AirSense™ 10 AutoSet turns itself off

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Sleepless in Ca
Posts: 18
Joined: Thu Aug 29, 2013 10:00 am

AirSense™ 10 AutoSet turns itself off

Post by Sleepless in Ca » Thu Nov 19, 2020 10:06 pm

My 3 year old AirSense™ 10 AutoSet now for the last 6 days turns itself off once to twice a night while I am sound asleep. I wake up feeling like I am suffocating and pull my mask off. I wait about 30 seconds, put the mask back on and it works fine until it does it again. Anyone know what is wrong? Thanks

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AirSense™ 10 AutoSet turns itself off

Post by LSAT » Fri Nov 20, 2020 5:26 am

Check to see if the "smart start" setting is turned on. If it is, turn it off.

User avatar
MarylandCPAPer2
Posts: 86
Joined: Tue Apr 14, 2020 9:39 pm

Re: AirSense™ 10 AutoSet turns itself off

Post by MarylandCPAPer2 » Sun Nov 22, 2020 10:09 am

Sleepless in Ca wrote:
Thu Nov 19, 2020 10:06 pm
My 3 year old AirSense™ 10 AutoSet now for the last 6 days turns itself off once to twice a night while I am sound asleep. I wake up feeling like I am suffocating and pull my mask off. I wait about 30 seconds, put the mask back on and it works fine until it does it again. Anyone know what is wrong? Thanks
The A10 should not be doing this. Is it possible you are having momentary power outages that cause the power to go on and off in your house? I have had 2 Airsense 10s over the last 6 years, both with the "smart start and stop" setting. I have only had the A10 go off during sleep during power outages, even ones that just flicked the power on and off. With the smart start and stop settings , it takes about 10 seconds after taking the mask off for the A10 to shut off.

I would have the machine checked out. This sounds like a defect.

Maryland CPAPer
Started CPAP 12/09. Currently use (since 2015) AirSense 10 AutoSet FOR HER.APAP Range: 11-20, AirFit P10 mask. Started home oxygen at 2L on 8/7/20.
Previously (2010 to 2015) APAP, Pressure 10-16. PR System One.. Was Maryland CPAPer 12/09 to 4/20.

grounded
Posts: 93
Joined: Sun May 17, 2020 1:19 pm

Re: AirSense™ 10 AutoSet turns itself off

Post by grounded » Mon Nov 23, 2020 8:44 pm

This has happened to me several times over the past 6 months but certainly not nightly. Horrible when it does happen tho

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AirSense™ 10 AutoSet turns itself off

Post by LSAT » Mon Nov 23, 2020 9:03 pm

It's always nice when a poster comes back after getting suggestions to indicate if anything works....too many never respond

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense™ 10 AutoSet turns itself off

Post by chunkyfrog » Tue Nov 24, 2020 12:03 am

Way too uncommon--but downright WONDERFUL when we get a little "attaboy".
I guess not this time.
(sniff)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AirSense™ 10 AutoSet turns itself off

Post by jimbud » Tue Nov 24, 2020 10:58 am

LSAT wrote:
Fri Nov 20, 2020 5:26 am
Check to see if the "smart start" setting is turned on. If it is, turn it off.
Sleepless in Ca wrote:
Thu Nov 19, 2020 10:06 pm
Thanks
JPB :D

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AirSense™ 10 AutoSet turns itself off

Post by LSAT » Tue Nov 24, 2020 11:35 am

The "thanks" was written in the original post asking for an answer...not in response to the suggestions.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AirSense™ 10 AutoSet turns itself off

Post by jimbud » Tue Nov 24, 2020 11:48 am

I know.
I just thought I would help the OP out.
I am sure your advice worked and OP is so happy the OP forgot to come back and do the right thing.
Probably still asleep.
Some people need a little boost.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

lkg
Posts: 2
Joined: Wed Nov 25, 2020 4:55 pm

Re: AirSense™ 10 AutoSet turns itself off

Post by lkg » Wed Nov 25, 2020 5:08 pm

My Airsense will shut off if too much air runs through it so check all your connections also e.g. Humidifier isn't snapped in all the way. It doesn't hurt to test the outlet. If you have an older house test with a multimeter. Trying another outlet or taking faceplate off to see if wires are loose if you don't have a multimeter. Outlets like all electronics do go bad(not often), GFI's go bad more often.