Inhaling much faster with VAUTO. Why?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Lightweaver
Posts: 21
Joined: Wed Jul 15, 2020 10:45 pm

Inhaling much faster with VAUTO. Why?

Post by Lightweaver » Sat Jul 25, 2020 3:41 pm

Yesterday, I used my VAuto machine for the first time during the night and I felt pretty weird upon waking up. I've noticed that my breathing is much faster with it. I used to have AutoSet for Her, and my inhale time was around 2 seconds. With Vauto, my medium inhale time is barely over 1s.
Any suggestions why am I breathing so fast with VAuto? I used exactly same pressure settings as my AutoSet.

_________________
Mask
Last edited by Lightweaver on Mon Jul 27, 2020 4:46 am, edited 1 time in total.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Breathing much faster with VAUTO. Why?

Post by Dog Slobber » Sat Jul 25, 2020 4:01 pm

Post some comparitive charts.
Post the settings of two machines.
Did you change any of BiLevel oriented settings?
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Breathing much faster with VAUTO. Why?

Post by palerider » Sat Jul 25, 2020 4:28 pm

Lightweaver wrote: ↑
Sat Jul 25, 2020 3:41 pm
I used exactly same pressure settings as my AutoSet.
That's highly unlikely.

What were your settings on the autoset, and what did you set the VAuto to?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Lightweaver
Posts: 21
Joined: Wed Jul 15, 2020 10:45 pm

Re: Breathing much faster with VAUTO. Why?

Post by Lightweaver » Sat Jul 25, 2020 8:55 pm

palerider wrote: ↑
Sat Jul 25, 2020 4:28 pm
Lightweaver wrote: ↑
Sat Jul 25, 2020 3:41 pm
I used exactly same pressure settings as my AutoSet.
That's highly unlikely.

What were your settings on the autoset, and what did you set the VAuto to?
Autoset: pressure 7-10, epr 3
Vauto: Min Epap 4, max Ipap 10, ps 3

_________________
Mask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Breathing much faster with VAUTO. Why?

Post by palerider » Sun Jul 26, 2020 1:00 am

Lightweaver wrote: ↑
Sat Jul 25, 2020 8:55 pm
palerider wrote: ↑
Sat Jul 25, 2020 4:28 pm
Lightweaver wrote: ↑
Sat Jul 25, 2020 3:41 pm
I used exactly same pressure settings as my AutoSet.
That's highly unlikely.

What were your settings on the autoset, and what did you set the VAuto to?
Autoset: pressure 7-10, epr 3
Vauto: Min Epap 4, max Ipap 10, ps 3
as highly unlikely as it is, if that's what you used, then they're the same.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

slowriter
Posts: 383
Joined: Sat Jul 20, 2019 4:37 am

Re: Breathing much faster with VAUTO. Why?

Post by slowriter » Sun Jul 26, 2020 7:45 am

Lightweaver wrote: ↑
Sat Jul 25, 2020 3:41 pm
I used to have AutoSet for Her, and my inhale time was around 2 seconds. With Vauto, my medium inhale time is barely over 1s.
That's not necessarily a problem. What's your median expiratory time? The "normal" ratio is 1:2.

So let's first establish there's an actual problem.

Also, you might post the other settings (Ti, Cycle, Trigger).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: UARS; VAuto Mode, 7-15, PS 5.8

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Breathing much faster with VAUTO. Why?

Post by palerider » Sun Jul 26, 2020 9:48 am

slowriter wrote: ↑
Sun Jul 26, 2020 7:45 am
Lightweaver wrote: ↑
Sat Jul 25, 2020 3:41 pm
I used to have AutoSet for Her, and my inhale time was around 2 seconds. With Vauto, my medium inhale time is barely over 1s.
That's not necessarily a problem. What's your median expiratory time? The "normal" ratio is 1:2.

So let's first establish there's an actual problem.

Also, you might post the other settings (Ti, Cycle, Trigger).
That's often a *target* for forced ventilation, it's not something that's significant for someone that's not on a ventilator.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

slowriter
Posts: 383
Joined: Sat Jul 20, 2019 4:37 am

Re: Breathing much faster with VAUTO. Why?

Post by slowriter » Sun Jul 26, 2020 1:51 pm

palerider wrote: ↑
Sun Jul 26, 2020 9:48 am
That's often a *target* for forced ventilation, it's not something that's significant for someone that's not on a ventilator.
Why I put "normal" in quotes.

My point was simply that the inspiratory number absent other data isn't very meaningful.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: UARS; VAuto Mode, 7-15, PS 5.8

Lightweaver
Posts: 21
Joined: Wed Jul 15, 2020 10:45 pm

Re: Inhaling much faster with VAUTO. Why?

Post by Lightweaver » Mon Jul 27, 2020 3:43 am

I've looked at the stats closer and it looks like respiratory rate only increased a little, but my inhale to exhale ratio has changed from about 1:1 to over 1:3.

On the Autoset, the inhale time is all over, but usually stays well above 1s
autoset1.png
autoset1.png (635.33 KiB) Viewed 6107 times

With VAuto, it stays at about 1s in a pretty controlled manner.
vauto1.png
vauto1.png (515.95 KiB) Viewed 6107 times

How does VAuto change the inhalation time, and keep it controlled like so?
Cycle and Trigger are at Med
I increased Ti Max to 4s as I was expecting my inhales to be over 2sec based on previous data, Ti min default.

_________________
Mask

Holden4th
Posts: 571
Joined: Wed Feb 24, 2016 3:15 am
Location: Gold Coast Australia

Re: Inhaling much faster with VAUTO. Why?

Post by Holden4th » Mon Jul 27, 2020 6:41 pm

Your Ti max setting will probably be the issue. What you set this to decides when the machine thinks you should be exhaling.

I had this issue when I got my VAuto and I increased the default Ti max until it felt right. For me it's 2.5 seconds. I left the Ti min as it was. (The default setting for Ti Max is 2.0 which was not long enough for me).

Can I also suggest that you set the Trigger to Very High (detects beginning of inhalation) and the Cycle to Very Low (detects beginning of exhalation. I found the trigger setting to be effective than cycle setting.

This thread is from 2018

viewtopic/t173578/explanation-needed-fo ... Timax.html

The second post is a very good explanation of the settings by Pale Rider.

_________________
MachineMask
Additional Comments: Sleepyhead software

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Inhaling much faster with VAUTO. Why?

Post by palerider » Mon Jul 27, 2020 7:32 pm

Holden4th wrote: ↑
Mon Jul 27, 2020 6:41 pm
Your Ti max setting will probably be the issue. What you set this to decides when the machine thinks you should be exhaling.
OP said they already set TiMax to 4.

I'm actually leaning towards erroneous reporting from the autoset, or at least something strange going on over there.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Holden4th
Posts: 571
Joined: Wed Feb 24, 2016 3:15 am
Location: Gold Coast Australia

Re: Inhaling much faster with VAUTO. Why?

Post by Holden4th » Tue Jul 28, 2020 2:28 am

palerider wrote: ↑
Mon Jul 27, 2020 7:32 pm
Holden4th wrote: ↑
Mon Jul 27, 2020 6:41 pm
Your Ti max setting will probably be the issue. What you set this to decides when the machine thinks you should be exhaling.
OP said they already set TiMax to 4.

I'm actually leaning towards erroneous reporting from the autoset, or at least something strange going on over there.
I didn't read that so thanks for pointing it out.

_________________
MachineMask
Additional Comments: Sleepyhead software

slowriter
Posts: 383
Joined: Sat Jul 20, 2019 4:37 am

Re: Inhaling much faster with VAUTO. Why?

Post by slowriter » Tue Jul 28, 2020 5:43 am

What does a two-minute close up of the flow rate graph look like?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: UARS; VAuto Mode, 7-15, PS 5.8

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Inhaling much faster with VAUTO. Why?

Post by raisedfist » Tue Jul 28, 2020 10:55 am

Your respiratory rate is more or less the same. With the VAuto new settings such as Trigger and Cycle are introduced. Try setting the Cycle to "Low" and see if that fixes the problem.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Inhaling much faster with VAUTO. Why?

Post by palerider » Tue Jul 28, 2020 12:11 pm

raisedfist wrote: ↑
Tue Jul 28, 2020 10:55 am
Your respiratory rate is more or less the same. With the VAuto new settings such as Trigger and Cycle are introduced. Try setting the Cycle to "Low" and see if that fixes the problem.
Who says there is a "problem"?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.