grounded wrote: ↑Tue May 26, 2020 2:38 pm
palerider wrote: ↑Tue May 26, 2020 1:25 pm
bump your min up to 12 and turn off EPR.
Interesting - my clinician says EPR is "just fine" and can be set to 2. She seems to have no problem with it.
Most of the time for the majority of the people using any form of exhale relief is indeed just fine and not a problem.
Unfortunately there are a few people (very small minority) who find that the inhale/exhale difference that using exhale relief will be a trigger for an increase in the number of central apneas. It has to do with the exchange of blood gases and primarily the blood levels of carbon dioxide.
Using exhale relief or bilevel pressures can cause a few people to blow off too much carbon dioxide so the levels don't increase in the blood to the point that the brain says "oh...carbon dioxide level is up...need to send the breathe signal"...so you don't breathe and that's a central apnea...no effort to breathe. It's actually the carbon dioxide levels in the blood that drive the "breathe human" trigger and not oxygen levels which is what most people think drive our breathing.
EPR creates a bilevel pressure situation...one pressure for inhale and one different pressure for exhale...we call that difference pressure support. It can be 1 cm to much higher difference when using a machine that will allow the pressure support to go higher than the 3 cm that EPR is limited to.
I have a friend who gets 15 to 20 centrals per hour when she uses a 4 cm difference between inhale and exhale. We thought for sure she needed one of those fancy high end machines that will treat centrals but we tried using a 3 cm pressure support and her centrals dropped down to less than 1.0 per hour. Go figure that one....a little 1 cm difference made all the difference in the world for her.
The reason we suggest turning EPR off...just to see if you are one of that small minority that has that gas exchange problem where your carbon dioxide levels just don't get high enough for the brain to send the "breathe human" signal.
If it makes a big difference then you can play with EPR to see if there is a line in the sand where EPR causes centrals and below that line it doesn't. Like my friend that found out at 4 it was a problem and at 3 no problem. She really needed the exhale relief because here pressure needs were in the high teens so going totally without any exhale relief would be a problem for her. We got lucky and found a workable compromise and she didn't have to go buy that high dollar fancy machine that would treat her centrals.
Now you might not be one of those people...it might make no difference at all but it is something you can do to see for yourself if you are in that very small minority or not...and it hurts nothing to give it a try.