Number in parenthesis in Oscar
Number in parenthesis in Oscar
What does the number in parenthesis mean?
_________________
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: Number in parenthesis in Oscar
Doesn't mean much on a Respironics machine report.
Close approximation to duration of events in seconds but not 100% because Respironics machines don't record/report event duration.
Now on a ResMed machine report it is event duration in seconds.
Close approximation to duration of events in seconds but not 100% because Respironics machines don't record/report event duration.
Now on a ResMed machine report it is event duration in seconds.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Number in parenthesis in Oscar
It means the number of seconds the event lasted.
However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.
*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.
*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Number in parenthesis in Oscar
ResMed doesn't measure event duration for hyponeas...only apneas.
So with SleepyHead/OSCAR you get a minimum 10 second duration for hyponeas because the hyponea has to have lasted at least 10 seconds to earn a flag.
Respironics machine use some sort of marker system and that's what is shown in the parentheses...it's a close approximation of event duration if one were to zoom in and actually count the duration of the event.
Sometimes it's fairly accurate and sometimes it's way off though.
So with SleepyHead/OSCAR you get a minimum 10 second duration for hyponeas because the hyponea has to have lasted at least 10 seconds to earn a flag.
Respironics machine use some sort of marker system and that's what is shown in the parentheses...it's a close approximation of event duration if one were to zoom in and actually count the duration of the event.
Sometimes it's fairly accurate and sometimes it's way off though.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Number in parenthesis in Oscar
With the old Sleepyhead, hypopneas were always reported in seconds using ResMed S9.Pugsy wrote: ↑Mon Feb 03, 2020 9:26 amResMed doesn't measure event duration for hyponeas...only apneas.
So with SleepyHead/OSCAR you get a minimum 10 second duration for hyponeas because the hyponea has to have lasted at least 10 seconds to earn a flag.
Respironics machine use some sort of marker system and that's what is shown in the parentheses...it's a close approximation of event duration if one were to zoom in and actually count the duration of the event.
Sometimes it's fairly accurate and sometimes it's way off though.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Number in parenthesis in Oscar
Apparently one of the things we lost with the AirSense/AirCurve models was hyponea duration recording.
I get 10 seconds on all my AirCurve 10 hyponeas as well as my AutoSet.
I guess I should modify my statement though.....S9 models gave us event duration for hyponeas....AirSense/AirCurve 10 models do not.
Obviously the software can't show something if the machine doesn't record it.
I have verified with ResScan that all I get is 10 seconds for all hyponeas so it isn't just the software.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
Joshua_CPAP
- Posts: 1
- Joined: Tue Feb 02, 2021 1:10 pm
Re: Number in parenthesis in Oscar
Where did you learn that it's not accurate for Hypopneas? Can you cite your sources please.Dog Slobber wrote: ↑Mon Feb 03, 2020 9:15 amIt means the number of seconds the event lasted.
However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.
*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
Re: Number in parenthesis in Oscar
.
deleted by author as what i said was a load of bollocks.
thanks, dog, for correcting me in private.
deleted by author as what i said was a load of bollocks.
thanks, dog, for correcting me in private.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Last edited by zonker on Thu Apr 29, 2021 8:50 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Number in parenthesis in Oscar
My sources:Joshua_CPAP wrote: ↑Thu Apr 29, 2021 4:30 pmWhere did you learn that it's not accurate for Hypopneas? Can you cite your sources please.Dog Slobber wrote: ↑Mon Feb 03, 2020 9:15 amIt means the number of seconds the event lasted.
However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.
*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
- experience
- the fact that every Hypopnea time on an Air10 device, that has ever been reported by both SleepyHead and OSCAR has been exactly 10 seconds. So either that 10 seconds is not accurate, or it is the biggest coincidence in the history in the 13.8 billion years of our universe's existence
- the fact that, as I stated, the timing value for Hypopnea's was removed in OSCAR
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- BlueDragon
- Posts: 567
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: Number in parenthesis in Oscar
In fact, the "10" you see was not reported by ResMed AirSense 10 CPAP machines. It was arbitrarily added to the number reported (0) by Sleepyhead and early versions of OSCAR. OSCAR no longer does that, so you don't see (10) any more.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: Number in parenthesis in Oscar
can you cite your sources please.Dog Slobber wrote: ↑Thu Apr 29, 2021 7:58 pm
in the history in the 13.8 billion years of our universe's existence
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Number in parenthesis in Oscar
Chicago Granny's birth certificatezonker wrote: ↑Thu Apr 29, 2021 8:53 pmcan you cite your sources please.Dog Slobber wrote: ↑Thu Apr 29, 2021 7:58 pm
in the history in the 13.8 billion years of our universe's existence
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Number in parenthesis in Oscar
Dog Slobber wrote: ↑Thu Apr 29, 2021 8:59 pmChicago Granny's birth certificatezonker wrote: ↑Thu Apr 29, 2021 8:53 pmcan you cite your sources please.Dog Slobber wrote: ↑Thu Apr 29, 2021 7:58 pm
in the history in the 13.8 billion years of our universe's existence

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Number in parenthesis in Oscar
Papillion, NE?
Are you USAF, or a family member?
If so, thank you for your service.
"They also serve, who stand and wait."
Are you USAF, or a family member?
If so, thank you for your service.
"They also serve, who stand and wait."
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Number in parenthesis in Oscar
Katerific, the OP hasn't been around since September 2020. Joshua_CPAP necro-bumped a year old, no longer relevant post to scrutinize something he doesn't know anything about.chunkyfrog wrote: ↑Fri Apr 30, 2021 10:53 amPapillion, NE?
Are you USAF, or a family member?
If so, thank you for your service.
"They also serve, who stand and wait."
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2