Do I need to change pressure for a different mask?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
KillMyApnea
Posts: 8
Joined: Sat Dec 19, 2020 1:11 pm

Do I need to change pressure for a different mask?

Post by KillMyApnea » Wed Dec 23, 2020 10:20 am

Hi everyone :)

I use a Philips AutoCPAP with a dreamwear nasal gel pillow mask. The titration in the lab was done using the same nasal pillows. The "Mask Type" setting on the machine is set to "2". I borrowed the Resmed Full Face F20 AirTouch mask and want to try it for a few days. Do I need to change the pressure settings for this new mask? Can just swap masks without effect to my therapy?

Thanks

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Do I need to change pressure for a different mask?

Post by chunkyfrog » Wed Dec 23, 2020 10:36 am

No need to change anything.
The respironics mask setting relates to the inside volume of the mask, and is only used for reporting, not therapy.
And the difference is very slight--not enough to matter.
It only refers to numbers on Respironics brand masks--mainly to fool you into thinking you need THEIR mask.
But we are all smarter than that.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need to change pressure for a different mask?

Post by Pugsy » Wed Dec 23, 2020 11:15 am

Technically Respironics says that when we are using a non Respironics mask that we should select 0 or off for the mask type (which is actually a resistance setting) in the menu.
In real life it's not a critical setting. You can make the change and I am betting you won't see or feel any difference in anything.
I have done that experiment...tried all the settings including off with my mask and nothing changed anywhere even in how it felt to me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

KillMyApnea
Posts: 8
Joined: Sat Dec 19, 2020 1:11 pm

Re: Do I need to change pressure for a different mask?

Post by KillMyApnea » Wed Dec 23, 2020 11:45 am

Thank you guys so much! These are great and very quick replies!

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Do I need to change pressure for a different mask?

Post by JLROhio » Wed Dec 23, 2020 2:02 pm

I just recently tried out a Respironics Bi-Pap ASV.

The comments below are based upon “MY Experience with a Philips device.

I had to change the mask setting to two (2) in order for it to work correctly (for me) with my ResMed P10 nasal pillows.

Philips does not provide any information about which mask setting to use for non-Philips masks and pillows. I had to research on my own and came across a couple of websites and YouTube videos that said to use number two (2) for my MASK SETTING to match my Resmed P10 nasal pillows.

You should go ahead and experiment and change the MASK settings to what best feels good to you… It does not take long… And you could accomplish all of this in under 5-10 minutes.

I am not at my computer right now (sending from phone) but Pugsy can send you a link to the clinician operator guide for your machine… You can order it via email… And that guide will explain the settings for the different mask settings for Philips mask. You can use that as a starting point.


EDIT: My post here refers to MASK Settings...found in the Comfort Setting of my Phlips BiPap ASV.
Last edited by JLROhio on Wed Dec 23, 2020 3:21 pm, edited 1 time in total.
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
ChicagoGranny
Posts: 15447
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Do I need to change pressure for a different mask?

Post by ChicagoGranny » Wed Dec 23, 2020 3:02 pm

KillMyApnea wrote:
Wed Dec 23, 2020 10:20 am
Do I need to change the pressure settings for this new mask?
What ARE your pressure settings?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Do I need to change pressure for a different mask?

Post by chunkyfrog » Wed Dec 23, 2020 3:54 pm

ChicagoGranny wrote:
Wed Dec 23, 2020 3:02 pm
KillMyApnea wrote:
Wed Dec 23, 2020 10:20 am
Do I need to change the pressure settings for this new mask?
What ARE your pressure settings?
KMA said it was set to 2.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need to change pressure for a different mask?

Post by Pugsy » Wed Dec 23, 2020 4:34 pm

chunkyfrog wrote:
Wed Dec 23, 2020 3:54 pm
ChicagoGranny wrote:
Wed Dec 23, 2020 3:02 pm
KillMyApnea wrote:
Wed Dec 23, 2020 10:20 am
Do I need to change the pressure settings for this new mask?
What ARE your pressure settings?
KMA said it was set to 2.
The setting of 2 mentioned by OP is just the mask type setting or what we used to see on the System One machines called Resistance setting. It's called mask type on the DreamStation and Resistance control on the System One machines. It's the same thing and for all non Respironics masks it is suggested that the menu selection just be set to 0 which effectively turns off this feature.
Now if someone wants to try to figure out the inherent resistance in a ResMed mask and then compare it to the inherent resistance to a Respironics mask and use the setting associated with that mask...they can sure do it but you have to be careful because with the Respironics setting options there is also a factor of the size of whatever is being used. Like the setting for a size small in something is different than a setting for a size large in the same mask.

It's an awful lot of work and not all that critical of work to be done IMHO. Respironics doesn't give resistance settings for non Respironics masks....duh...they of course want you to buy their masks. It's not that big of a deal and in real life the resistance isn't all that different and really doesn't affect much. Respironics makes out like it does because they want to make a big deal out of something they do.
I suspect ResMed does something similar with their generic mask setting but they don't ever really tell us for sure.

But to answer the OP's original question....it's not a critical setting change that needs to be made. Pressures sure don't need to be changed. Changing the actual Resistance setting doesn't really change the pressure used enough to even be measurable.
Technically though....Respironics says turn off Resistance control or mask type when not using a Respironics mask.
They have never ever talked about changing the pressure itself and that's certainly not what the resistance control does.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Do I need to change pressure for a different mask?

Post by chunkyfrog » Wed Dec 23, 2020 5:06 pm

Oops, my bad.
(Hides under a leaf)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
ChicagoGranny
Posts: 15447
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Do I need to change pressure for a different mask?

Post by ChicagoGranny » Wed Dec 23, 2020 5:17 pm

chunkyfrog wrote:
Wed Dec 23, 2020 5:06 pm
Oops, my bad.
(Hides under a leaf)
I've been able to delete my goofs if I catch it before anyone else posts in the thread.

User avatar
zonker
Posts: 11371
Joined: Fri Jun 19, 2015 4:36 pm

Re: Do I need to change pressure for a different mask?

Post by zonker » Wed Dec 23, 2020 6:36 pm

ChicagoGranny wrote:
Wed Dec 23, 2020 5:17 pm
chunkyfrog wrote:
Wed Dec 23, 2020 5:06 pm
Oops, my bad.
(Hides under a leaf)
I've been able to delete my goofs if I catch it before anyone else posts in the thread.
you are either very quick or seldom make goofs.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15447
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Do I need to change pressure for a different mask?

Post by ChicagoGranny » Thu Dec 24, 2020 9:20 am

zonker wrote:
Wed Dec 23, 2020 6:36 pm
you are either very quick
~.JPG
~.JPG (20.06 KiB) Viewed 2997 times

User avatar
zonker
Posts: 11371
Joined: Fri Jun 19, 2015 4:36 pm

Re: Do I need to change pressure for a different mask?

Post by zonker » Thu Dec 24, 2020 12:10 pm

ChicagoGranny wrote:
Thu Dec 24, 2020 9:20 am
zonker wrote:
Wed Dec 23, 2020 6:36 pm
you are either very quick
~.JPG
is she from pasadena?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg