Problems With CPAP Tubing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: Problems With CPAP Tubing

Post by zonker » Mon Dec 28, 2020 11:01 pm

Movieman wrote:
Mon Dec 28, 2020 10:45 pm

Thank you palerider. This was the issue. Everything is working great now. My current hose is slimline non-heated and the machine was set to slimline. The hose I purchased said it was standard size. When I went into settings and picked “standard” on the menu, the hose worked with no problems.

Thanks everyone!! Great forum.
thank YOU for letting us know. even though the solution had nothing to do with me, it's still great to hear good results!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65209
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Problems With CPAP Tubing

Post by Pugsy » Mon Dec 28, 2020 11:05 pm

I have been on cpap over 11 years now.
I am trying to remember when the last time I actually cleaned the hose after the initial 6 months when I first started therapy and was religious about doing a once a week cleaning of everything. That got old real fast. I abandoned that schedule.
I might have cleaned one 5 or 6 years ago but that was because I was checking for holes in it as I found my cat playing with it. :lol:
I had to use soapy water to check for leaks from kitty claw holes.

Your choice though. All depends on just how much it bugs you or not. Obviously it doesn't bug me very much but I cast no stones at people who want to adopt any sort of different cleaning (or not cleaning) schedule than mine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems With CPAP Tubing

Post by palerider » Tue Dec 29, 2020 12:06 pm

Movieman wrote:
Mon Dec 28, 2020 10:45 pm
palerider wrote:
Mon Dec 28, 2020 3:38 pm
Movieman wrote:
Mon Dec 28, 2020 12:55 pm
🙂🙂🙂 I still wonder why it only happens w/ new hose though? Lots of crazy quirks happen using CPAP 😆😆. Thanks!!
Is the new hose the same diameter (the hose itself, not the connectors at the end) as the old one? is the old one a heated hose? Did you adjust the host size in the machine's menu?
Thank you palerider. This was the issue. Everything is working great now. My current hose is slimline non-heated and the machine was set to slimline. The hose I purchased said it was standard size. When I went into settings and picked “standard” on the menu, the hose worked with no problems.

Thanks everyone!! Great forum.
Great! The 'slimline' setting causes the machine to operate at a higher pressure to compensate for pressure losses due to the greater resistance of the smaller hose. Not setting the hose properly would have caused you to have higher pressure at the mask, and strange other behavior :)

Glad it's fixed!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Problems With CPAP Tubing

Post by chunkyfrog » Tue Dec 29, 2020 1:10 pm

"Over cautious", that's it!
Too many lawyers, too many skeezebags, litgation city.
Everybody needs to cover the hinie more than ever.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Movieman
Posts: 82
Joined: Mon Nov 30, 2020 11:38 am

Re: Problems With CPAP Tubing

Post by Movieman » Tue Dec 29, 2020 4:48 pm

Pugsy wrote:
Mon Dec 28, 2020 11:05 pm
I have been on cpap over 11 years now.
I am trying to remember when the last time I actually cleaned the hose after the initial 6 months when I first started therapy and was religious about doing a once a week cleaning of everything. That got old real fast. I abandoned that schedule.
I might have cleaned one 5 or 6 years ago but that was because I was checking for holes in it as I found my cat playing with it. :lol:
I had to use soapy water to check for leaks from kitty claw holes.

Your choice though. All depends on just how much it bugs you or not. Obviously it doesn't bug me very much but I cast no stones at people who want to adopt any sort of different cleaning (or not cleaning) schedule than mine.
Ha! Lol I can relate to your cat playing with the hose. I have two cats, but I’ve been lucky so far. They seem to ignore the hose. 😆 For now!!

This is good to read. Others on the forum and in this thread also found that cleaning the hose isn’t necessary. It’s good to see that it won’t make me sick if I don’t clean it. After a little over a month on CPAP, I hate cleaning the hose. :D