Storing a Resmed airsense10, Autoset for a back up unit.
Storing a Resmed airsense10, Autoset for a back up unit.
Just purchased a new Resmed airsense 10, autoset to replace my oldie goldie (2015 model)Resmed airsense 10, autoset. Im going to clean the hose and make sure it goes into it's case with it slightly opened to prevent any air quality issues(not sure if that is a issue). Was wondering what to expect when I might have to use it 5 years down the road. Will it start up if needed, no doubt they have battery for power disruptions that might be dead. The 2015 model is working wonderfully at the moment and it is worth the effort to keep it for a back up just in case of a unknown. I have spare masks just for the unknown - just my nature to be prepared!
Re: Storing a Resmed airsense10, Autoset for a back up unit.
You indicate in your profile that your pressures are "locked" on your airsense 10... There is no way to lock the pressures. They can be changed in 30 seconds by pressing a couple buttons. You also asked about a battery to start your machine after it has been stored....no battery...just plug it in if it is needed. If you are worried about it, just use bit once a year.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Storing a Resmed airsense10, Autoset for a back up unit.
Thanks for letting me know about the no battery (I was thinking for internal Clock)and yes it is locked because only people who have this model would know how to access the parameter setting's on pressures. That is what I meant by locked. I always wanted to just have enough pressure to keep my Apnea's under 1 and not snore so with a ceiling max pressure of 10 it seems to work well.LSAT wrote: ↑Fri Jul 24, 2020 8:26 pmYou indicate in your profile that your pressures are "locked" on your airsense 10... There is no way to lock the pressures. They can be changed in 30 seconds by pressing a couple buttons. You also asked about a battery to start your machine after it has been stored....no battery...just plug it in if it is needed. If you are worried about it, just use bit once a year.
When it is on factory set Auto it can reach higher pressures that gave me mask issues and my Apnea count was higher & uncomfortable at times. I was able to use this machine for the last 5 years approx with excellent results that were under 1 98% of the time with minimal snore = one happy Cpap user
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Storing a Resmed airsense10, Autoset for a back up unit.
Twenty years ago, computers had battery powered internal clocks.
The time-keeping function is now assigned to a special low-power chip
that keeps time using an internally stored charge.
Even so, the clock on Resmed machines can be set by the user--so worst case--
you have to tell it when is "now" when it is powered up again.
Now Respironics--damifino.
The time-keeping function is now assigned to a special low-power chip
that keeps time using an internally stored charge.
Even so, the clock on Resmed machines can be set by the user--so worst case--
you have to tell it when is "now" when it is powered up again.
Now Respironics--damifino.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Storing a Resmed airsense10, Autoset for a back up unit.
WoW, you folks sure keep up with the Cpap equipment. Thanks again!chunkyfrog wrote: ↑Sat Jul 25, 2020 12:39 amTwenty years ago, computers had battery powered internal clocks.
The time-keeping function is now assigned to a special low-power chip
that keeps time using an internally stored charge.
Even so, the clock on Resmed machines can be set by the user--so worst case--
you have to tell it when is "now" when it is powered up again.
Now Respironics--damifino.
Re: Storing a Resmed airsense10, Autoset for a back up unit.
I have three identical APAP machines (long story why I have three not including a brick CPAP) and one is specifically for emergency backup. Naturally, it is stored in a standard ResMed travel case, but it also includes a new P10 mask with an addition pillow, and the complete standard power accessories. I have told my wife that if in an emergency situation where I suddenly endup in a hospital for some reason, that all she needs to bring me is that one bag; it has it all. Never know.....
Machine: ResMed AirSense 11 w/Humidifier
Mask Make & Model: Pillow mask
CPAP Pressure: 9.4
CPAP Reporting Software: OSCAR & SleepHQ
Mask Make & Model: Pillow mask
CPAP Pressure: 9.4
CPAP Reporting Software: OSCAR & SleepHQ
- Dog Slobber
- Posts: 4268
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Storing a Resmed airsense10, Autoset for a back up unit.
- Wipe it down.
- Next time your eligible for a new mask, get one and pack your current mask with your spare machine
- Clean the humidifier and mask, don't bother with hose. Make sure they are all completely dry.
- Pack them all up with everything you need, so it is ready to go.
- Put a couple silica moisture absorbing packs into the case and zip it up tight.
Take a couple pics of the setup screen (Mode, pressures, EPR, comfort)
No, zip it up tight. Why circulate dusty humid air into it?
Should be fine. Check on it once a year if you like.Was wondering what to expect when I might have to use it 5 years down the road.
[Will it start up if needed, no doubt they have battery for power disruptions that might be dead.
They have a little lithium Button Cell in them for the clock. Theres a lot of people still using their 2014 AS10, and older S9's with out battery issues.
Absolutely.The 2015 model is working wonderfully at the moment and it is worth the effort to keep it for a back up just in case of a unknown. I have spare masks just for the unknown
There's a lot of people who have had miserable nights, weekends, even longer because the got rid of their perfectly fine previous device.
Huh?and yes it is locked because only people who have this model would know how to access the parameter setting's on pressures.
I *think* what you're describing is *fixed*
Sounds like you experienced comfort/pressure/mask issues when you were first starting, as all of us do. Solved it by putting in some CPAP limitations.When it is on factory set Auto it can reach higher pressures that gave me mask issues and my Apnea count was higher & uncomfortable at times. I was able to use this machine for the last 5 years approx with excellent results that were under 1 98% of the time with minimal snore = one happy Cpap user
Now that you are completely stable and comfortable you may want to try optimizing your settings a bit.
People here can lend a hand.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Storing a Resmed airsense10, Autoset for a back up unit.
that's a most excellent point. hadn't thought of that.Dog Slobber wrote: ↑Sat Jul 25, 2020 10:44 am
[*]Put a couple silica moisture absorbing packs into the case and zip it up tight.
dead.
and i want to thank cannuck for bringing this up. it strikes me that the settings on my backup are WAY off as it was set more than a year ago and thus won't be the optimum settings.
time to drag it out and power it up. then make the appropriate changes and put it back in the bag with moisture absorbing packs.
where would i get those, anyway?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Dog Slobber
- Posts: 4268
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Storing a Resmed airsense10, Autoset for a back up unit.
I keep a Tupperware full of them. Every time I get some in packaging I keep them. It's important to keep them in a airtight, sealable container so they don't hit their moisture saturation point from the ambient humidity.zonker wrote: ↑Sat Jul 25, 2020 10:58 amthat's a most excellent point. hadn't thought of that.Dog Slobber wrote: ↑Sat Jul 25, 2020 10:44 am
[*]Put a couple silica moisture absorbing packs into the case and zip it up tight.
dead.
...
where would i get those, anyway?
Especially, if you're near the West Coast.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Storing a Resmed airsense10, Autoset for a back up unit.
i'll keep that in mind, should i ever move there.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Storing a Resmed airsense10, Autoset for a back up unit.
Dog Slobber wrote: ↑Sat Jul 25, 2020 10:44 am
- Wipe it down.
- Next time your eligible for a new mask, get one and pack your current mask with your spare machine
- Clean the humidifier and mask, don't bother with hose. Make sure they are all completely dry.
- Pack them all up with everything you need, so it is ready to go.
- Put a couple silica moisture absorbing packs into the case and zip it up tight.
Take a couple pics of the setup screen (Mode, pressures, EPR, comfort)No, zip it up tight. Why circulate dusty humid air into it?
Should be fine. Check on it once a year if you like.Was wondering what to expect when I might have to use it 5 years down the road.
[Will it start up if needed, no doubt they have battery for power disruptions that might be dead.
They have a little lithium Button Cell in them for the clock. Theres a lot of people still using their 2014 AS10, and older S9's with out battery issues.
Absolutely.The 2015 model is working wonderfully at the moment and it is worth the effort to keep it for a back up just in case of a unknown. I have spare masks just for the unknown
There's a lot of people who have had miserable nights, weekends, even longer because the got rid of their perfectly fine previous device.
Huh?and yes it is locked because only people who have this model would know how to access the parameter setting's on pressures.
I *think* what you're describing is *fixed*
Sounds like you experienced comfort/pressure/mask issues when you were first starting, as all of us do. Solved it by putting in some CPAP limitations.When it is on factory set Auto it can reach higher pressures that gave me mask issues and my Apnea count was higher & uncomfortable at times. I was able to use this machine for the last 5 years approx with excellent results that were under 1 98% of the time with minimal snore = one happy Cpap user
Now that you are completely stable and comfortable you may want to try optimizing your settings a bit.
People here can lend a hand.