PR Flex settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

PR Flex settings

Post by willy737 » Tue Jan 19, 2021 6:28 pm

On my dreamstation go, the different flex settings are giving me the following cm h2o epap relief noted in oscar. These are when the pressure is stable rather than climbing or falling since the difference varies then.

Cflex 1 - 0.2
Cflex 2 - 0.3
Cflex 3 - 0.4

Aflex 1 - 1.5
Aflex 2 - 1.6
Aflex 3 - 1.8

The user manual doesn't specify the exact amount of relief, just that the higher number gives more relief. Are this typical relief numbers others are setting with PR devices? Seems like I should seeing a greater difference than one tenth cm h20 between each setting...
Attachments
Cflex1.png
C FLEX 1
Cflex1.png (209.25 KiB) Viewed 1230 times
Aflex3.png
A FLEX 3
Aflex3.png (208.86 KiB) Viewed 1230 times
Aflex1.png
A FLEX 1
Aflex1.png (208.22 KiB) Viewed 1230 times
Last edited by willy737 on Tue Jan 19, 2021 6:34 pm, edited 3 times in total.

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: PR Flex settings

Post by willy737 » Tue Jan 19, 2021 6:29 pm

c-flex 3 attachment
Attachments
Cflex3.png
C FLEX 3
Cflex3.png (243.73 KiB) Viewed 1229 times

User avatar
zonker
Posts: 11496
Joined: Fri Jun 19, 2015 4:36 pm

Re: PR Flex settings

Post by zonker » Fri Jan 22, 2021 8:11 pm

bump
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65414
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR Flex settings

Post by Pugsy » Fri Jan 22, 2021 8:46 pm

There is no set amount of relief at each setting because the amount is based on your own flow rate or forcefulness of your breathing but the most you can get is 2 cm drop during exhale at a setting of 3 and you get that only if you breathe forcefully.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: PR Flex settings

Post by willy737 » Sat Jan 23, 2021 3:19 pm

Oh I see, interesting. Thank you for the info Pugsy!

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: PR Flex settings

Post by willy737 » Sun Jan 24, 2021 7:16 pm

Pugsy wrote: ↑
Fri Jan 22, 2021 8:46 pm
There is no set amount of relief at each setting because the amount is based on your own flow rate or forcefulness of your breathing but the most you can get is 2 cm drop during exhale at a setting of 3 and you get that only if you breathe forcefully.

Image
Do resmed machines work in a similar manner? Or do they provide the cm h2o corresponding to the epr level regardless of flow rate?

User avatar
Pugsy
Posts: 65414
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR Flex settings

Post by Pugsy » Sun Jan 24, 2021 7:25 pm

willy737 wrote: ↑
Sun Jan 24, 2021 7:16 pm
Do resmed machines work in a similar manner? Or do they provide the cm h2o corresponding to the epr level regardless of flow rate?
ResMed EPR or exhale relief is a strict per cm drop up to 3 cm setting. So setting of 1 is a 1 cm drop, 3 is a 3 cm drop and force of breathing or flow rate doesn't change it. You get what you select.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
willy737
Posts: 30
Joined: Thu Nov 19, 2020 8:47 pm
Location: DENVER

Re: PR Flex settings

Post by willy737 » Sun Jan 24, 2021 9:38 pm

Got it. Thanks again for the info