Apnic episodes with chest pounding

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Kewilso1
Posts: 2
Joined: Wed Apr 07, 2021 7:25 pm

Apnic episodes with chest pounding

Post by Kewilso1 » Thu Apr 08, 2021 6:29 pm

I have rem sleep osa. I wake up with all episodes with chest pounding. I use apap pressures 12-20. My seal is good with a nasal pillow but I still have episodes.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Apnic episodes with chest pounding

Post by Julie » Thu Apr 08, 2021 8:21 pm

A little info please, like... have you tried different (possibly lower min.) pressures?
Which apap are you using? Mask? Take any meds?

Can you post a couple of Oscar charts in this thread so we can see what's happening?

How do you know you have REM osa?

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: Apnic episodes with chest pounding

Post by zonker » Thu Apr 08, 2021 8:55 pm

Kewilso1 wrote:
Thu Apr 08, 2021 6:29 pm
I have rem sleep osa. I wake up with all episodes with chest pounding. I use apap pressures 12-20. My seal is good with a nasal pillow but I still have episodes.
welcome to the zoo!

apnic as far as i can see is another term for apnea. so you're in the right place.

please give us some more info. not everyone has the same machine or mask. so it would be helpful if you can supply us with that information. it would be great if the machine you have is compatible with the sleep apnea software, oscar.

how did it come about that your pressure settings are 12 and 20? again, specific settings are generally used in each individual case.

let us know more so that we can help.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Apnic episodes with chest pounding

Post by khauser » Fri Apr 09, 2021 11:19 am

Kewilso1 wrote:
Thu Apr 08, 2021 6:29 pm
I have rem sleep osa. I wake up with all episodes with chest pounding. I use apap pressures 12-20. My seal is good with a nasal pillow but I still have episodes.
So you have an APAP, which is good. Can you use Oscar and get the data from it? It would be useful to try and determine if your apnea is obstructive or central in nature. That will likely directly impact the guidance.