Number in parenthesis in Oscar

Post by Katerific » Mon Feb 03, 2020 9:08 am

What does the number in parenthesis mean?
Annotation 2020-02-03 090541.png
Annotation 2020-02-03 090541.png (6.98 KiB) Viewed 1826 times

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Number in parenthesis in Oscar

Post by Pugsy » Mon Feb 03, 2020 9:12 am

Doesn't mean much on a Respironics machine report.
Close approximation to duration of events in seconds but not 100% because Respironics machines don't record/report event duration.
Now on a ResMed machine report it is event duration in seconds.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Number in parenthesis in Oscar

Post by Dog Slobber » Mon Feb 03, 2020 9:15 am

It means the number of seconds the event lasted.

However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.

*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Number in parenthesis in Oscar

Post by Pugsy » Mon Feb 03, 2020 9:26 am

ResMed doesn't measure event duration for hyponeas...only apneas.
So with SleepyHead/OSCAR you get a minimum 10 second duration for hyponeas because the hyponea has to have lasted at least 10 seconds to earn a flag.

Respironics machine use some sort of marker system and that's what is shown in the parentheses...it's a close approximation of event duration if one were to zoom in and actually count the duration of the event.
Sometimes it's fairly accurate and sometimes it's way off though.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Number in parenthesis in Oscar

Post by LSAT » Mon Feb 03, 2020 10:28 am

Pugsy wrote:
Mon Feb 03, 2020 9:26 am
ResMed doesn't measure event duration for hyponeas...only apneas.
So with SleepyHead/OSCAR you get a minimum 10 second duration for hyponeas because the hyponea has to have lasted at least 10 seconds to earn a flag.

Respironics machine use some sort of marker system and that's what is shown in the parentheses...it's a close approximation of event duration if one were to zoom in and actually count the duration of the event.
Sometimes it's fairly accurate and sometimes it's way off though.
With the old Sleepyhead, hypopneas were always reported in seconds using ResMed S9.

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Number in parenthesis in Oscar

Post by Pugsy » Mon Feb 03, 2020 10:33 am

LSAT wrote:
Mon Feb 03, 2020 10:28 am
With the old Sleepyhead, hypopneas were always reported in seconds using ResMed S9.
Apparently one of the things we lost with the AirSense/AirCurve models was hyponea duration recording.
I get 10 seconds on all my AirCurve 10 hyponeas as well as my AutoSet.

I guess I should modify my statement though.....S9 models gave us event duration for hyponeas....AirSense/AirCurve 10 models do not.
Obviously the software can't show something if the machine doesn't record it.

I have verified with ResScan that all I get is 10 seconds for all hyponeas so it isn't just the software.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Joshua_CPAP
Posts: 1
Joined: Tue Feb 02, 2021 1:10 pm

Re: Number in parenthesis in Oscar

Post by Joshua_CPAP » Thu Apr 29, 2021 4:30 pm

Dog Slobber wrote:
Mon Feb 03, 2020 9:15 am
It means the number of seconds the event lasted.

However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.

*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
Where did you learn that it's not accurate for Hypopneas? Can you cite your sources please.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Number in parenthesis in Oscar

Post by zonker » Thu Apr 29, 2021 7:11 pm

.


deleted by author as what i said was a load of bollocks.
:oops:

thanks, dog, for correcting me in private.
Last edited by zonker on Thu Apr 29, 2021 8:50 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Number in parenthesis in Oscar

Post by Dog Slobber » Thu Apr 29, 2021 7:58 pm

Joshua_CPAP wrote:
Thu Apr 29, 2021 4:30 pm
Dog Slobber wrote:
Mon Feb 03, 2020 9:15 am
It means the number of seconds the event lasted.

However, I understand that it's not accurate on AirSense 10's for Hypopneas and will be removed in a future version.

*EDIT*
My answer was based on ResMed, wasn't aware that Respironics were different.
Where did you learn that it's not accurate for Hypopneas? Can you cite your sources please.
My sources:
  • experience
  • the fact that every Hypopnea time on an Air10 device, that has ever been reported by both SleepyHead and OSCAR has been exactly 10 seconds. So either that 10 seconds is not accurate, or it is the biggest coincidence in the history in the 13.8 billion years of our universe's existence
  • the fact that, as I stated, the timing value for Hypopnea's was removed in OSCAR
  • Hypop_10_sec.jpg
Will you be requiring a larger sample size for the hypopnea's I presented, or possibly a source for the age of the universe?
Battery Backup: EcoFlow Delta 2

User avatar
BlueDragon
Posts: 565
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Number in parenthesis in Oscar

Post by BlueDragon » Thu Apr 29, 2021 8:06 pm

In fact, the "10" you see was not reported by ResMed AirSense 10 CPAP machines. It was arbitrarily added to the number reported (0) by Sleepyhead and early versions of OSCAR. OSCAR no longer does that, so you don't see (10) any more.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Number in parenthesis in Oscar

Post by zonker » Thu Apr 29, 2021 8:53 pm

Dog Slobber wrote:
Thu Apr 29, 2021 7:58 pm

in the history in the 13.8 billion years of our universe's existence
can you cite your sources please.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Number in parenthesis in Oscar

Post by Dog Slobber » Thu Apr 29, 2021 8:59 pm

zonker wrote:
Thu Apr 29, 2021 8:53 pm
Dog Slobber wrote:
Thu Apr 29, 2021 7:58 pm

in the history in the 13.8 billion years of our universe's existence
can you cite your sources please.
Chicago Granny's birth certificate
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Number in parenthesis in Oscar

Post by zonker » Thu Apr 29, 2021 9:50 pm

Dog Slobber wrote:
Thu Apr 29, 2021 8:59 pm
zonker wrote:
Thu Apr 29, 2021 8:53 pm
Dog Slobber wrote:
Thu Apr 29, 2021 7:58 pm

in the history in the 13.8 billion years of our universe's existence
can you cite your sources please.
Chicago Granny's birth certificate
Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Number in parenthesis in Oscar

Post by chunkyfrog » Fri Apr 30, 2021 10:53 am

Papillion, NE?
Are you USAF, or a family member?
If so, thank you for your service.
"They also serve, who stand and wait."

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Number in parenthesis in Oscar

Post by Dog Slobber » Fri Apr 30, 2021 11:17 am

chunkyfrog wrote:
Fri Apr 30, 2021 10:53 am
Papillion, NE?
Are you USAF, or a family member?
If so, thank you for your service.
"They also serve, who stand and wait."
Katerific, the OP hasn't been around since September 2020. Joshua_CPAP necro-bumped a year old, no longer relevant post to scrutinize something he doesn't know anything about.