ResMed s9 humidifier

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Cpaplynnie123
Posts: 1
Joined: Thu Apr 16, 2020 12:55 am

ResMed s9 humidifier

Post by Cpaplynnie123 » Thu Apr 16, 2020 1:00 am

When I turn the dial on my Resmed 9, I can either adjust the ramp setting or the temperature. (There’s an icon with a thermometer and water drop) But now there doesn’t seem to be a way to change the humidity - I’ve been using it for years and I used to set it at 6. Now there suddenly eems to be no way to get to that water drop screen when I turn the dial and the air seems dry. Does anyone have a suggestion? Thank you!

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed s9 humidifier

Post by LSAT » Thu Apr 16, 2020 6:58 am

Are you using the heated hose?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: ResMed s9 humidifier

Post by chunkyfrog » Thu Apr 16, 2020 9:23 am

I believe that it is possible to change the temp/humidity from automatic to manual
in the clinician's menu.
I tried it once, but I liked auto better.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed s9 humidifier

Post by Pugsy » Thu Apr 16, 2020 9:34 am

The humidity level is probably set to "automatic" which is why you can't seem to change it.
You need to go into the setup menu area and change it to "manual".

https://www.apneaboard.com/adjust-cpap- ... tup-manual

I suggest you go to the above link and request the provider manual which explains how to get to the setup area where you can change the humidity control main setting.