Newbie here

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
REDKBRECH
Posts: 8
Joined: Sun May 24, 2020 3:26 pm
Location: Kokomo, Indiana

Newbie here

Post by REDKBRECH » Mon Jun 08, 2020 5:21 pm

what does it indicate when the CPAP machine after it has been on for about 5 hours starts making sounds in the tubing like popping, knocking, etc. My humidifier is set on 5 & the tubing temp is set on 68 which blows out cooler air which I like. The humidifier on 4 makes my nose stuffy but sometimes on 5 I have water dripping from my mask. Please give me some suggestions or advice. Thanks

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: Not sure about humidifier, machine says Humid Air S

User avatar
Pugsy
Posts: 65374
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie here

Post by Pugsy » Mon Jun 08, 2020 5:26 pm

You've got rain out or condensation in the hose and/or mask.
When water gets in the hose it makes an awful racket.
Increase your hose air temp.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13408
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Newbie here

Post by LSAT » Mon Jun 08, 2020 8:25 pm

Usually cause by condensation..cool room and warm moist air in the tube. Get a tube cover to insulate it, get a heated hose, reduce humidity, keep the hose under the covers, try a hose lift to make the water run back to the humidifier. Any 1 or a combination of these will solve the problem.

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Newbie here

Post by Sheriff Buford » Tue Jun 09, 2020 6:02 am

LSAT eluded to it, but be sure the machine is lower than the bed.

Sheriff