Reducing rebreathing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Reducing rebreathing

Post by littorine » Sun Jan 19, 2020 2:00 pm

Newbie here, looking for recommendations for types of masks and/or tubing to minimize rebreathing as much as possible. I know all masks have sufficient venting to be safe, but I'm very sensitive to elevated CO2 levels (can't go in malls, airplanes, or many office buildings because the CO2 is too high for me; and it takes hours for the cognitive effects to wear off if the CO2 is too high), so even mild, generally safe levels of rebreathing are problematic for me. I'm currently using a ResMed Mirage FX mask with pinhole vents and a Respironics REMstar Auto A-Flex System One machine, but I've seen that some masks and tubes are advertised with valves to reduce rebreathing. I'd appreciate any recommendations for equipment that minimizes rebreathing. Thanks!

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Reducing rebreathing

Post by raisedfist » Sun Jan 19, 2020 3:19 pm

Any mask with a smaller surface area, such as a nasal pillows mask, has less dead space - therefore less chance of re breathing. If you're a mouth breather you can always wear a nasal mask and tape your mouth shut. Many people here do just that.

If you're using the APAP mode, increasing the min pressure will also increase the amount of expired air being vented. Each mask has the vent rates in relation to the pressure set in the manual.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Reducing rebreathing

Post by Wulfman... » Sun Jan 19, 2020 4:28 pm

Read and then fill out your profile.......including all your settings.

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Janknitz
Posts: 8511
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Reducing rebreathing

Post by Janknitz » Sun Jan 19, 2020 5:27 pm

I think you can look at the specs of any mask and see what the vent rate is. That might help in your decision.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
ragtopcircus
Posts: 220
Joined: Mon Sep 09, 2019 11:55 am

Re: Reducing rebreathing

Post by ragtopcircus » Mon Jan 20, 2020 10:15 am

littorine wrote:
Sun Jan 19, 2020 2:00 pm
I've seen that some masks and tubes are advertised with valves to reduce rebreathing. I'd appreciate any recommendations for equipment that minimizes rebreathing. Thanks!
If you are referring to anti-asphyxiation valves, they are not intended to reduce rebreathing. The valve remains closed in normal operation. It opens if the machine shuts off so you can get enough air. They are typically found only on full face masks, because if the machine shuts off (for example, in a power failure) when you are are wearing a pillow mask, you can just open your mouth.

Some people have the opposite problem: low CO2 levels can cause Central Apnea, because CO2 is a breathing trigger. This can be induced if the difference between inhalation and exhalation pressure is too high. Unfortunately, the options are limited with your machine, but it might be worthwhile to at least set the Flex to 3 if you have not already.

Do you use supplemental oxygen?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: BiPAP VAuto 10-25 PS 4.4, and wherever my tempestuous turbinates take me ....
Last edited by ragtopcircus on Mon Jan 20, 2020 10:53 am, edited 1 time in total.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Reducing rebreathing

Post by chunkyfrog » Mon Jan 20, 2020 10:22 am

littorine wrote:
Sun Jan 19, 2020 2:00 pm
. . .
I'm very sensitive to elevated CO2 levels
Has this sensitivity been medically confirmed?
If so, then you should be referred to an appropriate specialist for treatment.
The source could be physical or psychological, but not treatable online.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Miss Emerita
Posts: 3752
Joined: Sun Nov 04, 2018 8:07 pm

Re: Reducing rebreathing

Post by Miss Emerita » Mon Jan 20, 2020 10:44 am

chunkyfrog wrote:
Mon Jan 20, 2020 10:22 am
littorine wrote:
Sun Jan 19, 2020 2:00 pm
. . .
I'm very sensitive to elevated CO2 levels
Has this sensitivity been medically confirmed?
If so, then you should be referred to an appropriate specialist for treatment.
The source could be physical or psychological, but not treatable online.
+1
Oscar software is available at https://www.sleepfiles.com/OSCAR/

littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Re: Reducing rebreathing

Post by littorine » Mon Jan 20, 2020 1:02 pm

chunkyfrog wrote:
Mon Jan 20, 2020 10:22 am
Has this sensitivity been medically confirmed?
If so, then you should be referred to an appropriate specialist for treatment.
The source could be physical or psychological, but not treatable online.
Yes, it has been confirmed medically, and the treatment is to avoid high CO2 environments. I'm not looking online for treatment, just advice about CPAP masks. Thanks.

littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Re: Reducing rebreathing

Post by littorine » Mon Jan 20, 2020 1:08 pm

ragtopcircus wrote:
Mon Jan 20, 2020 10:15 am
If you are referring to anti-asphyxiation valves, they are not intended to reduce rebreathing. The valve remains closed in normal operation. It opens if the machine shuts off so you can get enough air.
Thanks, that's very helpful information.
ragtopcircus wrote:
Mon Jan 20, 2020 10:15 am
Unfortunately, the options are limited with your machine, but it might be worthwhile to at least set the Flex to 3 if you have not already.
Do you use supplemental oxygen?
Thanks, I'll try that! No supplemental oxygen, since I don't have any problems in that regard.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing rebreathing

Post by palerider » Mon Jan 20, 2020 1:14 pm

littorine wrote:
Mon Jan 20, 2020 1:02 pm
chunkyfrog wrote:
Mon Jan 20, 2020 10:22 am
Has this sensitivity been medically confirmed?
If so, then you should be referred to an appropriate specialist for treatment.
The source could be physical or psychological, but not treatable online.
Yes, it has been confirmed medically, and the treatment is to avoid high CO2 environments. I'm not looking online for treatment, just advice about CPAP masks. Thanks.
I've never encountered a CPAP mask that has any significant amount of CO2 rebreathing.

It's very easy to tell, as I'm breathing, I can feel the vent air change temperature as it's purging my exhaled breath and goes cool again before the next inhalation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Re: Reducing rebreathing

Post by littorine » Mon Jan 20, 2020 2:40 pm

palerider wrote:
Mon Jan 20, 2020 1:14 pm
I've never encountered a CPAP mask that has any significant amount of CO2 rebreathing.

It's very easy to tell, as I'm breathing, I can feel the vent air change temperature as it's purging my exhaled breath and goes cool again before the next inhalation.
That doesn't mean all the CO2 is gone, but that isn't a problem for most people since they can tolerate some CO2 rebreathing. I've checked the CO2 outside the vents of my mask using a CO2 monitor and can confirm that it's well higher than ambient air (about 1400 ppm in a room that's kept below 550 ppm, compared to about 1900 ppm when holding the monitor at the same distance without the mask), but the monitor also shows higher CO2 inside the mask just prior to inspiration. It's impossible to get a good reading inside the mask because of having to stick the monitor underneath it, which causes a big air leak and diffusion of the air inside the mask and therefore artificially low levels, but I've had readings from 750-1500 ppm. Again, this is NOT an issue for the average person - I've been in hotel rooms with CO2 levels of 1500 ppm, and no one thinks twice about sleeping in them, except those of us who carry around CO2 monitors! ;)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing rebreathing

Post by palerider » Mon Jan 20, 2020 3:57 pm

littorine wrote:
Mon Jan 20, 2020 2:40 pm
palerider wrote:
Mon Jan 20, 2020 1:14 pm
I've never encountered a CPAP mask that has any significant amount of CO2 rebreathing.

It's very easy to tell, as I'm breathing, I can feel the vent air change temperature as it's purging my exhaled breath and goes cool again before the next inhalation.
That doesn't mean all the CO2 is gone, but that isn't a problem for most people since they can tolerate some CO2 rebreathing. I've checked the CO2 outside the vents of my mask using a CO2 monitor and can confirm that it's well higher than ambient air (about 1400 ppm in a room that's kept below 550 ppm, compared to about 1900 ppm when holding the monitor at the same distance without the mask), At what point in the respiration cycle? the co2 concentration varies wildly.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Re: Reducing rebreathing

Post by littorine » Mon Jan 20, 2020 4:09 pm

palerider wrote:
Mon Jan 20, 2020 3:57 pm
At what point in the respiration cycle? the co2 concentration varies wildly.
The inside mask readings? I stick the meter in just before I would naturally breathe in (but holding my breath). My CO2 monitor is designed to reflect ambient CO2, and doesn't respond quickly enough to record the ups and downs of a respiratory cycle, so the measurements outside the vents and at an equivalent distance from my mouth reflect the maximum it's exposed to, rather than any particular instant.

littorine
Posts: 6
Joined: Sun Jan 19, 2020 11:49 am

Re: Reducing rebreathing

Post by littorine » Tue Jan 21, 2020 12:38 pm

ragtopcircus wrote:
Mon Jan 20, 2020 10:15 am
It might be worthwhile to at least set the Flex to 3 if you have not already.
Thank you! I tried that last night and felt much clearer and energetic this morning.:)

User avatar
zonker
Posts: 11333
Joined: Fri Jun 19, 2015 4:36 pm

Re: Reducing rebreathing

Post by zonker » Tue Jan 21, 2020 1:11 pm

littorine wrote:
Tue Jan 21, 2020 12:38 pm
ragtopcircus wrote:
Mon Jan 20, 2020 10:15 am
It might be worthwhile to at least set the Flex to 3 if you have not already.
Thank you! I tried that last night and felt much clearer and energetic this morning.:)