Getting AHI lower

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Fri Jan 31, 2020 9:53 am

We have a lot of full face mask users complaining of dry mouth and using some sort of oral hydrating product.
Sometimes the machine just can't crank out enough moisture to get the mouth not to be so dry. It simply can't.

You could give it a try though. Some people find that a full face mask and a high humidity setting helps out...there's no guarantee though.
Won't help much if the main culprit is meds or just overall tendency to have dry mouth though.

You could maybe try taping your mouth one night and see if totally eliminating any chance of mouth breathing makes any difference in the dry mouth...and then decide what to do about it once you know exactly the cause.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Getting AHI lower

Post by Miss Emerita » Fri Jan 31, 2020 9:58 am

You might try Xylimelts. They are little discs that stick to your gums or teeth and help to relieve dry mouth by stimulating the production of saliva. They are designed to keep working for hours. I've never tried them, but other people have posted that they can be useful.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Getting AHI lower

Post by palerider » Fri Jan 31, 2020 10:52 am

mgundy wrote:
Fri Jan 31, 2020 5:13 am
I've upped my min pressure to 12 for the last few nights. Still no luck... I'm getting a little worried that this is the best I can expect :(
You've got a long way to go before you get to that point, pressure wise. Don't start worrying yet.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11462
Joined: Fri Jun 19, 2015 4:36 pm

Re: Getting AHI lower

Post by zonker » Fri Jan 31, 2020 12:08 pm

Miss Emerita wrote:
Fri Jan 31, 2020 9:58 am
You might try Xylimelts. They are little discs that stick to your gums or teeth and help to relieve dry mouth by stimulating the production of saliva. They are designed to keep working for hours. I've never tried them, but other people have posted that they can be useful.
meerkat.gif
to the OP: yup. was just about to suggest these. wee discs with a bit of "stickum" on one side to help it adhere to the outside of your teeth. it slowly melts overnight promoting your saliva.

find them in the oral healthcare section at your drugstore. i've seen them at cvs and rite aid. also available at amazon.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Sat Feb 01, 2020 6:59 am

Caught myself breathing out of my mouth. That's what's causing my dry mouth and tongue irritation. That started happening at the higher pressure setting (12+).

I'll try a full face mask. I took a look at some popular ones from ResMed. I'm assuming they can be used with the Dreamstation. In looking at the mask manual I couldn't find a mask setting, x1, x2, etc. Would anyone have insight on that?

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Sat Feb 01, 2020 8:00 am

Yes, you can use any brand mask with the DreamStation.
The only brands that use the mask resistance number are the Respironics brands. We are told by Respironics when using a non Respironics mask to just set it to 0 which effectively turns the Resistance setting off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Feb 03, 2020 3:30 am

I raised the min pressure to 15 and I'm still plateaued at an AHI of around 8. It's weird... every time I raise the min pressure the max pressure during the night increases. This time is was over 18. I guess I should increase it more? What happens if I'm still like this and there is no more increase available?

screenshot-20200203-041636.png
screenshot-20200203-041636.png (857.35 KiB) Viewed 7813 times

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Mon Feb 03, 2020 7:29 am

mgundy wrote:
Mon Feb 03, 2020 3:30 am
What happens if I'm still like this and there is no more increase available?
different model machine that can go higher than 29 cm.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Feb 03, 2020 9:46 am

Any thoughts on the next min pressure setting? 16? 16.5? 17?

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Getting AHI lower

Post by palerider » Mon Feb 03, 2020 9:50 am

mgundy wrote:
Mon Feb 03, 2020 9:46 am
Any thoughts on the next min pressure setting? 16? 16.5? 17?
17

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65359
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Mon Feb 03, 2020 9:52 am

0.5 cm isn't going to do much to change results but it might be needed in terms of comfort adjustments.
I would keep using at least 1 cm increments and not bother with 0.5 unless I had significant problems going to sleep at the higher minimum.

If you can do 17....I would try 17.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4331
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Getting AHI lower

Post by Dog Slobber » Mon Feb 03, 2020 10:14 am

I think we are approaching the point where you might want to consider switching to a BiLevel machine, such as the AirCurve 10 VAuto.
  • You are approaching your machine's maximum of 20 cm, the VAuto has a maximum of 25cm
  • The responsiveness of ResMed's algorithm might do a better job of treating your events.
  • Even if your events get controlled in the 17-20 cm we have left to adjust your minimum to, the better pressure support might increase your comfort level.
Battery Backup: EcoFlow Delta 2

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Feb 03, 2020 10:57 am

Anyone want to buy a barely used dreamstation? :lol: *smh*

I called where I ordered from any there’s no trial periods/returns on home machines. Maybe I’ll keep it as a backup if it comes to that.

User avatar
Dog Slobber
Posts: 4331
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Getting AHI lower

Post by Dog Slobber » Mon Feb 03, 2020 11:06 am

mgundy wrote:
Mon Feb 03, 2020 10:57 am
Anyone want to buy a barely used dreamstation? :lol: *smh*

I called where I ordered from any there’s no trial periods/returns on home machines. Maybe I’ll keep it as a backup if it comes to that.
It's just a thought and might be a bit premature, just yet. There's still some room for adjustment.
Battery Backup: EcoFlow Delta 2

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Feb 03, 2020 11:56 am

Is EPAP related to Flex? I searched around and I couldn't find anything definitive.

I zoomed in on a problem area from last night a few times. I saw that when I did have a bunch of events, what finally brought my breathing back to normal was a pressure of 17 (sometimes a bit higher) and when the EPAP catches up to being between 15.5. and 16.
screenshot-20200203-125111.png
screenshot-20200203-125111.png (872.57 KiB) Viewed 7763 times
screenshot-20200203-125617.png
screenshot-20200203-125617.png (870.23 KiB) Viewed 7763 times