Puzzled by Pressures

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 4:06 pm

My REMstar Auto (System One 60 Series) when set at a Fixed 8.5 (cmH2O), is showing "Pressure set: 11.5" in Oscar software. Anyone know why? BTW, this is my first post in a long time. I joined this board 15 years ago. I've searched the Internet for an answer-zilch. This board has some very savvy people, so once again, I turn to you.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 4:11 pm

screen shot please of detailed nightly report.
Make sure calendar and pie chart is hidden.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 4:23 pm

https://lh3.googleusercontent.com/DUxQ6 ... authuser=0

This is the screenshot with issues circled in red.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 4:26 pm

You didn't hide the calendar. I need to see the machine settings below the statistics.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 4:35 pm


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 4:40 pm

I can't see your most recent image as it requires a log in password
but I think your problem is the cpap check mode you are using.
http://incenter.medical.philips.com/doc ... %3d9752807

It may be set at 8.5 but using that mode will cause the machine to increase the pressure if it thinks you need it.
It's giving you 11.5 so it must think you need it.

I would get out of that mode and use just regular cpap mode.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 5:06 pm

Please see if this link gives you access: https://photos.app.goo.gl/EB7c8WnActDaJbHJ8

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 5:20 pm

I can see the image now and it confirms what I suspected.

I think that while the "official" setting is probably 8.5 if you go into the machine that is what you will likely see....the cpap check mode way of doing things has essentially decided it wants you to use something else.
It's obvious from the pressure graph that you are seeing 11.5 fixed with the slightly lower pressure on exhale being a result of the Flex setting.

So that explains the weird picture and reporting....now if you are sleeping great and feeling great it probably doesn't hurt anything to use it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 5:43 pm

Thanks for looking at this. I'd like to know more about "CPAP flex". The link "http://incenter.medical.philips.com/doc ... %3D9752807" is not working for me, however.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 5:45 pm

When I click on the link, I see "Uh oh! Seems like your file is not a PDF or is corrupted. Please check that your file is a valid PDF and try again."

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 5:46 pm

I should have said, "CPAP check mode".

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 5:55 pm

I found this on another board, "The patient's user's manual says this about C-Check:

"CPAP Check (C-Check) - If available on your device, this mode delivers CPAP therapy while automatically adjusting the pressure level to meet your needs. Pressure adjustments while in CPAP-Check mode are more gradual than those that occur in Auto-Trial mode and the amount of adjustment that can be made over time is limited."

That's wild. "CPAP Check" adjusts the pressure even though you set the machine to "Fixed". How the hell do you know what you will get? :lol:

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 6:00 pm

Not sure why the link doesn't work for you...it works for me just fine
http://incenter.medical.philips.com/doc ... %3D9752807

But I got to it with a google search for "system one cpap check mode" and I started reading.
https://www.google.com/search?client=fi ... Check+mode

But I took a screen shot.
Apparently it is an option for the auto adjusting capable machines...so instead of auto/apap mode where it will auto adjust all night long it does this fixed thing.

Bunch of hooey IMHO. I prefer regular auto mode instead of using a higher pressure all night just because the machine thinks I need a higher pressure all night.
cpapcheck.JPG
cpapcheck.JPG (134.52 KiB) Viewed 1192 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Puzzled by Pressures

Post by Pugsy » Thu Jan 28, 2021 6:02 pm

https://www.apneaboard.com/adjust-cpap- ... tup-manual

You might go to the above link and request the provider/clinical manual for your machine.
It's all explained in the manual.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: Puzzled by Pressures

Post by Jerry69 » Thu Jan 28, 2021 7:03 pm

Thanks very much, Pugsy. I watched the YouTube on CPAP Check Mode. I had already found the user manual for my machine and learned, "• CPAP-Check – If available on your device, this mode delivers CPAP therapy while automatically adjusting the pressure level to meet your needs. Pressure adjustments while in CPAP-Check mode are more gradual than those that occur in Auto-Trial mode and the amount of adjustment that can be made over time is limited.

Since the CPAP Check Mode has decided that my pressure should be 11.5 cm, I'm going to test by setting the pressure to 10 cm in the CPAP Test Mode and see what happens.

I've gained a lot of insight from our exchange. I thought I'd be able lo get answers on this forum.

THANK YOU

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm