2 days ago, I bought ResMed's AutoSet for Her. I was gifted an AirFit P10 starter pack, and the technician installed the Medium for me. My Dr. prescribed 4-8 cm H2O, and I felt asleep quickly.
But I felt the Medium P10 was too big for my nostrils, and switched to the Small. Yesternight, even at 8 cm H2O, I wasn't receiving enough air to inhale and my lungs started to hurt. I then increased the pressure to 6-20 cm H2O. Only then could I fall asleep. The machine displayed a pressure of 9-11 cm H2O.
So what exactly happened? What did my changing from the Medium to Small AirFit P10 cause?
How ought changing the AirFit P10's nostril size affect you?
-
antinatalism
- Posts: 21
- Joined: Sun Oct 28, 2018 11:43 pm
Re: How ought changing the AirFit P10's nostril size affect you?
The overall volume difference of the Small nasal pillow and the Medium nasal pillow is actually quite small but it is enough that some people feel like they can't move enough air with the small. I don't know why you had a problem with 8 cm and didn't with the 6 to 20 setting because you were using 6 cm while awake.
Is the ramp feature on your machine turned on by some chance? When ramp is engaged you are only getting 4 cm to start with no matter what the settings are. I suggest turning ramp off if you find it is turned on.
I don't know what all your settings are but maybe adding in EPR to the mix would help you with the not enough air flow feeling.
https://www.respshop.com/manuals/ResMed ... %20her.pdf
If you don't know what EPR is....it's explained in the manual above.
To know if ramp is turned on...you have to go into the clinical menu area to turn it off.
Is the ramp feature on your machine turned on by some chance? When ramp is engaged you are only getting 4 cm to start with no matter what the settings are. I suggest turning ramp off if you find it is turned on.
I don't know what all your settings are but maybe adding in EPR to the mix would help you with the not enough air flow feeling.
https://www.respshop.com/manuals/ResMed ... %20her.pdf
If you don't know what EPR is....it's explained in the manual above.
To know if ramp is turned on...you have to go into the clinical menu area to turn it off.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: How ought changing the AirFit P10's nostril size affect you?
Can you describe why you felt the mediums were too large?
Remember, the pillows are not supposed to actually go into the nose. The openings should just be breaking the plane of the nostril openings. A lot of novices try to insert the pillow into the nose. This can result in leaks, blisters, discomfort, and crimping of the pillow which can also explain reduced flow.
Remember, the pillows are not supposed to actually go into the nose. The openings should just be breaking the plane of the nostril openings. A lot of novices try to insert the pillow into the nose. This can result in leaks, blisters, discomfort, and crimping of the pillow which can also explain reduced flow.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: How ought changing the AirFit P10's nostril size affect you?
NOW he tells me!Dog Slobber wrote: ↑Wed Jul 29, 2020 11:30 amCan you describe why you felt the mediums were too large?
Remember, the pillows are not supposed to actually go into the nose. The openings should just be breaking the plane of the nostril openings. A lot of novices try to insert the pillow into the nose. This can result in leaks, blisters, discomfort, and crimping of the pillow which can also explain reduced flow.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: How ought changing the AirFit P10's nostril size affect you?
I use the small--but the extra small felt like I was being deprived of air.
So, yes, it's true.
So, yes, it's true.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: How ought changing the AirFit P10's nostril size affect you?
When I went from the large pillows to medium on my P30i, it felt a little air starved. Not much, but enough to notice. I've since gotten used to it.chunkyfrog wrote: ↑Thu Jul 30, 2020 6:21 pmI use the small--but the extra small felt like I was being deprived of air.
So, yes, it's true.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2