my machine dries my mouth out so bad. What should the settings be set on as far as the water goes?

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Respironics System One

Post by Janknitz » Mon Apr 13, 2020 5:34 pm

1. Dry mouth is almost always due to mouth breathing. You can crank the humidifier all the way up, but you will still have the problem. And if you don't have a full face mask, you are getting no benefit from your CPAP if the air is going right out of your mouth. If you have a full face mask, at least your treatment can work. So please fill out your member profile (Click on your name on the right hand side of the screen under the red cross and fill in your member profile) with your equipment so we can help you. There are some ways to determine if it is mouth breathing and what to do about it.

2. The System One is an old, outdated machine. Hope you didn't just buy it new. It has a known flaw with the humidifier (manufacturer's defect) so it's likely you are not getting any humidification anyway. And they are not that great at humidification, even when they do work. Tell us how you ended up with that machine.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
jhuckabyjr
Posts: 32
Joined: Tue May 23, 2017 9:14 pm

Re: Respironics System One

Post by jhuckabyjr » Mon Apr 13, 2020 6:42 pm

My first CPAP was a PR System One, brand new. Humidifier didn't work from day one. The DME replaced it with another new one and it worked. I migrated to a Dreamstation, then an Airsense 10 Autoset. I've never looked back.
Airsense 10 Autoset with humidifier, Climateline Heated Tube, Airfit N30i.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Respironics System One

Post by zonker » Mon Apr 13, 2020 6:45 pm

interesting!

i once owned a system one and did not have this problem.

must've dodged a bullet.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Respironics System One

Post by squid13 » Tue Apr 14, 2020 8:18 am

If your humidifier is not working here is a thread on how to repair it, this is a known problem. This could be your problem. viewtopic.php?f=1&t=102275&p=959965&hil ... ir#p959965