CPAP-induced central sleep apnea?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
kevinpz
Posts: 1
Joined: Thu Jan 16, 2020 8:49 pm

CPAP-induced central sleep apnea?

Post by kevinpz » Thu Jan 16, 2020 8:59 pm

Hi!

First, thank you so much for this website, it's a really amazing source for people who just discovered they have sleep apnea.

I've been diagnosed in October 2019 with mild sleep apnea, here's the report (Philips Alice NightOne):
  • AHI: 8.4 events / hour -> 0 central
  • Average SpO2: 95%
  • Lowest SpO2: 91%
So, I've been using my CPAP since mid October (ResMed AirSense 10 AutoSet for Her, AirFit P30i).
I'm starting to sleep a little better in my opinion, but I'm seeing since the beginning of the CPAP some central apneas. The AHI is considered as treated (less than 5), but I'm wondering what causes them, and how could I lower them. (they are all around 15-30 seconds long, and all seem the same zooming in OSCAR)

In December I tried to increase the min pressure to 7, but seems like it was the same.

Do you have any advice, or settings I can change to try to lower them? Is it real central sleep apnea, or hidden OSA, or I'm just awake a lot in the night (even if I don't remember)?

I'm including 2 nights (one with min at 7, the other with min at 6, with 1 zoom of the CA).
Screen Shot 2020-01-16 at 21.56.58.png
Screen Shot 2020-01-16 at 21.56.58.png (490.35 KiB) Viewed 1771 times
Screen Shot 2020-01-16 at 21.38.09.png
Screen Shot 2020-01-16 at 21.38.09.png (507.85 KiB) Viewed 1771 times
Screen Shot 2020-01-16 at 19.33.17.png
Screen Shot 2020-01-16 at 19.33.17.png (452.13 KiB) Viewed 1771 times
Thanks!

SleepyPaolo
Posts: 57
Joined: Sat Nov 23, 2019 1:08 am

Re: CPAP-induced central sleep apnea?

Post by SleepyPaolo » Fri Jan 17, 2020 4:30 pm

I notice you've got EPR set to 3.
I've been told on here that this pressure differential can cause CAs.
It might be worth trying EPR on 1 (or off) for 3 or 4 nights to see if they reduce consistently. It's worked for me, I've got a thread on here with a similar experiment (same machine too) and it worked.

The CA close up you've should appears to be when you've woken briefly. From what I've been reading here the erratic pattern just before it happens is normally awake or changing position and so that CA is not worth considering.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: CPAP-induced central sleep apnea?

Post by LSAT » Fri Jan 17, 2020 4:55 pm

You really can't make a significant decrease in CAs with pressure change. I agree that your events look like slight wake ups or restless turning.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: CPAP-induced central sleep apnea?

Post by JDS74 » Sat Jan 18, 2020 7:10 am

CA events on a CPAP machine are reported differently than on a PSG report. The in-lab machine (Alice) sees them just the same but the person scoring the data takes into account the arousal state of the patient.
All events that occur related to an arousal are not reported.

So … if, as suggested, the CA's you are showing on the OSCAR report is related to an arousal (and it seems so to me also) the in-lab study would not have reported it and, unless the time duration gets to be longer, they can be safely ignored.

However, when you see you sleep doctor next, do mention them to him and show him the same detail that you have posted - he will have his own opinion as to their significance.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP-induced central sleep apnea?

Post by Pugsy » Sat Jan 18, 2020 7:28 am

http://freecpapadvice.com/sleepyhead-free-software

Watch the videos at the above link.
Awake flagged events don't count.

Try turning EPR off and see if the number of centrals reduce or not....at least the number that can't be blamed on awake breathing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.