pressures too high,need imput

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: pressures too high,need imput

Post by JDS74 » Sat Jan 18, 2020 7:26 am

It would be helpful if you could post your machine settings.
In particular, the setting for RAMP.
New users often have difficulty adjusting to their machines, especially if the pressures are as high as yours are.
Setting the RAMP function to ON, the initial pressure to something like 7 or 8 cmH2O, and the time to 30 minutes or so can help in this transition.
What this does is to start you out with a lower pressure and gradually increase it to your therapeutic pressure over the time interval.
This will give you a chance to get to sleep before the higher pressures kick in.

A couple of other things:
1) Ask for a copy of your sleep study report. It has all kinds of interesting data in it that will be helpful.
2) Discuss this issue with your sleep doctor - this is especially important for you since the DOT rules re CPAP apply.
3) Ask for a copy of your CPAP prescription and keep a copy with you. If something happens to your machine on the road it will be necessary to have when you go to a local DME where you are. Every large city will have a DME so you can get accessory parts wherever you are.
4) Keep your sleep doctor informed.

Hope everything works out and this issue gets resolved.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
ChicagoGranny
Posts: 15509
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: pressures too high,need imput

Post by ChicagoGranny » Sat Jan 18, 2020 8:43 am

He ran for the hills!!???

User avatar
zonker
Posts: 11458
Joined: Fri Jun 19, 2015 4:36 pm

Re: pressures too high,need imput

Post by zonker » Sat Jan 18, 2020 12:45 pm

pugsy has done some editing here because OP had posted more than one post.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg