altitude factor

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
larryn
Posts: 33
Joined: Thu Jan 16, 2020 9:43 am

altitude factor

Post by larryn » Thu Nov 12, 2020 12:52 pm

how does altitude change effect pressure settings? I will be going from 1000ft to7000ft.What changes might be needed?

_________________
MachineMask

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: altitude factor

Post by Wulfman... » Thu Nov 12, 2020 1:14 pm

larryn wrote:
Thu Nov 12, 2020 12:52 pm
how does altitude change effect pressure settings? I will be going from 1000ft to7000ft.What changes might be needed?
Read your User Manual. It's in there. Page 16 in the English part.

If you don't have one, here's a link:

https://www.resmed.com/us/dam/documents ... er_eng.pdf


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

RNeil
Posts: 76
Joined: Tue Jul 31, 2018 7:58 am
Contact:

Re: altitude factor

Post by RNeil » Fri Nov 13, 2020 9:30 am

The reference says it will work at altitudes 0-8500 feet. I think that means that is the altitude range that it was tested at. I think the OP was asking how or if he should change the settings.

As I understand the CPAP theory, the machine makes the pressure in your airway higher than the outside so that your airway doesn't collapse. The absolute pressure doesn't mater, just the difference between the pressure in the room and in your airway. The pressure that the machine measures and regulates is the difference between the atmospheric pressure and the outlet pressure. This pressure difference does not change with altitude. So I don't think you need to change your settings.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: altitude factor

Post by Wulfman... » Fri Nov 13, 2020 2:04 pm

RNeil wrote:
Fri Nov 13, 2020 9:30 am
The reference says it will work at altitudes 0-8500 feet. I think that means that is the altitude range that it was tested at. I think the OP was asking how or if he should change the settings.

As I understand the CPAP theory, the machine makes the pressure in your airway higher than the outside so that your airway doesn't collapse. The absolute pressure doesn't mater, just the difference between the pressure in the room and in your airway. The pressure that the machine measures and regulates is the difference between the atmospheric pressure and the outlet pressure. This pressure difference does not change with altitude. So I don't think you need to change your settings.
For all intents and purposes, the machine automatically compensates for altitude.
In the "old days" and cheaper, more basic models, there were settings for different altitudes. However, it appears that the newer ones (even in the more basic models) now have automatic altitude compensation.

The OP should be just fine.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

McSleepy
Posts: 499
Joined: Fri Nov 26, 2010 10:32 pm
Location: USA

Re: altitude factor

Post by McSleepy » Sat Nov 14, 2020 5:16 am

One consideration in altitude change is the change in absolute humidity; possibly exacerbated by the fact that most high-altitude locations also have drier climate, so relative humidity will also be lower. Ultimately, be prepared to adjust (increase) your humidification levels.