Dry throat with DreamStation Auto CPAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
valiant
Posts: 3
Joined: Thu Oct 15, 2020 1:18 am

Dry throat with DreamStation Auto CPAP

Post by valiant » Thu Oct 15, 2020 5:00 am

I get dry throat maybe twice a week using the machine with humidifier
my pressure is 7.0 cmH2o
Ramp 4.0
Ramp time 30sec
Flex 3
Heated tube
Mask type : Bleep dreamport

User avatar
Pugsy
Posts: 65374
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dry throat with DreamStation Auto CPAP

Post by Pugsy » Thu Oct 15, 2020 9:08 am

The most common suspect for random dry mouth or dry throat....mouth breathing when we can't blame maybe medication or hydration as the cause.

Depends on how annoying it is and how much large leak the mouth breathing might cause what we might want to do about it.

It doesn't take much mouth breathing to dry out the mouth or throat. I have had it happen with as little as 5 minutes of probable mouth breathing.

Twice a week...I don't know I would do much about it depending on how bad it is. You could try increasing the humidity setting but I don't know that it would help much but might be worth trying.

I do suggest that you use available free software to monitor leaks just to make sure that the mouth breathing is causing massive prolonged leaks that could potentially impact therapy effectiveness. I don't worry about 15 minutes of big leak but I would worry about 2 hours of big leak.
OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

valiant
Posts: 3
Joined: Thu Oct 15, 2020 1:18 am

Re: Dry throat with DreamStation Auto CPAP

Post by valiant » Fri Oct 16, 2020 12:57 am

Theres dreammapper app, the reports I get daily doesn't really show leakage at all.
CPAP.jpg
CPAP.jpg (82.52 KiB) Viewed 1158 times

User avatar
Pugsy
Posts: 65374
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dry throat with DreamStation Auto CPAP

Post by Pugsy » Fri Oct 16, 2020 9:07 am

All the DreamMapper app shows in terms of leakage is you didn't hit large leak territory with any leaks.
Large leak territory on Respironics machines is up there in terms of excess leakage. So doesn't mean no excess leaks happened...just means leaks didn't hit large leak territory.
Mouth breathing leaks don't have to hit large leak territory to dry out the mouth or throat.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.