To address a couple of points if I maypalerider wrote: ↑Tue Apr 06, 2021 7:34 amThis is incorrect, the pressure at the mask is NOT equal to the pressure in the mask, unless no air is flowing through the patient circuit.
Resistances to airflow cause lowered pressure (to avoid cough drop comments) at each stage of the circuit between the blower motor itself and your face. whether you have a humidifier attached, what diameter and length of tube, filters, and mask type (ever noticed the "System One Resistance"?) all affect the actual pressure, *AT THE MASK*. The machines have to compensate for these flow resistances by increasing the pressure at the machine.Totally incorrect, the machine increases the FLOW to compensate for leaks, in order to *maintain* pressure.
Again, incorrect. many Philips machines never even report mask pressure, just target pressure.
Most doctors don't have any idea what 'mask pressure' is.
"the pressure at the mask is NOT equal to the pressure in the mask"
It's as near as makes no difference. Obviously there are *very* slight pressure variances, and obviously I'm aware some are caused depending on what components are present in the circuit - but these are accounted for as the machine knows which components are connected and the mask type (resistance) is user specified. The point is, we're talking about a 14 cmH20 difference between actual and that reported in OSCAR here.
"Totally incorrect, the machine increases the FLOW to compensate for leaks, in order to *maintain* pressure."
OK, if we were being pedantic about it you could state it that way, however my assertion was indeed correct. By increasing flow, it is of course increasing pressure to compensate for the drop in pressure caused by the leak. A PAP unit is just a 'fan in a box'.. not exactly rocket science.
"Again, incorrect. many Philips machines never even report mask pressure, just target pressure."
Mask pressure is of course the only figure that truly matters - whether it's close to the target pressure or not. It doesn't matter what's 'reported', the actual delivered pressure is based on measurement within a closed circuit. Anyway, as we're talking about a 14cm difference that's a different story...