Another Resmed Airsense 10 issue - Blowing too hard

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Gator5000e
Posts: 134
Joined: Mon May 07, 2018 1:07 pm

Another Resmed Airsense 10 issue - Blowing too hard

Post by Gator5000e » Wed Sep 09, 2020 2:03 pm

A few times over the last few weeks, including last night, my Airsense 10 has just started blowing so hard that the mask is being lifted off my face. It's like it's blowing at 30 instead of the 14 I have it set at. It sounds like a jet engine.

I have no idea what's causing the issue. I shut it off, unplugged it for a few minutes, changed the settings (increased the ramp, turned the ramp off, changed the pressure, went from CPAP to Auto CPAP and back) and nothing worked. I finally let it sit and changed my mask (Dreamwear FF) and headgear and it seemed to stop. I didn't detect any leaks in the mask or headgear so I have no idea what I did to solve or stop the issue.

I bought the machine from someone here on this board so I don't have a dealer I can take it to if it's malfunctioning.

Any suggestions would be appreciated. TIA.



PS - just thought of this, but does ResMed ever update the software or firmware of these machines? If so, how do they do it? Do I have to take it in or can I do it myself?

User avatar
Dog Slobber
Posts: 4258
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Another Resmed Airsense 10 issue - Blowing too hard

Post by Dog Slobber » Wed Sep 09, 2020 2:23 pm

Your signatures states an AirSense 10 CPAP, but in a previous post you mentioned you had an AutoSet.

Is it a CPAP model?

If it's an AutoSet, we can look at the graphs to see what your settings are and what pressure it made to.

And all though ResMed is capable of over-the-air firmware updates, I'm only aware of one, and that was years ago to address a humidifier problem. ResMed hasn't documented any user initiated firmware upgrades nor methods to upgrade via an authorized reseller.
Last edited by Dog Slobber on Wed Sep 09, 2020 2:33 pm, edited 1 time in total.
Battery Backup: EcoFlow Delta 2

Janknitz
Posts: 8512
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Another Resmed Airsense 10 issue - Blowing too hard

Post by Janknitz » Wed Sep 09, 2020 2:27 pm

I agree with Dog Slobber that the data will help to see what's going on.

When a machine "sounds like a jet engine" I suspect a mask leak, but you are an experienced user who should be able to tell the difference. So try to access the data.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Gator5000e
Posts: 134
Joined: Mon May 07, 2018 1:07 pm

Re: Another Resmed Airsense 10 issue - Blowing too hard

Post by Gator5000e » Wed Sep 09, 2020 2:59 pm

Thanks for the replies. Yes, it's an autoset machine that I am running on straight CPAP as suggested by my doctor. I will make the change in my signature. And I will will look a the data on Oscar (I assume that was what you were referring to when you say data). I probably should have thought of that first.