OSCAR high leak rate and grey bands

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
haughki
Posts: 7
Joined: Sat Nov 14, 2020 5:42 pm
Location: San Francisco, CA USA

OSCAR high leak rate and grey bands

Post by haughki » Sat Nov 14, 2020 7:30 pm

Hi,

Two questions:
- can anyone give me an idea about what would cause such high leak rates? It doesn't happen every night, and I can't think of what might be causing it when it does.
- What are the grey vertical bars on the Flow Rate and Leak Rate graphs?

Thanks!
oscar_leaks.jpg
oscar_leaks.jpg (476.19 KiB) Viewed 1248 times

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ResMed S9 AutoSet

User avatar
Dog Slobber
Posts: 4249
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: OSCAR high leak rate and grey bands

Post by Dog Slobber » Sat Nov 14, 2020 8:17 pm

High leak rates are caused by:
  • Poor mask fits
  • Mouth breathing on Nasal masks
  • Opening one's mouth and causing a poor seal on full face masks
On ResMeds the threshold for acceptable leaking is 24 L/M, anything above that contributes to reducing the machines ability to detect apneas (or the type of apnea), and therefore reduces its ability to address them properly. The grey bars are the times when the leak rate is above 24 Litres / Minute.
Battery Backup: EcoFlow Delta 2

User avatar
Miss Emerita
Posts: 3757
Joined: Sun Nov 04, 2018 8:07 pm

Re: OSCAR high leak rate and grey bands

Post by Miss Emerita » Sat Nov 14, 2020 8:44 pm

I realize you didn’t ask about this, but I’m curious. Why are you using a relatively low and fixed pressure? You’re having a fair number of obstructive apneas with that setting.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
haughki
Posts: 7
Joined: Sat Nov 14, 2020 5:42 pm
Location: San Francisco, CA USA

Re: OSCAR high leak rate and grey bands

Post by haughki » Tue Nov 17, 2020 11:56 am

@Dog Slobber -- tx, crystal clear.
Miss Emerita wrote:
Sat Nov 14, 2020 8:44 pm
I realize you didn’t ask about this, but I’m curious. Why are you using a relatively low and fixed pressure? You’re having a fair number of obstructive apneas with that setting.
@Miss Emerita -- tx for asking. Basically, I was doing something ridiculous :) I've been struggling with mask leaks, and thought I would try a low constant pressure just to see if I could get through a night without leaks (I guess: not a very well-formed plan :). Having read a bunch more about mask fitting, etc., I realize that it didn't really make much sense.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: ResMed S9 AutoSet

User avatar
Miss Emerita
Posts: 3757
Joined: Sun Nov 04, 2018 8:07 pm

Re: OSCAR high leak rate and grey bands

Post by Miss Emerita » Tue Nov 17, 2020 12:00 pm

Sounds like you're on it, haughki. Do give yourself a higher minimum, and while you're at it, give yourself a range. I would suggest 9 to 20 just to see how that goes, but you may have had previous settings that were working well on heading off the obstructive apneas.
Oscar software is available at https://www.sleepfiles.com/OSCAR/