ResMed S9 Stopped working

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CjD1973
Posts: 1
Joined: Thu Feb 11, 2021 2:57 am

ResMed S9 Stopped working

Post by CjD1973 » Thu Feb 11, 2021 3:15 am

I have A ResMed S9 that i got in 2013 and it quit working tonight. I put it on tonight and didn't notice if it was working right then as I just rolled over and tried to sleep. i tossed and turned for a couple of hours and finally looked at it and the power button was blinking green and the humidifier was blinking blue at the same time and the screen is blank/off. I checked the plug at the machine, the power block and the outlet and all are firmly connected. I tried it without the humidifier connected, it was the same. i tried with the SD card seated and ejected. same thing. i noticed that the light on the power brick was blinking at the same rate as the lights on the machine and when i unplugged the cord from the machine the power brick light comes on steady and bright. plug it back in and it flashes with the machine. every once in awhile it would look like the screen and everything would start to work for a second or two, the screen would show the ResMed logo an start to load and the button light would be solid but it would stop quickly and blink again.

is my machine dead or is there a fix out there for it? anyone seen this before?

User avatar
ChicagoGranny
Posts: 15537
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: ResMed S9 Stopped working

Post by ChicagoGranny » Thu Feb 11, 2021 7:47 am

It's in its eighth year. Shouldn't you consider replacing it?

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed S9 Stopped working

Post by Dog Slobber » Thu Feb 11, 2021 2:18 pm

ChicagoGranny wrote: ↑
Thu Feb 11, 2021 7:47 am
It's in its eighth year. Shouldn't you consider replacing it?
I don't know CjD1973 (the OP) or anything about them, but why let that stop me from answering on their behalf.

Yes, they should consider replacing their machine.
Battery Backup: EcoFlow Delta 2

User avatar
ChicagoGranny
Posts: 15537
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: ResMed S9 Stopped working

Post by ChicagoGranny » Thu Feb 11, 2021 6:35 pm

Dog Slobber wrote: ↑
Thu Feb 11, 2021 2:18 pm
Yes, they should consider replacing their machine.
I suspected something like this but was too shy to mention it.

User avatar
zonker
Posts: 11498
Joined: Fri Jun 19, 2015 4:36 pm

Re: ResMed S9 Stopped working

Post by zonker » Thu Feb 11, 2021 7:58 pm

ChicagoGranny wrote: ↑
Thu Feb 11, 2021 6:35 pm
Dog Slobber wrote: ↑
Thu Feb 11, 2021 2:18 pm
Yes, they should consider replacing their machine.
I suspected something like this but was too shy to mention it.
YOU?!?!?

Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: ResMed S9 Stopped working

Post by Muse-Inc » Thu Feb 11, 2021 10:58 pm

Try plugging it into another outlet, then try a different electrical circuit.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

gosman
Posts: 55
Joined: Sun Oct 17, 2010 3:44 pm
Contact:

Re: ResMed S9 Stopped working

Post by gosman » Thu Feb 11, 2021 11:00 pm

I have a S9 and love mine and dread the day it dies. The Airsense 10 is a pitiful replacement. Noisy and inadequate water supply. But the doc says CPAP machines last about 7 years. Yours is overdue and so is mine. They don't make them like the use to!

_________________
MaskHumidifier

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed S9 Stopped working

Post by palerider » Fri Feb 12, 2021 12:10 am

gosman wrote: ↑
Thu Feb 11, 2021 11:00 pm
I have a S9 and love mine and dread the day it dies. The Airsense 10 is a pitiful replacement. Noisy and inadequate water supply.
The water tank on the Air10 series holds *EXACTLY* the same amount of water as the water tank on the S9 series. (380ml)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
elsueno
Posts: 100
Joined: Fri Oct 09, 2020 2:07 pm

Re: ResMed S9 Stopped working

Post by elsueno » Fri Feb 12, 2021 3:34 am

x

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: ResMed S9 VPAP Auto (I've had 4), I don't use the humidifier
Last edited by elsueno on Sat Feb 20, 2021 1:20 am, edited 1 time in total.

User avatar
HD4Mark
Posts: 41
Joined: Wed Oct 21, 2020 3:06 pm
Location: Marathon, FL or Cicero, NY

Re: ResMed S9 Stopped working

Post by HD4Mark » Fri Feb 12, 2021 7:01 am

Does the power supply blink while plugged into an outlet but not your machine? It would be nice if it was just the power supply.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Machine is a ResMed Aircurve 10 ASV

User avatar
cyberdreamer
Posts: 129
Joined: Mon Jun 24, 2019 10:57 am
Location: Brazil

Re: ResMed S9 Stopped working

Post by cyberdreamer » Fri Feb 12, 2021 9:29 am

gosman wrote: ↑
Thu Feb 11, 2021 11:00 pm
I have a S9 and love mine and dread the day it dies. The Airsense 10 is a pitiful replacement. Noisy and inadequate water supply. But the doc says CPAP machines last about 7 years. Yours is overdue and so is mine. They don't make them like the use to!
I'm surprised you think that way. I had a S9 loaner for a year before buying my Airsense 10 apap and they both work perfectly. The new Airsense 10 even has a new humidity level (8).
Machine: Airsense 10 Autoset (humidity 7, slimline hose) (5 years)
Mask: Resmed F30 (medium) (1 year)
Previous masks: Swift FX Nano (5 years); Dreamwear Nasal (few days); Swift FX pillow (few days)
Sleep study in 2016 with AHI 12 and some oxy desats

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: ResMed S9 Stopped working

Post by klm49 » Fri Feb 12, 2021 10:19 am

CjD1973 wrote: ↑
Thu Feb 11, 2021 3:15 am
I have A ResMed S9 that i got in 2013 and it quit working tonight. I put it on tonight and didn't notice if it was working right then as I just rolled over and tried to sleep. i tossed and turned for a couple of hours and finally looked at it and the power button was blinking green and the humidifier was blinking blue at the same time and the screen is blank/off. I checked the plug at the machine, the power block and the outlet and all are firmly connected. I tried it without the humidifier connected, it was the same. i tried with the SD card seated and ejected. same thing. i noticed that the light on the power brick was blinking at the same rate as the lights on the machine and when i unplugged the cord from the machine the power brick light comes on steady and bright. plug it back in and it flashes with the machine. every once in awhile it would look like the screen and everything would start to work for a second or two, the screen would show the ResMed logo an start to load and the button light would be solid but it would stop quickly and blink again.

is my machine dead or is there a fix out there for it? anyone seen this before?
If you have access to a different power brick you could try it or maybe have someone test yours, otherwise I think I would get a different machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
ChicagoGranny
Posts: 15537
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: ResMed S9 Stopped working

Post by ChicagoGranny » Fri Feb 12, 2021 12:05 pm

When the S9 was canceled, I bought one to stretch out the years I could enjoy the machine. Then, I bought an AirSense 10. It works just as well. Maybe a little better.

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: ResMed S9 Stopped working

Post by bwexler » Sat Feb 13, 2021 4:44 pm

LSAT is currently advertising an A10 for Her on this site.
You might take a look.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
zonker
Posts: 11498
Joined: Fri Jun 19, 2015 4:36 pm

Re: ResMed S9 Stopped working

Post by zonker » Sat Feb 13, 2021 5:50 pm

CjD1973 wrote: ↑
Thu Feb 11, 2021 3:15 am
I have A ResMed S9 that i got in 2013 and it quit working tonight.
welp, SOMETHING must've happened as you haven't been in for a couple of nights.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein