A CRY FOR HELP, Is this only sleep apnea
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
Good job on the chart, Barebones. I'm not sure whether you are using the ramp feature; if you are, do you think you'd miss it if you turned it off? If you're willing to experiment, give it a try.
What jumps out to me is that you are having a lot of obstructive apneas, with a total AHI on CPAP almost at a diagnostic level. To reduce the OAs, you would probably benefit from a higher minimum pressure. Although your machine is quite nimble, it isn't able to head off OAs with a minimum this low.
You may wind up needing a higher minimum, but I'd suggest you start with 8, which I bet you'll find comfortable. If that's OK, leave it there for a few nights and then try inching it up if you're still seeing a lot of OAs.
Other things to think about:
* Do you mostly sleep on your back? sides? stomach? If you sleep on your back, I wonder whether you can try gently to train yourself to sleep on your sides. For many people, OAs are more frequent with back-sleeping, when the tongue can start sliding back and narrowing the airway.
* Also, do you use a high pillow or a stack of pillows? If so, you might try a flatter, firmer pillow so you'll be less likely to tuck your chin toward your chest. That, too, can narrow the airway, making OAs more prevalent.
What jumps out to me is that you are having a lot of obstructive apneas, with a total AHI on CPAP almost at a diagnostic level. To reduce the OAs, you would probably benefit from a higher minimum pressure. Although your machine is quite nimble, it isn't able to head off OAs with a minimum this low.
You may wind up needing a higher minimum, but I'd suggest you start with 8, which I bet you'll find comfortable. If that's OK, leave it there for a few nights and then try inching it up if you're still seeing a lot of OAs.
Other things to think about:
* Do you mostly sleep on your back? sides? stomach? If you sleep on your back, I wonder whether you can try gently to train yourself to sleep on your sides. For many people, OAs are more frequent with back-sleeping, when the tongue can start sliding back and narrowing the airway.
* Also, do you use a high pillow or a stack of pillows? If so, you might try a flatter, firmer pillow so you'll be less likely to tuck your chin toward your chest. That, too, can narrow the airway, making OAs more prevalent.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: A CRY FOR HELP, Is this only sleep apnea
much more better!Barebones wrote: ↑Sat Feb 06, 2021 2:52 pmwoops sorry. i thought i was meant to remove that section so i left it out. Thanks for letting me know![]()
hows this....
Also do you need multiple days for a good analysis?
still trying to understand this all. but thanks so much for all your support. i check this thread daily for replys. so thanks
yes, please, more charts so we can see what sort of trends are going on.
i have nothing to add to the most excellent advice that miss e has provided.
always ALWAYS ask if you don't understand something. some of "old timers" have a tendency to speak in jargon and assume the new user is following along.
continued good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: A CRY FOR HELP, Is this only sleep apnea
Im keen to try anything at this stage. My life sucks! So I will check my settings tonight and try that. I mostly sleep on my side or my back. Always been a side sleeper tho. And don't have too many pillows as I get a sore neck. So mostly one flat pillow.Miss Emerita wrote: ↑Sat Feb 06, 2021 5:29 pmGood job on the chart, Barebones. I'm not sure whether you are using the ramp feature; if you are, do you think you'd miss it if you turned it off? If you're willing to experiment, give it a try.
What jumps out to me is that you are having a lot of obstructive apneas, with a total AHI on CPAP almost at a diagnostic level. To reduce the OAs, you would probably benefit from a higher minimum pressure. Although your machine is quite nimble, it isn't able to head off OAs with a minimum this low.
You may wind up needing a higher minimum, but I'd suggest you start with 8, which I bet you'll find comfortable. If that's OK, leave it there for a few nights and then try inching it up if you're still seeing a lot of OAs.
Other things to think about:
* Do you mostly sleep on your back? sides? stomach? If you sleep on your back, I wonder whether you can try gently to train yourself to sleep on your sides. For many people, OAs are more frequent with back-sleeping, when the tongue can start sliding back and narrowing the airway.
* Also, do you use a high pillow or a stack of pillows? If so, you might try a flatter, firmer pillow so you'll be less likely to tuck your chin toward your chest. That, too, can narrow the airway, making OAs more prevalent.
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
Keep us posted, would you?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: A CRY FOR HELP, Is this only sleep apnea
yes i will!
last night i changed my settings as you suggested.
i did notice a difference and struggled a little to get comfortable. i felt the air in my lungs which is a strange feeling but think i can get used to it. Ive been thinking ..because i breath through my nose and not my mouth and i have a deviated septum. i wonder how effective cpap can be? i mean its only a small hole that has to be forced into to stop and event. ??? just something i have been thinking about....
..setting from last night and graph ...
min pressure 8 was 4
max 17
response standard
ramp time 25min- was 35min
start pressure 4
- Attachments
-
- oscar 3.png (109.85 KiB) Viewed 1770 times
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
These are encouraging results, Barebones, though as always, one night does not a trend make. Your obstructive apnea is better controlled, which tells us you're on a good path here. There's a little uptick in central apneas, but I suspect the cluster right before your 11:15 break was during a transition from sleep to waking and so of no real consequence. Your pressure levels are smoother, which perhaps makes for a more restful night.
I think if you're willing, you'd do well to continue with these settings for at least a few more nights so you can get used to the slightly higher median pressure. If all goes well, then next step might be to raise the minimum a little bit at a time, and perhaps reduce the ramp time gradually as well.
Keep us posted!
I think if you're willing, you'd do well to continue with these settings for at least a few more nights so you can get used to the slightly higher median pressure. If all goes well, then next step might be to raise the minimum a little bit at a time, and perhaps reduce the ramp time gradually as well.
Keep us posted!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
p.s. Your machine produces continuous pressure -- sort of pressurizing your whole airway all the time. At the extreme, if the airway is quite constricted, your breathing will be constricted too. But it will be less constricted with pressure continuously at work than without it.
I hope others will weigh in on this, as there are aspects of it I feel I don't understand as well as I'd like to.
I hope others will weigh in on this, as there are aspects of it I feel I don't understand as well as I'd like to.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: A CRY FOR HELP, Is this only sleep apnea
x
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: ResMed S9 VPAP Auto (I've had 4), I don't use the humidifier |
Last edited by elsueno on Sat Feb 20, 2021 1:21 am, edited 1 time in total.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: A CRY FOR HELP, Is this only sleep apnea
I would definitely see about going to a neurologist. I have been having balance issues for several months. I went the ENT. He couldn't find anything. So he sent me for vestibular testing. Results were borderline abnormal, but ruled out the basic stuff. She he ordered s brain MRI. Small white spots that are basically lesions, but too small to be called lesions. Also having a variety of other symptoms. I am currently waiting for multiple test results and follow up. He is testing my MS, Lyme, and a few other things. It sounds like you have symptoms that could really use some testing.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: A CRY FOR HELP, Is this only sleep apnea
THANKS EVERYONE. You are being so helpful
I havnt changed any settings since last time we spoke but will lower the ramp a little more. i slept quite well last night.
The next set of test results im waiting to hear back are for an Allergy test and Thyroid. I and my family have discussed a MRI Brain scan
but at the time the doctor was not for it. I will rediscuss with him soon i feel. anyway here is last nights data. Thanks
I havnt changed any settings since last time we spoke but will lower the ramp a little more. i slept quite well last night.
The next set of test results im waiting to hear back are for an Allergy test and Thyroid. I and my family have discussed a MRI Brain scan
but at the time the doctor was not for it. I will rediscuss with him soon i feel. anyway here is last nights data. Thanks
- Attachments
-
- oscar 4.png (111.47 KiB) Viewed 1710 times
- cyberdreamer
- Posts: 129
- Joined: Mon Jun 24, 2019 10:57 am
- Location: Brazil
Re: A CRY FOR HELP, Is this only sleep apnea
Great leak rate. I'm jealous. Sleep quality should improve as you get used to the machine.Barebones wrote: ↑Thu Feb 11, 2021 2:11 pmTHANKS EVERYONE. You are being so helpful
I havnt changed any settings since last time we spoke but will lower the ramp a little more. i slept quite well last night.
The next set of test results im waiting to hear back are for an Allergy test and Thyroid. I and my family have discussed a MRI Brain scan
but at the time the doctor was not for it. I will rediscuss with him soon i feel. anyway here is last nights data. Thanks
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
Machine: Airsense 10 Autoset (humidity 7, slimline hose) (5 years)
Mask: Resmed F30 (medium) (1 year)
Previous masks: Swift FX Nano (5 years); Dreamwear Nasal (few days); Swift FX pillow (few days)
Sleep study in 2016 with AHI 12 and some oxy desats
Mask: Resmed F30 (medium) (1 year)
Previous masks: Swift FX Nano (5 years); Dreamwear Nasal (few days); Swift FX pillow (few days)
Sleep study in 2016 with AHI 12 and some oxy desats
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
You had a good sleep—that’s excellent. Stick with your current settings at least a few more days, then try going up a little. But do go ahead and shorten the ramp if you can hack that.
I hope the additional medical investigations will pay off. Meanwhile, do keep us posted.
I hope the additional medical investigations will pay off. Meanwhile, do keep us posted.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: A CRY FOR HELP, Is this only sleep apnea
hi guys i posted a reply yesterday but doesnt look like it has showed up . ah well .. i will post up last night and the days before data.
still struggling with fatigue and brain fog but feel like the new machine settings are better...
still struggling with fatigue and brain fog but feel like the new machine settings are better...
- Attachments
-
- oscarshot.png (110.39 KiB) Viewed 1636 times
Re: A CRY FOR HELP, Is this only sleep apnea
more data...... let me know if you notice anything you think i can change to help better my therapy.
thanks
thanks
- Attachments
-
- oscarrecent.png (109.02 KiB) Viewed 1636 times
- Miss Emerita
- Posts: 3760
- Joined: Sun Nov 04, 2018 8:07 pm
Re: A CRY FOR HELP, Is this only sleep apnea
How would you feel about raising your minimum to 8.8? I'm going to guess that little flurry of CAs at the end of the night were transitional and so not much of a red flag. The higher minimum might continue to help you have fewer obstructive events.
I'd also suggest raising your minimum during ramp to something higher that you'd still be comfortable with -- your call what that might be. With an EPR of 3, a minimum of 7 is very comfortable for most people.
I'd also suggest raising your minimum during ramp to something higher that you'd still be comfortable with -- your call what that might be. With an EPR of 3, a minimum of 7 is very comfortable for most people.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/