Monitoring data wirelessly

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mcs2015
Posts: 31
Joined: Sat Dec 05, 2015 1:52 pm

Monitoring data wirelessly

Post by mcs2015 » Tue Sep 22, 2020 2:37 pm

I have a modem installed on my DreamStation and would like to monitor the detailed data on my card without having to pull the card every day. What I need to know is how to import the card wirelessly using Oscar. Appreciate any advice.

User avatar
Dog Slobber
Posts: 4272
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Monitoring data wirelessly

Post by Dog Slobber » Tue Sep 22, 2020 3:07 pm

The modem on your DreamStation is intended to upload data for your provider. There isn't a way to wirelessly transfer the data to OSCAR using the DreamStation's Modem.

The only supported way to wirelessly transfer the detail data from your DreamStation (or ResMed) to OSCAR, is to use the Toshiba Flashair WiFi SD Card and a third party transfer agent such as SleepMaster or FlashPAP.

I *believe* the Toshiba FlashAir is no longer in production. They are becoming increasingly difficult to find and the price on them is sky-rocketing. There are other WiFi SD cards, but completely unsupported.

I've been using the FlashAir for over 2 years and love it.
Battery Backup: EcoFlow Delta 2

mcs2015
Posts: 31
Joined: Sat Dec 05, 2015 1:52 pm

Re: Monitoring data wirelessly

Post by mcs2015 » Thu Oct 08, 2020 1:53 am

I just got the Toshiba FlashAir III W03 16GB Wireless SD Memory Card as I was advised.

But I haven't a clue what the next step is.

Can anyone post a link to or have step-by-step instructions to use it to transmit the data using my OSCAR software??

I use both Windows 10 and OSCAR loaded on that.

Would greatly appreciate, as I am totally lost.