2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
age: 37
height: 6'1
weight:170
health: fine otherwise
Time on cpap: 2 months
Type: Philips Respironics Dreamstation APAP
Mask: Dreamwear Full Facemask
Compliance: good
PROBLEM: my exhaustion is profound. It hits around 4pm at which time I become non-functional. It is impossible to get work done, be a functioning family member etc. Until then I seem to be reasonably alright. When I started the APAP the time from wake to 4pm was more full of energy. Now I'm getting tired throughout the day. AHI has on average risen since the first two weeks.
Please let me know what OSCAR data you need in addition to the data I have provided.
Included:
1. Summary Data
2. Example of two different days where pressures were very different but similar results.
height: 6'1
weight:170
health: fine otherwise
Time on cpap: 2 months
Type: Philips Respironics Dreamstation APAP
Mask: Dreamwear Full Facemask
Compliance: good
PROBLEM: my exhaustion is profound. It hits around 4pm at which time I become non-functional. It is impossible to get work done, be a functioning family member etc. Until then I seem to be reasonably alright. When I started the APAP the time from wake to 4pm was more full of energy. Now I'm getting tired throughout the day. AHI has on average risen since the first two weeks.
Please let me know what OSCAR data you need in addition to the data I have provided.
Included:
1. Summary Data
2. Example of two different days where pressures were very different but similar results.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
- Attachments
-
- nov2PressureAhi.png (142.77 KiB) Viewed 1576 times
-
- exampleDay.png (147.12 KiB) Viewed 1576 times
-
- statisticsCpap.png (111.84 KiB) Viewed 1576 times
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Try increasing your minimum pressure to 7 cm and see what happens.
You have a lot of "clutter" showing up on your reports that could be impacting overall sleep quality.
FLs and RERAs along with the OAs and hyponeas all point to the airway either collapsing or trying to collapse and could be causing sleep disturbances impacting the restorative powers of sleep.
You have a lot of "clutter" showing up on your reports that could be impacting overall sleep quality.
FLs and RERAs along with the OAs and hyponeas all point to the airway either collapsing or trying to collapse and could be causing sleep disturbances impacting the restorative powers of sleep.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
I've upped it to 5.5 from 4 which corresponded to a decrease in performance. So I'm hesitant to keep going since it should A-uto adjust.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
It's a well known fact that Respironics auto adjusting algorithm is a bit on the slow side and often people need to increase the minimum to get better results.
If you don't want to try an increase in the minimum pressure then I suggest you talk to your doctor about your issues.
If you don't want to try an increase in the minimum pressure then I suggest you talk to your doctor about your issues.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
I'll give it a shot then post the data. Thanks
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Note: I did experiment with nasal mask, which could explain some leak as opposed to around the mouth usage. Higher min pressure used. I'm mainly surprised at the higher ahi around non-leak periods.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
- Attachments
-
- summaryDataDec4Cpap.png (119.77 KiB) Viewed 1499 times
-
- data4Dec2020CPAP.png (154.97 KiB) Viewed 1499 times
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Could you reformat this most recent chart so that we can see these graphs:
Events
Flow rate
Pressure
Leaks
Snores.
To fit them all into a single screenshot, you can grab the horizontal gray bars that separate the graphs and push them up.
Thanks!
Events
Flow rate
Pressure
Leaks
Snores.
To fit them all into a single screenshot, you can grab the horizontal gray bars that separate the graphs and push them up.
Thanks!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Here ya go. Thanks for the help.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Here is results with new Nasal Pillows mask. I remember no leaks of any kind. Woke up with no leaks. Not sure why those leaks were showing early on. NOTE: I WAS AWAKE UNTIL 3.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Strangely enough, it looks like your maximum pressure is 7.5, even though on the left panel it says 12 or 15. Even with the conservative PR algorithm, your pressure should have gone up with your big batches of obstructive apneas. Could you double check what your maximum is?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
I just learned from Pugsy that the problem might be a PR feature called EZ start, which limits the maximum pressure for a period of time at the beginning of therapy. Could you check whether it's on? And if it is, turn it off and let's see where the pressure "wants" to go with a true maximum of 15.
Pugsy provided this image of the description of the EZ start function, if you want more information.
Pugsy provided this image of the description of the EZ start function, if you want more information.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
Anytime we see a Respironics pressure line not going up when it obviously should be going up or we would expect it to go up because of the flagged events I always suspect EZ-Start hindering things.
So when we see the pressure not going up and a truckload of apnea events happening....always look to see if EZ-Start somehow got turned on and if it did....turn it off. Dumbest idea Respironics ever came up with was EZ-Start.
So when we see the pressure not going up and a truckload of apnea events happening....always look to see if EZ-Start somehow got turned on and if it did....turn it off. Dumbest idea Respironics ever came up with was EZ-Start.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
diabellical
- Posts: 14
- Joined: Tue Dec 01, 2020 9:00 pm
Re: 2 months on APAP: hitting a wall at 4pm : barely functioning. OSCAR images inside
I’ll disable ez start and report back in a couple nights. Thanks.
_________________
| Mask: AirFit F20 Complete Mask + AirMini Mask Setup Pack Bundle |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: standard heated hose. Typical settings: Humidification 2-3, heated tube 3-4 |