Cellular signal low

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
dawnfogle
Posts: 1
Joined: Sun Feb 23, 2020 4:45 pm

Cellular signal low

Post by dawnfogle » Mon Feb 24, 2020 8:24 pm

Help! I just started the cpap ResMed airsense 10 my insurance compliance requires 21 days with at least 4 hours of usage to keep machine. I live in a VERY rural area, as in most of my neighbors are amish... My problem is I only show one bar signal on machine and its not showing all hours of usage so maybe I don't have enough signal sometime during the night to show all hours being used? Is there another way to show hours used?
Im hoping this makes sense. I work in medical field (hospice) am used to putting on people, just not using on myself

_________________
MachineMask

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Cellular signal low

Post by Okie bipap » Mon Feb 24, 2020 9:08 pm

See if your machine has the SD card installed. If not, purchase one at either Walmart, Best Buy, or somewhere else. Your doctor or DME will be able to recover information from the card. In addition, the card will enable you to use OSCAR to monitor your treatment.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Cellular signal low

Post by jimbud » Mon Feb 24, 2020 9:48 pm

dawnfogle wrote: ↑
Mon Feb 24, 2020 8:24 pm
Help! I just started the cpap ResMed airsense 10 my insurance compliance requires 21 days with at least 4 hours of usage to keep machine. I live in a VERY rural area, as in most of my neighbors are amish... My problem is I only show one bar signal on machine and its not showing all hours of usage so maybe I don't have enough signal sometime during the night to show all hours being used? Is there another way to show hours used?
Im hoping this makes sense. I work in medical field (hospice) am used to putting on people, just not using on myself
You can just take the machine with you to show whoever the compliance data on the device screen through the sleep report section. It stores 365 days of usage.

Best is to do what Okie Bipap said and get a SD card. You can let them get the data off it.

Get Oscar.

https://www.sleepfiles.com/OSCAR/

You might/will need it in the future. :D

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Cellular signal low

Post by LSAT » Mon Feb 24, 2020 9:50 pm

I would think that the DME would contact you if they are not getting compliance data. Get an SD card anyway. You will need it to monitor your progress.

User avatar
loggerhead12
Posts: 585
Joined: Fri Sep 01, 2017 4:00 pm

Re: Cellular signal low

Post by loggerhead12 » Mon Feb 24, 2020 11:33 pm

Bars aren't a great indicator of signal. It doesn't take much for the device to connect. Mine consistently shows no bars but my results still show up in MyAir.

Speaking of which, does your data show up in MyAir?

You say "some" of your hours don't show up - if any show up, they will all show up. It's an all or nothing thing for any particular day. A low signal won't give partial results, it won't show any results at all.

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: Cellular signal low

Post by D.H. » Tue Feb 25, 2020 9:26 am

The machine is recording the data internally, whether it is being transmitted or not. I believe that this is the case even in the absence of an SD card.

Therefore, the data are not lost, merely unreported. It can be recovered with an SD Card, bringing the machine in, or even one "lucky" successful transmission.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max
Last edited by D.H. on Tue Feb 25, 2020 6:47 pm, edited 1 time in total.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Cellular signal low

Post by prodigyplace » Tue Feb 25, 2020 10:02 am

Okie bipap wrote: ↑
Mon Feb 24, 2020 9:08 pm
See if your machine has the SD card installed. If not, purchase one at either Walmart, Best Buy, or somewhere else. Your doctor or DME will be able to recover information from the card. In addition, the card will enable you to use OSCAR to monitor your treatment.
If your machine does not have an SD card, ask your DME for the card that was shipped with the unit. Resmed machines are shipped with SD Cards.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Cellular signal low

Post by Dog Slobber » Tue Feb 25, 2020 10:07 am

dawnfogle wrote: ↑
Mon Feb 24, 2020 8:24 pm
My problem is I only show one bar signal on machine and its not showing all hours of usage so maybe I don't have enough signal sometime during the night to show all hours being used? Is there another way to show hours used?
How are you determining that it is not showing all hours of usage?

Are you using MyAir?
Looking at the report on screen?
Battery Backup: EcoFlow Delta 2

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Cellular signal low

Post by jimbud » Tue Feb 25, 2020 10:11 am

Dog Slobber wrote: ↑
Tue Feb 25, 2020 10:07 am
dawnfogle wrote: ↑
Mon Feb 24, 2020 8:24 pm
My problem is I only show one bar signal on machine and its not showing all hours of usage so maybe I don't have enough signal sometime during the night to show all hours being used? Is there another way to show hours used?
How are you determining that it is not showing all hours of usage?

Are you using MyAir?
Looking at the report on screen?
Good point.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Cellular signal low

Post by palerider » Tue Feb 25, 2020 10:17 am

dawnfogle wrote: ↑
Mon Feb 24, 2020 8:24 pm
My problem is I only show one bar signal on machine and its not showing all hours of usage so maybe I don't have enough signal sometime during the night to show all hours being used?
The two have absolutely nothing to do with each other, all the cell signal does is report your data back to Resmed. It's totally optional.
dawnfogle wrote: ↑
Mon Feb 24, 2020 8:24 pm
Is there another way to show hours used?
Oscar, if you actually want to know what's going on.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Cellular signal low

Post by Janknitz » Tue Feb 25, 2020 12:55 pm

People think the machine isn't showing all hours of usage because of the way it records. It doesn't put in the night before until noon, so that if you go back to sleep in the morning that part is also recorded. Check again this evening and it should all be there.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Cellular signal low

Post by chunkyfrog » Tue Feb 25, 2020 1:42 pm

Janknitz wrote: ↑
Tue Feb 25, 2020 12:55 pm
People think the machine isn't showing all hours of usage because of the way it records. It doesn't put in the night before until noon, so that if you go back to sleep in the morning that part is also recorded. Check again this evening and it should all be there.
Unless you slept past noon (on the cpap's internal clock)
At "noon", a new day is started, and only the time after noon is displayed.
But all time is recorded, though--each "day" with its own total time "slept", even in pieces.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

RobertS975
Posts: 211
Joined: Tue Feb 11, 2014 6:03 pm
Location: MA

Re: Cellular signal low

Post by RobertS975 » Tue Feb 25, 2020 7:40 pm

prodigyplace wrote: ↑
Tue Feb 25, 2020 10:02 am
Okie bipap wrote: ↑
Mon Feb 24, 2020 9:08 pm
See if your machine has the SD card installed. If not, purchase one at either Walmart, Best Buy, or somewhere else. Your doctor or DME will be able to recover information from the card. In addition, the card will enable you to use OSCAR to monitor your treatment.
If your machine does not have an SD card, ask your DME for the card that was shipped with the unit. Resmed machines are shipped with SD Cards.
My wife and I now own 4 Resmed 10 Airsense machines, all acquired new. None of them came with an SD card.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Cellular signal low

Post by prodigyplace » Tue Feb 25, 2020 7:49 pm

I stopped my DME from keeping mine by insisting my doctor said I needed to bring the card in on my next visit. He was just going to throw it in his desk drawer saying I did not need it.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

arhdc
Posts: 19
Joined: Tue Feb 04, 2020 11:17 am

Re: Cellular signal low

Post by arhdc » Tue Feb 25, 2020 7:58 pm

I live in the boondocks, one bar of cellular signal is rare and precious. ResMed still gets data from my Airsense 10 Auto.
~Aaron~

Resmed Airsense 10 Auto
Dreamwear Nasal Pillow