Huge amounts of pure central apneas (~50 per hour) with high pressure & EPR

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
msreef
Posts: 17
Joined: Mon Nov 09, 2015 3:18 am

Huge amounts of pure central apneas (~50 per hour) with high pressure & EPR

Post by msreef » Wed Mar 10, 2021 8:35 am

KInda freaking out a bit now..
Have been playing with my machine in attempts to fix what I believe is undiagnosed UARS, as I'm not getting as effective treatment as what I'd like.
I set the pressure to 19 and EPAP of 9, so effectively a PS of 10.
Woke up after around 4 hours with really sore aerophagia and lots of gas and looked at my machine, AHI of ~50 of purely centrals. Confirmed via OSCAR.
When I have the pressures much lower (around 9/12) I get basically zero centrals each night, maybe one or two max.

Is this anything to be cornered about (heart failure etc), or can I put it simply down to the crazy high pressure?

Note that may tidal volume is ~700, it's typically like ~460.
Image

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Huge amounts of pure central apneas (~50 per hour) with high pressure & EPR

Post by Pugsy » Wed Mar 10, 2021 9:14 am

PS of 10 is causing your centrals. It's causing you to sort of hyperventilate and blow of too much carbon dioxide.

It was the wrong experiment for you to try. That much PS used (unless someone has major respiratory issues) is going to trigger centrals.

More IPAP this way is NOT the way to deal with UARS.

Go back to more normal settings.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

msreef
Posts: 17
Joined: Mon Nov 09, 2015 3:18 am

Re: Huge amounts of pure central apneas (~50 per hour) with high pressure & EPR

Post by msreef » Wed Mar 10, 2021 9:17 am

Pugsy wrote:
Wed Mar 10, 2021 9:14 am
PS of 10 is causing your centrals. It's causing you to sort of hyperventilate and blow of too much carbon dioxide.

It was the wrong experiment for you to try. That much PS used (unless someone has major respiratory issues) is going to trigger centrals.

More IPAP this way is NOT the way to deal with UARS.

Go back to more normal settings.
Thank you, now I can go back to sleep and stop excessively worrying!