Icon Motor Swap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
GutsyRoman
Posts: 3
Joined: Fri Feb 12, 2021 6:50 pm

Icon Motor Swap

Post by GutsyRoman » Sat Feb 13, 2021 8:56 am

I wanted to leave this here in case it helps somebody else.

I have two Fisher & Paykel Icons. One is the icon+ (plus) [ICONAAN-HT] the other is the standard icon [ICONAAN]. Both of these crapped out on me.

Icon
This unit failed me. The unit would start to power on but the motor sounded terrible and the screen would show ERROR. I used the Icon+ as the replacement.

Icon+
This unit suffered from the internal humidity sensor issue where whenever it was unplugged, I had to remove the base and blow the boards with a hairdryer to get it to turn on. This was annoying, but I kept it running this way until it completely failed to work. One night it just shut off on me and wouldn't come back on.

I only attempted this repair, because I was without a CPAP. I slept one night with a snoreRX guard so I'd have something, and the next day was terrible. Since I cannot seem to figure out how to completely disassemble these units non-destructively to see if a part burnt out on a board, the original icon seemed the most likely to attempt a repair.

According to a manual I found online https://www.cpapman.com/manuals/ICON_.pdf The error codes could be Flow Sensing Fualts, Pressure Sensing Faults, Ambient Sensing Faults, Heater Plate Faults, Motor Faults, USB Faults, Model Configuration Faults, or Power Faults.

First I tried swapping the USB from the Icon+ (InfoUSB 2) to the Icon (smartstick). This did power on the Icon initially, but the same bad motor noise came back followed with ERROR.

I took the base of the icon and manually spun the motor. There's definitely a bearing issue. The motor/blower assembly appears to be identical on both units so I decided to swap them.



These have some sort of potting material on the to cover the connections and PCB that in my experience is normally used to make service difficult. I used a pocketknife to cut away this material on the broken motor so I could figure out how it connected, and then used that information to non destructively do the same to the icon+ motor. I ended up pulling the plastic connector header off of the PCB on the Icon+. Now I know where and how to cut on the good motor.
icon_plus_motor_removed.jpg
icon motor connector destroyed
icon_plus_motor_removed.jpg (191.43 KiB) Viewed 438 times
I carefully cut the potting material away from the good motor from the icon+ to expose the PCB connector, and then carefully cut out the wire harness and it's connector. I separated the two and swapped over to the other unit.
icon_repair2.jpg
Icon+ motor connector header
icon_repair2.jpg (102.63 KiB) Viewed 438 times
I'll continue on in another message since I can only seem to attach 3 images to the post.
Attachments
icon_assembly.jpg
Pic of the inside of an Icon
icon_assembly.jpg (179.02 KiB) Viewed 438 times

GutsyRoman
Posts: 3
Joined: Fri Feb 12, 2021 6:50 pm

Re: Icon Motor Swap

Post by GutsyRoman » Sat Feb 13, 2021 2:35 pm

The icon motor harness and connector
icon_motor_connector1.jpg
icon_motor_connector1.jpg (194.07 KiB) Viewed 415 times
A better view of the connector
icon_connector.jpg
icon_connector.jpg (97.52 KiB) Viewed 415 times

I did order a new CPAP machine yesterday morning, but it won't arrive until this afternoon. I did this so I'd have *something* to use. It worked fine last night without any issues.

Note: I don't use the humidifier at all, but I wouldn't expect this repair would cause any issues with that either. In order to get this to work, I did have to use the usb stick from the Icon+ in the Icon. It's much more likely that both the motor, and original USB flash drive failed than that there's some special firmware embedded in there that requires this.
Last edited by GutsyRoman on Sat Feb 13, 2021 2:45 pm, edited 1 time in total.

GutsyRoman
Posts: 3
Joined: Fri Feb 12, 2021 6:50 pm

Re: Icon Motor Swap

Post by GutsyRoman » Sat Feb 13, 2021 2:37 pm

For fun, I disassembled the broken motor. Here's what the PCB for it looks like with the potting material removed (mostly). Note: The connector for the wiring harness broke off of the PCB when I did the initial disassembly, and the connector wires were bent upward.
IMG_20210213_143248.jpg
IMG_20210213_143248.jpg (136.22 KiB) Viewed 415 times

And the back side has the motor coils actually embedded in. I think it's pretty safe to say this is a custom part. It's actually a very clever design.
IMG_20210213_143319.jpg
IMG_20210213_143319.jpg (107.58 KiB) Viewed 415 times

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Icon Motor Swap

Post by chunkyfrog » Sat Feb 13, 2021 3:07 pm

Cutting away that potting material had to be a bear.
That stuff is 2 part EPOXY.
Most solvents remotely able to get through it can destroy the componenrs underneath.
Evil, but effective--but not for a really determined DIY-er.
Icon is not on my recommended list for a new machine.
IMHO, Resmed Airsense 10 Autosets are worth the extra coin.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

RNeil
Posts: 69
Joined: Tue Jul 31, 2018 7:58 am
Contact:

Re: Icon Motor Swap

Post by RNeil » Mon Feb 15, 2021 9:09 am

Thankyou for an interesting post. I used to have a F&P Icon+, but now use a Resmed Autoset S10. I agree that it is better than F&P.

I have worked with non-CPAP electric motors. Every one that I have seen that has embedded electronics uses epoxy potting. I think that is it to protect the circuitry rather than to foil any repair. I also noticed corrosion in the pictures. My F&P Icon+ did not have the cooldown cycle that Resmed has. If I turned off the machine and did not empty the humidifier, I would see heavy condensation in the tube and humidifier. You can't see it, but you would have the same condensation in the blower. This can't be good for it.