ResMed Airsense 10 Shutdown While Sleeping

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
pudgypoet
Posts: 1
Joined: Wed Oct 28, 2020 9:34 pm

ResMed Airsense 10 Shutdown While Sleeping

Post by pudgypoet » Wed Oct 28, 2020 10:10 pm

I have on 3 occasions been awaken by the AS10 shutting down while I was in deep sleep. On 2 other occasions it has lowered the air pressure substantially. I've called the place of purchase and the lady worked diligently with me trying to find the problem. We never found a resolution. In 2 of the shutdown experiences the water tank became empty. It's scary waking up fighting for air in the middle of a deep sleep. I'm interested in finding out what the problem(s) are. Any help would be greatly appreciated. One other thing...the ramp is turned off but the machine ramps up to the minimum level when turned on. Thanks.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed Airsense 10 Shutdown While Sleeping

Post by LSAT » Thu Oct 29, 2020 3:14 am

I would check to see if the "Smart Start" feature is turned on. That feature automatically turns on the machine when you put on the mask and start breathing. It turns off the machine when you take off the mask. If you have a large leak during the night, it could think you took off the mask and turn off...Also...If you have a long apnea event. However, it should turn back on also. If that feature is turned on...turn it off and use the on/off button...See if that solves the problem. An empty water tank should not turn the machine off, but it could burn the humidifier tub. With the ramp turned off, the machine is doing exactly what it should by going directly to the minimum pressure setting. (What are you expecting it to do?).

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: ResMed Airsense 10 Shutdown While Sleeping

Post by Julie » Thu Oct 29, 2020 5:11 am

Not to make light of it, but my cat's stepped on the Off button more than once... :wink:

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: ResMed Airsense 10 Shutdown While Sleeping

Post by khauser » Thu Oct 29, 2020 11:47 am

pudgypoet wrote: ↑
Wed Oct 28, 2020 10:10 pm
I have on 3 occasions been awaken by the AS10 shutting down while I was in deep sleep. On 2 other occasions it has lowered the air pressure substantially. I've called the place of purchase and the lady worked diligently with me trying to find the problem. We never found a resolution. In 2 of the shutdown experiences the water tank became empty. It's scary waking up fighting for air in the middle of a deep sleep. I'm interested in finding out what the problem(s) are. Any help would be greatly appreciated. One other thing...the ramp is turned off but the machine ramps up to the minimum level when turned on. Thanks.
I'm going to recommend you get OSCAR and look at your charts. To me, the part about the humidifier going dry suggests you had a rather large leak, and if a leak is large enough the machine will 'think' you've taken the mask off and power down (unless you turn off that feature as LSAT suggests).

Airsense 10 machines (the only Resmed machine I have experience) always come up to pressure semi-gradually. A real Ramp feature takes minutes, not the few seconds that the machine spends turning on air.

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: ResMed Airsense 10 Shutdown While Sleeping

Post by jimbud » Thu Oct 29, 2020 12:18 pm

Approximately, a year ago my Resmed Airsense 10 Autoset on Smart Start shut down by itself two separate times Both times the Sleepyhead flow rate showed very shallow breathing. (Tidal volume and Minute vent were OK) Both times when I breathed out forcefully it turned back on.

I simply turned off Smart Start and use the on/off button.
That still works to this day. :D

I am not the only one to have had this happen to them. Some machines seem to have a "glitch" in them.
Your machine may/may not be one of these.

Just thought I would chime in.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.