OSCAR Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jeffdanger
Posts: 14
Joined: Sun Jul 20, 2008 5:46 pm

OSCAR Question

Post by jeffdanger » Mon Dec 28, 2020 5:54 am

Last night I had two sleep sessions interrupted by a two minute lapse where I turned off the machine and turned back on. All the files are on the sd card but OSCAR is only reporting one session--the first one that was one hour 40 minutes. The second session that was 5 hours 50 minutes is not being captured. I've never seen this happen before as most of my sessions are two or three nightly and it always combines all sessions.

Any ideas?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: OSCAR Question

Post by Julie » Mon Dec 28, 2020 6:24 am

Post at least the first one (in your previous thread with repeat of above message), maybe there are hints there..

jeffdanger
Posts: 14
Joined: Sun Jul 20, 2008 5:46 pm

Re: OSCAR Question

Post by jeffdanger » Mon Dec 28, 2020 6:57 am

Last night I had two sleep sessions interrupted by a two minute lapse where I turned off the machine and turned back on. All the files are on the sd card but OSCAR is only reporting one session--the first one that was one hour 40 minutes. The second session that was 5 hours 50 minutes is not being captured. I've never seen this happen before as most of my sessions are two or three nightly and it always combines all sessions.

Any ideas? IHere is the session data and the files on the sd card. I also purged the day and reimported but same thing.
Attachments
Screenshot 2020-12-28 075345.jpg
Screenshot 2020-12-28 075345.jpg (62.84 KiB) Viewed 995 times
screenshot-20201228-074149.png
screenshot-20201228-074149.png (97.08 KiB) Viewed 995 times

User avatar
Dog Slobber
Posts: 4340
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: OSCAR Question

Post by Dog Slobber » Mon Dec 28, 2020 10:37 am

In OSCAR, delete the days data and then re-import.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65437
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: OSCAR Question

Post by Pugsy » Mon Dec 28, 2020 11:38 am

There is a purge feature in OSCAR and you can select the date to purge.
It's under Data and then the Advanced choice in the menu.
It just removes it from OSCAR...then do another import from your SD card and see what happens.
I may have to RISE but I refuse to SHINE.

User avatar
BlueDragon
Posts: 575
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR Question

Post by BlueDragon » Mon Dec 28, 2020 3:05 pm

jeffdanger wrote: ↑
Mon Dec 28, 2020 5:54 am
Last night I had two sleep sessions interrupted by a two minute lapse where I turned off the machine and turned back on. All the files are on the sd card but OSCAR is only reporting one session--the first one that was one hour 40 minutes. The second session that was 5 hours 50 minutes is not being captured. I've never seen this happen before as most of my sessions are two or three nightly and it always combines all sessions.

Any ideas?
Is your machine set to DST and your computer to standard time, and did you load your SD card into OSCAR less than one hour after you stopped using the machine? In that situation, the last session is not imported. The workaround is to wait an hour and import again, and/or reset your CPAP time to the same as your computer. This problem has been fixed for the next release of OSCAR.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

jeffdanger
Posts: 14
Joined: Sun Jul 20, 2008 5:46 pm

Re: OSCAR Question

Post by jeffdanger » Mon Dec 28, 2020 3:38 pm

Thank you, Blue Dragon, you have solved the mystery! When DST went away recently, I didn't bother to change the time on the CPAP. I see now that I couldn't expect the program to import something that technically hasn't happened yet.

This is a great group and I thank you all for your participation.

User avatar
zonker
Posts: 11507
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR Question

Post by zonker » Mon Dec 28, 2020 5:53 pm

jeffdanger wrote: ↑
Mon Dec 28, 2020 3:38 pm
I see now that I couldn't expect the program to import something that technically hasn't happened yet.
HA! that's funny. maybe the oscar team can incorporate time travel into the next release?
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4340
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: OSCAR Question

Post by Dog Slobber » Mon Dec 28, 2020 9:19 pm

zonker wrote: ↑
Mon Dec 28, 2020 5:53 pm
jeffdanger wrote: ↑
Mon Dec 28, 2020 3:38 pm
I see now that I couldn't expect the program to import something that technically hasn't happened yet.
maybe the oscar team can incorporate time travel into the next release?

Dr_Zonker.jpg
Dr_Zonker.jpg (218.04 KiB) Viewed 924 times
Battery Backup: EcoFlow Delta 2