On my dreamstation go, the different flex settings are giving me the following cm h2o epap relief noted in oscar. These are when the pressure is stable rather than climbing or falling since the difference varies then.
Cflex 1 - 0.2
Cflex 2 - 0.3
Cflex 3 - 0.4
Aflex 1 - 1.5
Aflex 2 - 1.6
Aflex 3 - 1.8
The user manual doesn't specify the exact amount of relief, just that the higher number gives more relief. Are this typical relief numbers others are setting with PR devices? Seems like I should seeing a greater difference than one tenth cm h20 between each setting...
PR Flex settings
PR Flex settings
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
- Attachments
-
- C FLEX 1
- Cflex1.png (209.25 KiB) Viewed 1228 times
-
- A FLEX 3
- Aflex3.png (208.86 KiB) Viewed 1228 times
-
- A FLEX 1
- Aflex1.png (208.22 KiB) Viewed 1228 times
Last edited by willy737 on Tue Jan 19, 2021 6:34 pm, edited 3 times in total.
Re: PR Flex settings
c-flex 3 attachment
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
- Attachments
-
- C FLEX 3
- Cflex3.png (243.73 KiB) Viewed 1227 times
Re: PR Flex settings
bump
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: PR Flex settings
There is no set amount of relief at each setting because the amount is based on your own flow rate or forcefulness of your breathing but the most you can get is 2 cm drop during exhale at a setting of 3 and you get that only if you breathe forcefully.


_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: PR Flex settings
Oh I see, interesting. Thank you for the info Pugsy!
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
Re: PR Flex settings
Do resmed machines work in a similar manner? Or do they provide the cm h2o corresponding to the epr level regardless of flow rate?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
Re: PR Flex settings
ResMed EPR or exhale relief is a strict per cm drop up to 3 cm setting. So setting of 1 is a 1 cm drop, 3 is a 3 cm drop and force of breathing or flow rate doesn't change it. You get what you select.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: PR Flex settings
Got it. Thanks again for the info