Fisher Paykel Icon Statistics not Updating

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
starchild
Posts: 2
Joined: Sun Jan 31, 2021 10:48 am

Fisher Paykel Icon Statistics not Updating

Post by starchild » Sun Jan 31, 2021 10:53 am

Hi all, brand new to the forum, first post.

I have a 10-year-old Fisher Paykel Icon that I stopped using for a while, but have started with again. The flow is fine, but I just noticed over the last few days that the statistics aren't changing. It doesn't seem to be registering any new data. Has anyone else had this problem. I don't know the right keywords to search without getting too many to look through.

Do I need to reset something? If so, how?

User avatar
zonker
Posts: 11424
Joined: Fri Jun 19, 2015 4:36 pm

Re: Fisher Paykel Icon Statistics not Updating

Post by zonker » Sun Jan 31, 2021 12:06 pm

starchild wrote:
Sun Jan 31, 2021 10:53 am
Hi all, brand new to the forum, first post.

I have a 10-year-old Fisher Paykel Icon that I stopped using for a while, but have started with again. The flow is fine, but I just noticed over the last few days that the statistics aren't changing. It doesn't seem to be registering any new data. Has anyone else had this problem. I don't know the right keywords to search without getting too many to look through.

Do I need to reset something? If so, how?
welcome to the zoo!

not sure if this will help, but i found some videos while doing a google search. didn't look at them myself and i don't own the machine you're talking about.

hope it helps!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Fisher Paykel Icon Statistics not Updating

Post by Pugsy » Sun Jan 31, 2021 12:51 pm

Have no idea what the problem might be but I suggest starting by getting the provider/clinical manual for your model machine and see if anything in it gives you a clue.

You can request that manual via email from here

https://www.apneaboard.com/adjust-cpap- ... tup-manual

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

starchild
Posts: 2
Joined: Sun Jan 31, 2021 10:48 am

Re: Fisher Paykel Icon Statistics not Updating

Post by starchild » Sun Jan 31, 2021 11:24 pm

OK thanks folks I'll try both suggestions.

User avatar
zonker
Posts: 11424
Joined: Fri Jun 19, 2015 4:36 pm

Re: Fisher Paykel Icon Statistics not Updating

Post by zonker » Sun Jan 31, 2021 11:54 pm

starchild wrote:
Sun Jan 31, 2021 11:24 pm
OK thanks folks I'll try both suggestions.

Image

all part of the zonker service; not necessarily correct, but quick!
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

RNeil
Posts: 76
Joined: Tue Jul 31, 2018 7:58 am
Contact:

Re: Fisher Paykel Icon Statistics not Updating

Post by RNeil » Mon Feb 01, 2021 2:35 pm

I used to use an Icon+. I am annoyed every time I think about how they try to hide data from the user. I trusted Sleepyhead, but the F&P summaries seemed to show lower AHIs. I am happier with the Resmed S10 Autoset that I bought used.