Been a while

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Jack o
Posts: 58
Joined: Wed Jul 24, 2013 4:25 pm
Location: New Jersey

Been a while

Post by Jack o » Sun Mar 22, 2020 12:48 pm

Been a while. I am a mouth breather. I've gone up and down in high pressure between 8.0 to 13.0 (no gas and gas) and receive varying AHI's - avg. 8. Lately whether set higher or lower I am continuously getting AHI's averaging about 8. I switched on 3/14 from an F30 to an F30i but probably getting more leakage. Anyone have any suggestions? I have tried to attach my report - not sure if it worked. Won't allow me to drag in a PDF or just not doing correctly.
Thanks.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 1:50 pm

Jack o wrote: ↑
Sun Mar 22, 2020 12:48 pm
Been a while. I am a mouth breather. I've gone up and down in high pressure between 8.0 to 13.0 (no gas and gas) and receive varying AHI's - avg. 8. Lately whether set higher or lower I am continuously getting AHI's averaging about 8. I switched on 3/14 from an F30 to an F30i but probably getting more leakage. Anyone have any suggestions? I have tried to attach my report - not sure if it worked. Won't allow me to drag in a PDF or just not doing correctly.
Thanks.
jack, see this thread from the last time you asked-

viewtopic.php?t=176717

re-read what i told you via PM or see chicago grannies answer below.

if you still have trouble, ask. we can get you through this.

good luck!
Last edited by zonker on Sun Mar 22, 2020 2:20 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15542
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Been a while

Post by ChicagoGranny » Sun Mar 22, 2020 2:09 pm

Jack o wrote: ↑
Sun Mar 22, 2020 12:48 pm
I have tried to attach my report
At the bottom of your next post, click on Attachments/Add Files/Place inline. It's a good idea to click Preview before clicking Submit.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Been a while

Post by palerider » Sun Mar 22, 2020 5:36 pm

Jack o wrote: ↑
Sun Mar 22, 2020 12:48 pm
Been a while. I am a mouth breather. I've gone up and down in high pressure between 8.0 to 13.0 (no gas and gas) and receive varying AHI's - avg. 8. Lately whether set higher or lower I am continuously getting AHI's averaging about 8. I switched on 3/14 from an F30 to an F30i but probably getting more leakage. Anyone have any suggestions? I have tried to attach my report - not sure if it worked. Won't allow me to drag in a PDF or just not doing correctly.
Thanks.
Nobody wants to see a PDF.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 6:03 pm

palerider wrote: ↑
Sun Mar 22, 2020 5:36 pm
Jack o wrote: ↑
Sun Mar 22, 2020 12:48 pm
Been a while. I am a mouth breather. I've gone up and down in high pressure between 8.0 to 13.0 (no gas and gas) and receive varying AHI's - avg. 8. Lately whether set higher or lower I am continuously getting AHI's averaging about 8. I switched on 3/14 from an F30 to an F30i but probably getting more leakage. Anyone have any suggestions? I have tried to attach my report - not sure if it worked. Won't allow me to drag in a PDF or just not doing correctly.
Thanks.
Nobody wants to see a PDF.
do you mean in general or just in this case?
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Jack o
Posts: 58
Joined: Wed Jul 24, 2013 4:25 pm
Location: New Jersey

Re: Been a while

Post by Jack o » Sun Mar 22, 2020 6:14 pm

I followed Granny's procedure- clicked attachments, then add files. That took me to a listing of my downloads. Nothing there. Then I clicked on Desktop listing and it was not in the listing even though I have it on my desktop. What am I doing wrong?

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 6:45 pm

Jack o wrote: ↑
Sun Mar 22, 2020 6:14 pm
I followed Granny's procedure- clicked attachments, then add files. That took me to a listing of my downloads. Nothing there. Then I clicked on Desktop listing and it was not in the listing even though I have it on my desktop. What am I doing wrong?
look in the documents folder. then oscar data folder (or sleepyhead if that's what you are using) then screenshots folder.

see it there?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Been a while

Post by palerider » Sun Mar 22, 2020 7:06 pm

zonker wrote: ↑
Sun Mar 22, 2020 6:03 pm
palerider wrote: ↑
Sun Mar 22, 2020 5:36 pm
Jack o wrote: ↑
Sun Mar 22, 2020 12:48 pm
Been a while. I am a mouth breather. I've gone up and down in high pressure between 8.0 to 13.0 (no gas and gas) and receive varying AHI's - avg. 8. Lately whether set higher or lower I am continuously getting AHI's averaging about 8. I switched on 3/14 from an F30 to an F30i but probably getting more leakage. Anyone have any suggestions? I have tried to attach my report - not sure if it worked. Won't allow me to drag in a PDF or just not doing correctly.
Thanks.
Nobody wants to see a PDF.
do you mean in general or just in this case?
:lol: :lol: :lol:
In this case.
We want a screenshot, as the numerous "how to" pages say, as you link says :)
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 7:09 pm

palerider wrote: ↑
Sun Mar 22, 2020 7:06 pm
zonker wrote: ↑
Sun Mar 22, 2020 6:03 pm
palerider wrote: ↑
Sun Mar 22, 2020 5:36 pm


In this case.
We want a screenshot, as the numerous "how to" pages say, as you link says :)
just goofin' you, ya big lug.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Jack o
Posts: 58
Joined: Wed Jul 24, 2013 4:25 pm
Location: New Jersey

Re: Been a while

Post by Jack o » Sun Mar 22, 2020 7:14 pm

Zonker, I went into documents, then oscar data file, then screenshots and there are old screenshots in there. How do I make new screen shots and put them in there and then how do I get them out of there?

User avatar
Pugsy
Posts: 65439
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Been a while

Post by Pugsy » Sun Mar 22, 2020 7:55 pm

What operating system are you using on your computer?
If using Windows....use the Snipping Tool to take screen shots and save them to your desktop so they are easy to find to link to. Google "how to use the Snipping TooL"
If using some other operating system...there's usually some way to take a screen shot. Google it. "How to take a screen shot on a so and so computer".
OSCAR will even take a screen shot for you...Click on View on the top left side and see "take screen shot or F12".
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 8:02 pm

Jack o wrote: ↑
Sun Mar 22, 2020 7:14 pm
Zonker, I went into documents, then oscar data file, then screenshots and there are old screenshots in there. How do I make new screen shots and put them in there and then how do I get them out of there?
those screenshots, made by hitting F12 if you are on a windows machine, should go there. if not, search your computer for "screenshot-2020*" but without the quotation marks. they should show up.

i know you can do this because you've done it before. please see the link i gave above to your original thread. or let me just put it here again-

viewtopic/t176717/Improving.html

you can do this. i've got faith in you. try again and let us know how it works out.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Jack o
Posts: 58
Joined: Wed Jul 24, 2013 4:25 pm
Location: New Jersey

Re: Been a while

Post by Jack o » Sun Mar 22, 2020 8:54 pm

I think I got it?

screenshot-20200322-225312.png
screenshot-20200322-225312.png (67.48 KiB) Viewed 7807 times
screenshot-20200322-225107.png
screenshot-20200322-225107.png (61.7 KiB) Viewed 7807 times
screenshot-20200322-224638.png
screenshot-20200322-224638.png (80.64 KiB) Viewed 7807 times

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Been a while

Post by zonker » Sun Mar 22, 2020 9:07 pm

Jack o wrote: ↑
Sun Mar 22, 2020 8:54 pm
I think I got it?
you got it!!

now, one more thing. (you had to know there would be one more thing, right?)

please put your graphs in this order-

Suggested Support Order for your charts
Event Flags
Flow Rate
Pressure
Leak Rate
Flow Limit (Only for Resmed machines, not on Respironics)

this way, the experts (I AM NOT ONE) can see what they need and can help you easier.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Been a while

Post by Miss Emerita » Sun Mar 22, 2020 9:14 pm

Good job on the screenshot and attaching. Next time you post a chart, do work on the layout so that it looks like the sample Zonker posted.

* Turn off the pie chart (File - Preferences - Appearance - uncheck the pie box)

* Stack these five graphs at the top:

Events
Flow rate
Pressure
Leaks
Flow limitations

Don't include any other graphs in the screenshot; move them down out of the way. People will ask for them if they want to see them.

If you have trouble squeezing in those five charts, you can grab the horizontal gray bars separating the graphs and move them up until those five graphs all fit.

It's true you have leak problems, but the bigger problem is that you are seeing too many obstructive events and flow limitations. An expert will need to give you advice about how to address that.
Oscar software is available at https://www.sleepfiles.com/OSCAR/