Getting AHI lower

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Getting AHI lower

Post by mgundy » Mon Jan 27, 2020 6:20 am

Hello,

I'm really having difficulty lowering my AHI below 10 with my apap. I currently use a philips respironics dreamstation with a dreamwear mask (w/ nasal cushion) and A-flex of 2. I originally had my minimum pressure set to 8 cmH2O and max at 16 cmH2O. My 90% pressure ranged anywhere from 10 to 13, with most nights right at 11. After reading some of the posts here with the same issue I tried to increase my minimum pressure to 9 cmH2O. This resulted in my AHI increasing to 16... It caused everything to increase (clear airway apneas, obstructive apneas, and hypopneas).

I have Sleepyhead but don't really know how to look at the data. I was hoping some kind soul would take pity on me and offer some suggestions. What charts should I post for analysis?

Thank you for reading!

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: Getting AHI lower

Post by jnk... » Mon Jan 27, 2020 9:52 am

You may find numbers 3 and 4 to be of particular help to you when you click on the following link:

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Getting AHI lower

Post by jimbud » Mon Jan 27, 2020 10:00 am

mgundy wrote:
Mon Jan 27, 2020 6:20 am
Hello,

I'm really having difficulty lowering my AHI below 10 with my apap. I currently use a philips respironics dreamstation with a dreamwear mask (w/ nasal cushion) and A-flex of 2. I originally had my minimum pressure set to 8 cmH2O and max at 16 cmH2O. My 90% pressure ranged anywhere from 10 to 13, with most nights right at 11. After reading some of the posts here with the same issue I tried to increase my minimum pressure to 9 cmH2O. This resulted in my AHI increasing to 16... It caused everything to increase (clear airway apneas, obstructive apneas, and hypopneas).

I have Sleepyhead but don't really know how to look at the data. I was hoping some kind soul would take pity on me and offer some suggestions. What charts should I post for analysis?

Thank you for reading!
Read this:

Sticky: Newbies PLEASE READ BEFORE POSTING
Post by Pugsy » Fri Jul 27, 2018 4:36 am

It is on the Home Page.
It is the first announcement.
Below the big orange Post a New Topic (with a down pointing arrow) sign.

That should get you started.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Jan 27, 2020 1:21 pm

Hi all,

Thank you! I've attached a screen shot from sleepyhead from before and after I raised my min pressure from 8 to 9. Hope someone has a few insights!
screenshot-20200127-141837.png
min pressure 9
screenshot-20200127-141837.png (307.81 KiB) Viewed 15368 times
screenshot-20200127-142114.png
min pressure 8
screenshot-20200127-142114.png (295.49 KiB) Viewed 15368 times

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Getting AHI lower

Post by Jas_williams » Mon Jan 27, 2020 2:10 pm

Your machine is not raising pressure correctly can you make sure you have optistart set to off

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Mon Jan 27, 2020 3:28 pm

I just verified that Optistart was set to ON. I turned it OFF. I will try this setup for a few nights and report back.

Thank you!

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Mon Jan 27, 2020 5:44 pm

Make sure you aren't using EZStart either.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Tue Jan 28, 2020 7:54 am

I verified both optistart and ezstart were off. The AHI was the same, but the pressures certainly changed a lot. Does my pressure plot look normal now at least?
screenshot-20200128-085156.png
screenshot-20200128-085156.png (983.51 KiB) Viewed 15280 times

User avatar
Dog Slobber
Posts: 4332
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Getting AHI lower

Post by Dog Slobber » Tue Jan 28, 2020 8:13 am

Yes, your pressure charts are now representing the range of your min/max settings.

Your minimum pressure is low. As soon your therapy starts, it climbs. Rarely below 10.

I would set it 10, perhaps a little more depending on your comfort level and see how things go for a few days.

Your machine hit your maximum (16) a few times. You may want to just set it to 20, it will only go there if it needs to.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Tue Jan 28, 2020 8:37 am

Did you actually fall asleep that first 90 minutes or so and turn the machine off?
or did you have a lot of trouble falling asleep?

If the answer is yes you were asleep....definitely need more minimum pressure.

Probably at least 10 cm minimum but that's quite a jump to start with from the 6.5 cm you are starting with now.
So if you can't handle that big of a jump then do it in 1 cm increments over a new days.
Some people can handle big jumps rather easily and some can barely handle 1 cm increases. I don't know which one you are. :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Tue Jan 28, 2020 8:55 am

Pugsy wrote:
Tue Jan 28, 2020 8:37 am
Did you actually fall asleep that first 90 minutes or so and turn the machine off?
or did you have a lot of trouble falling asleep?

If the answer is yes you were asleep....definitely need more minimum pressure.

Probably at least 10 cm minimum but that's quite a jump to start with from the 6.5 cm you are starting with now.
So if you can't handle that big of a jump then do it in 1 cm increments over a new days.
Some people can handle big jumps rather easily and some can barely handle 1 cm increases. I don't know which one you are. :lol:
I did fall asleep the first 90 minutes and then woke up. I turn the machine off when I get up to go to the bathroom and that's what I did.

Thank you both! I'll try 8 tonight with the plan to increase to 10 over the next few nights. *fingers crossed*

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Getting AHI lower

Post by palerider » Tue Jan 28, 2020 11:02 am

I'd do a minimum of 12 *tonight* with a max of 20, all you're doing by setting the min to 8 is causing more ineffective treatment as the machine spends the night trying to get back to that min pressure, which you can see is too low.

If you have trouble getting to sleep with that pressure, use the ramp.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Fri Jan 31, 2020 5:13 am

I've upped my min pressure to 12 for the last few nights. Still no luck... I'm getting a little worried that this is the best I can expect :(

A minor thing is I've noticed dry mouth/throat that is causing me to cough later in the evening from the higher pressure (I'm guessing). I have the humidifier all the way up and the heated tube setting at 3. I don't think my mouth is opening during the night.
screenshot-20200131-060550.png
screenshot-20200131-060550.png (887.41 KiB) Viewed 15170 times

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Getting AHI lower

Post by Pugsy » Fri Jan 31, 2020 8:52 am

Try minimum of 14...use ramp if you need to. We don't give up....we keep trying more pressure until you can't go higher with this machine and then we switch machines that will go higher if we have to.

Dry mouth happens more easily to some people than others and we don't know why and it isn't always mouth breathing but of course mouth breathing is a well known culprit.
It doesn't take a lot of mouth breathing either for some people....as little as 10 minutes of mouth breathing (even with a full face mask and humidity maxed out) to develop dry mouth.

The humidifier was never really designed to crank out enough moisture to rehydrate the oral cavity...it was designed for the much smaller nasal cavity.
Maybe someone takes some meds that accelerate the mouth drying or maybe they have other health issues that might accelerate mouth drying or maybe they are just unlucky.

Dry mouth even with a full face mask and humidifier set to max....still happens a lot and is a common complaint.
Those people add some sort of mouth hydrating agent at bedtime...to help keep the mouth moister.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mgundy
Posts: 47
Joined: Mon Jan 27, 2020 6:07 am

Re: Getting AHI lower

Post by mgundy » Fri Jan 31, 2020 9:49 am

Will do! I’ll give 14 a shot. I’ve read a full face mask can help with dry mouth. What’s your opinion on that?