Resmed aircurve 10 - whistling noise

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
pruettdj
Posts: 1
Joined: Tue Jun 16, 2020 5:30 am

Resmed aircurve 10 - whistling noise

Post by pruettdj » Tue Jun 16, 2020 5:47 am

So, the last few nights my machine has been making a gentle whistling noise, I think it’s whenever it decides it needs to turn up the pressure? On its face the noise isn’t terrible, but I’m sleeping with a six month old and the whistling sounds just like him waking up? Like not the screaming times, like the ‘hey, what’s going on’ baby squeak. Anyway, basically it wakes me up constantly. Is this common? It sounds like an internal seal is coming loose or something, but it sounds like opening one up will effectively break it?

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed aircurve 10 - whistling noise

Post by Pugsy » Tue Jun 16, 2020 7:44 am

Pull the water chamber out and look inside and feel around for the 2 seals in the back. That's an easy place to look. Maybe one is crimped or something. One has a fairly easy replacement to get. It's the outlet air seal.
https://www.cpap.com/productpage/resmed ... air-outlet
The other one I haven't been able to find but I would assume it fails less since the other one is easily found.

Is it still under warranty? How long have you had it?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed aircurve 10 - whistling noise

Post by LSAT » Tue Jun 16, 2020 7:46 am

If you have had the machine less than 2 years it is under warranty. Take it back to your DME. But, first....make sure the seal on the water chamber is installed properly and be sure there is no leak in your hose.

User avatar
Dog Slobber
Posts: 4261
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed aircurve 10 - whistling noise

Post by Dog Slobber » Tue Jun 16, 2020 7:48 am

  • Check for leaks
  • Tighten and ensure hose and mask are seated properly
  • Ensure humidifier and seal are seated properly. The ResMed humidifier not being pushed all the way in is a very common cause of noise
  • Open filter door, ensure filter is seated properly and doesn't need changing
  • Place device on mat or rubber cupboard stop buttons to dampen noise.
Battery Backup: EcoFlow Delta 2

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Resmed aircurve 10 - whistling noise

Post by Okie bipap » Tue Jun 16, 2020 10:10 am

I get a whistle from my machine when the water chamber is not seated correctly. When my wife adds water to the machine, she does not get it pushed in enough when she puts the water chamber into the machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

WorkingAtDad
Posts: 1
Joined: Thu Sep 26, 2024 6:10 am

Re: Resmed aircurve 10 - whistling noise

Post by WorkingAtDad » Thu Sep 26, 2024 6:36 am

I'm not sure if anyone from the original discussion is still checking this, but I too have begun to have a whistling noise from my ResMed aircurve 10 Vauto device. The difference I have from the other posts is that the whistling noise only occurs when I exhale. I have learned to sleep with it but it is very annoying to my wife. Any ideas?

dmoss74
Posts: 30
Joined: Fri Nov 26, 2010 2:08 am

Re: Resmed aircurve 10 - whistling noise

Post by dmoss74 » Thu Sep 26, 2024 7:51 pm

check your hose

_________________
MachineMask

User avatar
babydinosnoreless
Posts: 2364
Joined: Fri Nov 02, 2018 2:53 pm

Re: Resmed aircurve 10 - whistling noise

Post by babydinosnoreless » Thu Sep 26, 2024 8:29 pm

Do you have a short hose ? Check the connection from hose to mask.