Leak Rate

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 9:01 am

I have two identical Air Curve 10 vauto machines. One I use nightly at home. The other is my back up and gets used when I travel. Settings are identical. I have always noticed the leak rate on my backup machine is lower. I usually use it when I'm in vegas at my daughters house visiting. I have always figured it was lower because I was sleeping without hubby in a different bed.

This weekend we were finally able to take a trip together (first one since I've had the machines) and I noticed the same phenomenon. Just going by the leak rate reported on the machine ( I haven't imported data yet) it is significantly lower.

Same mask, same hose, the only difference is the machine. My home machine reports leaks usually at just under 20 L/min. My backup machine reports leaks at under 5L/min.

Any WAG as to why this might be ?

User avatar
Dog Slobber
Posts: 4331
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Leak Rate

Post by Dog Slobber » Sun Jan 19, 2020 9:46 am

Are both machines set to the same mask type?

Try swapping the hose and mask between the two and see if the mask leak moves with the mask change.

When you look at the OSCAR leak trace, do you see considerable leaks? If you do that suggests a mask fit issue on the leaky mask.
Battery Backup: EcoFlow Delta 2

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Leak Rate

Post by jimbud » Sun Jan 19, 2020 10:32 am

I'm just throwing this out there.

Have you changed out humidifier tanks?
Maybe one is leaking just enough to make a difference?

Same hose and same mask. Not the same tank. Process of elimination.

Just a thought. :)

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 11:03 am

Dog Slobber wrote:
Sun Jan 19, 2020 9:46 am
Are both machines set to the same mask type?

Try swapping the hose and mask between the two and see if the mask leak moves with the mask change.

When you look at the OSCAR leak trace, do you see considerable leaks? If you do that suggests a mask fit issue on the leaky mask.
Settings are identical. It is the same mask and hose I always use. I take it off my home machine and bring with me.

Trying to pull up sleepyhead data now. I haven't tried Oscar yet.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 11:29 am

jimbud wrote:
Sun Jan 19, 2020 10:32 am
I'm just throwing this out there.

Have you changed out humidifier tanks?
Maybe one is leaking just enough to make a difference?

Same hose and same mask. Not the same tank. Process of elimination.

Just a thought. :)

JPB
Good idea... I will try the back up tank tonight. :)

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 11:35 am

Night out of town:
screenshot-20200119-101746.png
Normal night at home:
screenshot-20200119-101118.png

User avatar
Pugsy
Posts: 65356
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leak Rate

Post by Pugsy » Sun Jan 19, 2020 1:25 pm

Do a 30 minute test on each machine while awake and you are darn sure there are zero leaks and then compare the results.

Use SleepyHead to view the results because of that little bug in OSCAR not letting a person download more than one session.

The difference might be nothing more than normal variation from some normal mask movement.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leak Rate

Post by palerider » Sun Jan 19, 2020 1:51 pm

babydinosnoreless wrote:
Sun Jan 19, 2020 9:01 am
I have two identical Air Curve 10 vauto machines. One I use nightly at home. The other is my back up and gets used when I travel. Settings are identical. I have always noticed the leak rate on my backup machine is lower. I usually use it when I'm in vegas at my daughters house visiting. I have always figured it was lower because I was sleeping without hubby in a different bed.

This weekend we were finally able to take a trip together (first one since I've had the machines) and I noticed the same phenomenon. Just going by the leak rate reported on the machine ( I haven't imported data yet) it is significantly lower.

Same mask, same hose, the only difference is the machine. My home machine reports leaks usually at just under 20 L/min. My backup machine reports leaks at under 5L/min.

Any WAG as to why this might be ?
Swap the machines only when at home.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leak Rate

Post by palerider » Sun Jan 19, 2020 1:53 pm

babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Dog Slobber wrote:
Sun Jan 19, 2020 9:46 am
Are both machines set to the same mask type?

Try swapping the hose and mask between the two and see if the mask leak moves with the mask change.

When you look at the OSCAR leak trace, do you see considerable leaks? If you do that suggests a mask fit issue on the leaky mask.
Settings are identical. It is the same mask and hose I always use. I take it off my home machine and bring with me.
I have a separate hose and mask in the bag with my spare machine, so all that's needed is to grab the bag and go.
I suggest you should do the same.That way you can't forget something when packing, and if something happens while away, you've got your setup at home, waiting for when you get back.
babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Trying to pull up sleepyhead data now. I haven't tried Oscar yet.
*taps foot*.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 2:22 pm

palerider wrote:
Sun Jan 19, 2020 1:53 pm
babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Dog Slobber wrote:
Sun Jan 19, 2020 9:46 am
Are both machines set to the same mask type?

Try swapping the hose and mask between the two and see if the mask leak moves with the mask change.

When you look at the OSCAR leak trace, do you see considerable leaks? If you do that suggests a mask fit issue on the leaky mask.
Settings are identical. It is the same mask and hose I always use. I take it off my home machine and bring with me.
I have a separate hose and mask in the bag with my spare machine, so all that's needed is to grab the bag and go.
I suggest you should do the same.That way you can't forget something when packing, and if something happens while away, you've got your setup at home, waiting for when you get back.
Oh I keep a spare mask and hose in the bag. I just wanted the one I have adjusted to fit and with the cozy on it so I swapped it before I left.
babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Trying to pull up sleepyhead data now. I haven't tried Oscar yet.
palerider wrote:
Sun Jan 19, 2020 1:53 pm

*taps foot*.
Whines...but I don't want to learn new software. :lol: :lol:
I want something that will work on my ipad. ;)

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Sun Jan 19, 2020 2:27 pm

Pugsy wrote:
Sun Jan 19, 2020 1:25 pm
Do a 30 minute test on each machine while awake and you are darn sure there are zero leaks and then compare the results.

Use SleepyHead to view the results because of that little bug in OSCAR not letting a person download more than one session.

The difference might be nothing more than normal variation from some normal mask movement.
Thanks ! Good idea.

harrywr2
Posts: 72
Joined: Sun Nov 10, 2019 1:01 pm

Re: Leak Rate

Post by harrywr2 » Sun Jan 19, 2020 2:50 pm

The leaks seem to share similar curves and both machines register zero leaks.
DIfferent bed pillow firmness?

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leak Rate

Post by palerider » Sun Jan 19, 2020 5:20 pm

babydinosnoreless wrote:
Sun Jan 19, 2020 2:22 pm
palerider wrote:
Sun Jan 19, 2020 1:53 pm
babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Trying to pull up sleepyhead data now. I haven't tried Oscar yet.
*taps foot*.
Whines...but I don't want to learn new software. :lol: :lol:
I want something that will work on my ipad. ;)
Sleepyhead is dead. no more, gone, kaput, murdered.

There's very little to learn with Oscar, it's just better. and getting into it now will make it easier in the future, when new versions come out.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Mon Jan 20, 2020 11:44 am

harrywr2 wrote:
Sun Jan 19, 2020 2:50 pm
The leaks seem to share similar curves and both machines register zero leaks.
DIfferent bed pillow firmness?
Actually brought my pillow with me. Lol. I don't do well with strange pillows. :)

I was going to try changing out the humidifier last night and do the awake test that Pugsy suggested but seem to have picked up a cold so I will probably wait for a couple of days before trying experiments.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Leak Rate

Post by babydinosnoreless » Mon Jan 20, 2020 11:53 am

palerider wrote:
Sun Jan 19, 2020 5:20 pm
babydinosnoreless wrote:
Sun Jan 19, 2020 2:22 pm
palerider wrote:
Sun Jan 19, 2020 1:53 pm
babydinosnoreless wrote:
Sun Jan 19, 2020 11:03 am
Trying to pull up sleepyhead data now. I haven't tried Oscar yet.
*taps foot*.
Whines...but I don't want to learn new software. :lol: :lol:
I want something that will work on my ipad. ;)
Sleepyhead is dead. no more, gone, kaput, murdered.

There's very little to learn with Oscar, it's just better. and getting into it now will make it easier in the future, when new versions come out.
I know. I will get to Oscar eventually. :) I have an ancient old laptop with a built in card reader My husband set up with sleepyhead for me. The laptop and I have a mutual dislike of each other. Just trying to pull up sleepyhead yesterday I was getting the boot menu and a black screen of death. Got it working eventually but it is slow and cranky. Kind of like me first thing in the morning. :lol: :lol: :lol: