airsense 10 time to write/upload data

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

airsense 10 time to write/upload data

Post by willy737 » Tue Feb 16, 2021 7:16 am

Just got an airsense 10, and noticed it's taking quite some time to write data to the sdcard, as well as upload to myair. This morning I waited until it was done with it's cooldown cycle to pull the sdcard, but it still doesn't have the data from last night. I've been coming back in the afternoons to pull it and it's there by then. Does it usually take an hour or so for the data to be written to the card?

User avatar
LSAT
Posts: 13414
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: airsense 10 time to write/upload data

Post by LSAT » Tue Feb 16, 2021 8:14 am

I'm not sure, but since daily data is captured from noon to noon, the system may wait until after 12 to be sure it is getting the entire days data. I never look at mine on a daily basis...I just look at nightly AHI when I get up. Every week or two I download my data to Oscar.

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: airsense 10 time to write/upload data

Post by willy737 » Tue Feb 16, 2021 9:59 am

Ok that makes sense. Thanks

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: airsense 10 time to write/upload data

Post by Dog Slobber » Tue Feb 16, 2021 10:00 am

The data is written to the SD Card at the end of a session, there isn't a (measurable) delay.

Why do you think it's not immediately writing to the SD CARD.

With MyAir, there is a day delay upon first signing up, then about an hour after the session end my data shows up.
Battery Backup: EcoFlow Delta 2

User avatar
Nooblakahn
Posts: 426
Joined: Sat Feb 02, 2013 10:22 pm
Location: Denver, Colorado

Re: airsense 10 time to write/upload data

Post by Nooblakahn » Tue Feb 16, 2021 10:31 am

Yeah as others have said. I've not seen ac delay in the card being written to. Usually an hour after I shut the machine off the data is in my air

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP™ Adapt SV
“I wonder if I've been changed in the night. Let me think. Was I the same when I got up this morning? I almost think I can remember feeling a little different."
― Lewis Carroll, Alice in Wonderland

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: airsense 10 time to write/upload data

Post by Pugsy » Tue Feb 16, 2021 10:59 am

I have never had any delay in the machine writing to the SD card.
As to reporting to MyAir...never used it so don't know.

The noon reporting thing for LCD sleep report....doesn't really apply to the AirSense/AirCurve models...only to older models.
I checked last night on my AirCurve sleep report from the prior night and the report was still available at 8PM.
While the machine might start a new session at noon it doesn't create the 0.0 usage that we saw after 12 PM with the S9 and older model machines.

12 Noon really isn't any factor with the AirSense 10 machines in terms of showing sleep report or writing to the SD card or transmitting data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: airsense 10 time to write/upload data

Post by Pugsy » Tue Feb 16, 2021 11:01 am

willy737 wrote: ↑
Tue Feb 16, 2021 7:16 am
This morning I waited until it was done with it's cooldown cycle to pull the sdcard, but it still doesn't have the data from last night.
This is not normal. The data should be on the SD card within 5 minutes at the most.
Something weird is going on.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: airsense 10 time to write/upload data

Post by Dog Slobber » Tue Feb 16, 2021 11:29 am

Pugsy wrote: ↑
Tue Feb 16, 2021 11:01 am
willy737 wrote: ↑
Tue Feb 16, 2021 7:16 am
This morning I waited until it was done with it's cooldown cycle to pull the sdcard, but it still doesn't have the data from last night.
This is not normal. The data should be on the SD card within 5 minutes at the most.
Something weird is going on.
There is a bug in OSCAR; if the season end time is later than the computer time, the data will NOT get imported.

The work around is to ensure the time on the ResMed is aligned with your computer, or wait an hour, if there is an hour difference because of DST changes.

This bug is being addressed in future version.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: airsense 10 time to write/upload data

Post by Pugsy » Tue Feb 16, 2021 11:54 am

Dog Slobber wrote: ↑
Tue Feb 16, 2021 11:29 am
Pugsy wrote: ↑
Tue Feb 16, 2021 11:01 am
willy737 wrote: ↑
Tue Feb 16, 2021 7:16 am
This morning I waited until it was done with it's cooldown cycle to pull the sdcard, but it still doesn't have the data from last night.
This is not normal. The data should be on the SD card within 5 minutes at the most.
Something weird is going on.
There is a bug in OSCAR; if the season end time is later than the computer time, the data will NOT get imported.

The work around is to ensure the time on the ResMed is aligned with your computer, or wait an hour, if there is an hour difference because of DST changes.

This bug is being addressed in future version.

Ahhhh. I forgot about that little bug.....But how does the session end time be later than the computer time....if the clock is set correctly? I am now wondering if the OP in this thread maybe has the date on the calendar not set correctly?????
And even if the clock on the machine is an hour off because of DST....wouldn't all we need to do is wait an hour to download? We shouldn't need to wait any longer should we?

I haven't tested things in a long time though. I am doing good to download once in 3 months and I have no idea what my machine clock is set to. I will try to remember to check it out later and then if the clock is off maybe try to remember to try to download first thing in the morning.

Right now I am too busy trying to figure out if I need to take a hit out on you guys in Canada for not blocking this shit from coming down here and messing with my life so much. :lol: :lol: Seeing the data off my machine from last night didn't even make my "to do" list.
We are under a threat for rolling power outages where we live and when hubby got up this morning it was -13 F. We just don't see this kind of cold down where I live which is about an hour north of the Arkansas line. Certainly not for this prolonged length of time.
If my outside kitties don't have the heat lamp on...it's dangerous for them outside and to bring them inside is dangerous because of the cat from hell not being very nice.

So if something weird is going on...maybe it's the calendar and/or clock setting on the machine????? Worth checking out.
I once accidentally set my calendar date forward and didn't realize it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: airsense 10 time to write/upload data

Post by Pugsy » Tue Feb 16, 2021 11:56 am

To OP here....

Check your machine clock setting AND the calendar setting to make sure it is correct.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: airsense 10 time to write/upload data

Post by Dog Slobber » Tue Feb 16, 2021 12:49 pm

Pugsy wrote: ↑
Tue Feb 16, 2021 11:54 am
But how does the session end time be later than the computer time....if the clock is set correctly? I am now wondering if the OP in this thread maybe has the date on the calendar not set correctly?????
And even if the clock on the machine is an hour off because of DST....wouldn't all we need to do is wait an hour to download? We shouldn't need to wait any longer should we?
Yes, ensure the two clocks are aligned and bug won't happen otherwise wait.
Pugsy wrote: ↑
Tue Feb 16, 2021 11:54 am
Right now I am too busy trying to figure out if I need to take a hit out on you guys in Canada for not blocking this shit from coming down here and messing with my life so much.
We got crapped on last night as well.

Not much of a problem for me, I'm in a Condo and don't need to shovel, we're under a pretty tight stay-at-home order because of COVID, so no need to go out.
Battery Backup: EcoFlow Delta 2

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: airsense 10 time to write/upload data

Post by willy737 » Tue Feb 16, 2021 1:14 pm

That's probably it. Didn't think to set the clock on the new machine, and it was 2 hours ahead with Oscar set to the correct local time. Reset it and will check in the morning.

Thanks for the help!

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: airsense 10 time to write/upload data

Post by willy737 » Wed Feb 17, 2021 8:59 am

Yep that was it. You guys rock, thank you!