Concerning Noise from Resmed 10 Machine

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
matthewb
Posts: 4
Joined: Sun Mar 22, 2020 3:08 pm
Location: N.C.

Concerning Noise from Resmed 10 Machine

Post by matthewb » Sun Mar 22, 2020 3:15 pm

Hi All! Im new to the forums. I've been on sleep therapy since July of '19. I had a sleep study done and was told that my AHI was 55. I use the machine every night for at least 8 hours. My AHI is now down to readings ranging from .8 AHI - 2 AHI. I got used to the machine and feel like maybe I sleep better...but ever since starting I've been struggling with breathlessness all of the time. Anyways that's a side note.

Last night my machine started making this weird whining/whirring sound. I did everything that I could to figure out where it's coming from but can't seem to figure it out. It's not the humidifier reservoir. It almost sounds like it's coming from inside the unit. It's loud enough that it is disrupting mine and my wife's sleep now. Any thoughts?

Here is a link to a video showing the noise.
https://www.youtube.com/watch?v=cVwLGyOdmF0

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Started Therapy July, 2019

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Concerning Noise from Resmed 10 Machine

Post by Pugsy » Sun Mar 22, 2020 3:20 pm

As soon as possible notify your equipment supplier about the strange new noise.

These machines do up and fail. Motors or fans do run into trouble.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

matthewb
Posts: 4
Joined: Sun Mar 22, 2020 3:08 pm
Location: N.C.

Re: Concerning Noise from Resmed 10 Machine

Post by matthewb » Sun Mar 22, 2020 3:34 pm

Okay, thanks. I'll see if they are open tomorrow. Are they generally under a warranty for a certain amount of time? I literally just had to send off $300 to pay it off about a month ago when my insurance changed and they refused to continue paying the monthly amount for it.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Started Therapy July, 2019

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Concerning Noise from Resmed 10 Machine

Post by Pugsy » Sun Mar 22, 2020 3:40 pm

ResMed has a 2 year warranty on their machines providing you haven't been using the SoClean device which ResMed believes could be compromising the machines and if you have that opens up a whole new can of worms since they just announced this change to their warranty in Feb of this year.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4322
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Concerning Noise from Resmed 10 Machine

Post by Dog Slobber » Sun Mar 22, 2020 4:49 pm

Make sure the Humidifier is seated properly. ResMed 10's are notorious for making weird sounds if it's not in correct and not shut all the way.

Remove it, then push it back in.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Concerning Noise from Resmed 10 Machine

Post by zonker » Sun Mar 22, 2020 6:05 pm

Dog Slobber wrote:
Sun Mar 22, 2020 4:49 pm
Make sure the Humidifier is seated properly. ResMed 10's are notorious for making weird sounds if it's not in correct and not shut all the way.

Remove it, then push it back in.
came to see this, left satisfied.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

matthewb
Posts: 4
Joined: Sun Mar 22, 2020 3:08 pm
Location: N.C.

Re: Concerning Noise from Resmed 10 Machine

Post by matthewb » Sun Mar 22, 2020 8:21 pm

Dog Slobber wrote:
Sun Mar 22, 2020 4:49 pm
Make sure the Humidifier is seated properly. ResMed 10's are notorious for making weird sounds if it's not in correct and not shut all the way.

Remove it, then push it back in.
Hey, thanks for the reply. Yes I know of the sounds you are referring to. I occasionally have those fitment noises. But this is 100% not that. Those fitment noises go away when I take the reservoir out and put it back in. No matter what I do to the machine this noise won't go away. And it sounds like it's coming from inside the main unit. :(

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Started Therapy July, 2019

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Concerning Noise from Resmed 10 Machine

Post by Miss Emerita » Sun Mar 22, 2020 9:17 pm

I listened to the sound on the video, and it sounds SO WRONG! Hope you can get a replacement quickly, easily, and at no cost.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

recipefordisaster
Posts: 149
Joined: Wed Jan 13, 2016 6:32 am

Re: Concerning Noise from Resmed 10 Machine

Post by recipefordisaster » Mon Mar 23, 2020 5:28 am

I just went through this with my machine, which is about 2 years old. It was too loud to manage sleeping (even with all of my noise dampening tricks and a loud white noise machine) because the sound is so annoying... and I figure the motor wants to fail, so I bought a new one on Craigslist.

This is my second Airsense 10 to fail before it was 3 years old. Grrrrr. It’s still cheaper to buy barely used machines on Craigslist than to pay my copay. Now I have 2 spares that sort of work. The other has a humidifier fault.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Use an iPad for everything so rarely download to Sleepyhead. Fixed pressure at 6 with 3 EPR, AHI around 1.