yesterday and following a discussion they changed the settings on the machine. They told me the pressure has now been set to 10 but that it would go up and down three ie down to 7 and up to 13 as required. I tried it last night and had better sleep and an AHI of 2.8, which is a significant improvement. Looking at Oscar it shows that the pressure has just been set to run at 10 all night. It didn't go up or down. I presume if this works for me that I can just leave it at this fixed pressure without it having any bad effects in either the long or short term? The technician did say that if the new setting doesn't work for me I can go back to my doctor and ask to be transferred to a hospital that uses ResMeds. Copy of last nights Oscar results attached.
Screenshot 2021-04-29 at 21.06.20.png
Screenshot 2021-04-29 at 21.06.20.png (902.62 KiB) Viewed 1977 times

User avatar
zonker
Posts: 11500
Joined: Fri Jun 19, 2015 4:36 pm

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by zonker » Thu Apr 29, 2021 4:21 pm

Decameron wrote: ↑
Thu Apr 29, 2021 2:17 pm
I am based in the uk and get my equipment through the NHS at no charge. I saw a technician yesterday and following a discussion they changed the settings on the machine. They told me the pressure has now been set to 10 but that it would go up and down three ie down to 7 and up to 13 as required. I tried it last night and had better sleep and an AHI of 2.8, which is a significant improvement. Looking at Oscar it shows that the pressure has just been set to run at 10 all night. It didn't go up or down. I presume if this works for me that I can just leave it at this fixed pressure without it having any bad effects in either the long or short term? The technician did say that if the new setting doesn't work for me I can go back to my doctor and ask to be transferred to a hospital that uses ResMeds. Copy of last nights Oscar results attached. Screenshot 2021-04-29 at 21.06.20.png
unfortunately, either that technician is incompetent or s/he is telling porkies.

as you said, your machine staid at a pressure of 10 all night, per the pressure graph.

yet if you look at the events graph, you had several hypopnea events throughout the night and the pressure did NOT go up to combat said hypopnea. in fact, when there is but a single pressure selected and that selection is ten, the machine CAN'T vary in either direction.


the machine you are using is capable of being set up so you have differing lows and highs. that is, you could be setup to have say a low of seven and a high of twenty. with that kind of setting, your machine would be free to go wherever the pressure needs to go to vanquish those hypopneas and thus lower your ahi.

does that make sense?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Decameron
Posts: 4
Joined: Sun Apr 25, 2021 9:24 am

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Decameron » Sat May 08, 2021 11:01 am

Thanks, yes this makes complete sense. I will give them a call and ask them to change the settings so it can go up and down a bit from the level 10 pressure. I am reluctant to try to change the settings myself if they can't get it right!

User avatar
Whale Road
Posts: 33
Joined: Tue Aug 25, 2015 9:44 am

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by Whale Road » Sat May 08, 2021 11:07 am

Decameron wrote: ↑
Sat May 08, 2021 11:01 am
I am reluctant to try to change the settings myself if they can't get it right!
Huh?? The reason most of us are here is that we can get the setting right when the medical profession can't.

Learn to use OSCAR and take control of your own therapy. You will be happy you did.

User avatar
zonker
Posts: 11500
Joined: Fri Jun 19, 2015 4:36 pm

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by zonker » Sat May 08, 2021 11:17 am

Decameron wrote: ↑
Sat May 08, 2021 11:01 am
Thanks, yes this makes complete sense. I will give them a call and ask them to change the settings so it can go up and down a bit from the level 10 pressure. I am reluctant to try to change the settings myself if they can't get it right!
i'm in agreement with whale road below. we mostly do the changing ourselves.

but i can understand your reluctance. in any case, please come back and let us know how you make out, okay?

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: Moved from ResMed S9 AutoSet to Dreamstation AutoCPAP and having problems

Post by clownbell » Sat May 08, 2021 12:41 pm

I don't know about the laws in the UK. Here in the USA, it is most definitely NOT illegal for the user to change settings. When you think about it, the doc and techs don't have any skin in the game. If you don't sleep well or your symptoms are not mitigated, their life is unchanged. YOU are the person impacted, so YOU should be the person in charge. I don't know anything about the Dreamstation but I believe you can get the clinical manual and begin to monitor and change your own settings. Just remember YOU are the boss. The docs and tech get grumpy because their position of prominence is eroded, which is their problem not yours.
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask