extra noise from Air set 10 Autoset

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
flightco
Posts: 540
Joined: Mon Aug 03, 2015 6:58 pm
Location: Tinley Park, IL

extra noise from Air set 10 Autoset

Post by flightco » Thu Jan 16, 2020 1:46 pm

I have been using my Cpap for about 5 or 6 years now and have never had a problem. Over the past month it is making a louder noise in the area where it draws in air, enough that it wakes my wife up (she wakes up from the blue light on the cable box) I have taken the filter out and it does not change the noise so not sure if it is inside the unit or sucking air in through a leak.

Wondering if anyone else has experienced this and might have any ideas. Time to get a new one?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: Also Sleepyhead, P10, F10 and Amara View

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: extra noise from Air set 10 Autoset

Post by squid13 » Thu Jan 16, 2020 2:11 pm

Do you use an ozone cleaner ( sanitizer) on your machine?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Dog Slobber
Posts: 4320
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: extra noise from Air set 10 Autoset

Post by Dog Slobber » Thu Jan 16, 2020 2:40 pm

Try reseating the humidifier, or swap it out with your spare humidifier. You do have a spare, don't you? :)

Reseat the hoses. Look for a leak on the hose. Swap the hose out with your spare hose.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: extra noise from Air set 10 Autoset

Post by palerider » Thu Jan 16, 2020 4:18 pm

flightco wrote:
Thu Jan 16, 2020 1:46 pm
I have been using my Cpap for about 5 or 6 years now and have never had a problem. Over the past month it is making a louder noise in the area where it draws in air, enough that it wakes my wife up (she wakes up from the blue light on the cable box) I have taken the filter out and it does not change the noise so not sure if it is inside the unit or sucking air in through a leak.

Wondering if anyone else has experienced this and might have any ideas. Time to get a new one?
Two things, 1), what the slobbering dog said, and 2) https://www.lightdims.com/index.php

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: extra noise from Air set 10 Autoset

Post by jimbud » Thu Jan 16, 2020 5:47 pm

palerider wrote:
Thu Jan 16, 2020 4:18 pm
flightco wrote:
Thu Jan 16, 2020 1:46 pm
wakes my wife up (she wakes up from the blue light on the cable box)

Wondering if anyone else has experienced this and might have any ideas. Time to get a new one?
Two things, 1), what the slobbering dog said, and 2) https://www.lightdims.com/index.php

Wife? No. Not for that, anyway.



I use blue painters tape. Shines through. Dimmer the more layers.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
zonker
Posts: 11431
Joined: Fri Jun 19, 2015 4:36 pm

Re: extra noise from Air set 10 Autoset

Post by zonker » Thu Jan 16, 2020 8:04 pm

palerider wrote:
Thu Jan 16, 2020 4:18 pm

.... and 2) https://www.lightdims.com/index.php
that's hilarious!!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: extra noise from Air set 10 Autoset

Post by palerider » Thu Jan 16, 2020 9:39 pm

zonker wrote:
Thu Jan 16, 2020 8:04 pm
palerider wrote:
Thu Jan 16, 2020 4:18 pm

.... and 2) https://www.lightdims.com/index.php
that's hilarious!!!
I found the lightdims years ago... some things that used to light up the room? now, just a dim glow that I can see when I look for it... I don't know WHY companies put these crazy bright LEDs on things.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11431
Joined: Fri Jun 19, 2015 4:36 pm

Re: extra noise from Air set 10 Autoset

Post by zonker » Thu Jan 16, 2020 10:59 pm

palerider wrote:
Thu Jan 16, 2020 9:39 pm
zonker wrote:
Thu Jan 16, 2020 8:04 pm
palerider wrote:
Thu Jan 16, 2020 4:18 pm

.... and 2) https://www.lightdims.com/index.php
that's hilarious!!!
I found the lightdims years ago... some things that used to light up the room? now, just a dim glow that I can see when I look for it... I don't know WHY companies put these crazy bright LEDs on things.
the same reason that, in the 90's, they put a frickin' clock in EVERYTHING!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
flightco
Posts: 540
Joined: Mon Aug 03, 2015 6:58 pm
Location: Tinley Park, IL

Re: extra noise from Air set 10 Autoset

Post by flightco » Mon Jan 20, 2020 9:13 pm

squid13 wrote:
Thu Jan 16, 2020 2:11 pm
Do you use an ozone cleaner ( sanitizer) on your machine?
That is it, must have forgotten to turn it off overnight. You do know how opposed I am to these don't you?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: Also Sleepyhead, P10, F10 and Amara View

User avatar
flightco
Posts: 540
Joined: Mon Aug 03, 2015 6:58 pm
Location: Tinley Park, IL

Re: extra noise from Air set 10 Autoset

Post by flightco » Mon Jan 20, 2020 9:15 pm

Dog Slobber wrote:
Thu Jan 16, 2020 2:40 pm
Try reseating the humidifier, or swap it out with your spare humidifier. You do have a spare, don't you? :)

Reseat the hoses. Look for a leak on the hose. Swap the hose out with your spare hose.
Sorry, I have been away for a few days but the problem was the humidifier (even though I would have sworn the noise was coming from the area of the filter). I do have a spare but don't really use my humidifier. I have a plug (for lack of better wording) I put in when I travel because I would never use hotel water and i placed that in where the humidifier goes and end of problem. You were right on.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: Also Sleepyhead, P10, F10 and Amara View

User avatar
flightco
Posts: 540
Joined: Mon Aug 03, 2015 6:58 pm
Location: Tinley Park, IL

Re: extra noise from Air set 10 Autoset

Post by flightco » Mon Jan 20, 2020 9:16 pm

palerider wrote:
Thu Jan 16, 2020 4:18 pm
flightco wrote:
Thu Jan 16, 2020 1:46 pm
I have been using my Cpap for about 5 or 6 years now and have never had a problem. Over the past month it is making a louder noise in the area where it draws in air, enough that it wakes my wife up (she wakes up from the blue light on the cable box) I have taken the filter out and it does not change the noise so not sure if it is inside the unit or sucking air in through a leak.

Wondering if anyone else has experienced this and might have any ideas. Time to get a new one?
Two things, 1), what the slobbering dog said, and 2) https://www.lightdims.com/index.php
Light dims, great idea that I wish I had thought of

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: Also Sleepyhead, P10, F10 and Amara View

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: extra noise from Air set 10 Autoset

Post by palerider » Mon Jan 20, 2020 10:50 pm

flightco wrote:
Mon Jan 20, 2020 9:13 pm
squid13 wrote:
Thu Jan 16, 2020 2:11 pm
Do you use an ozone cleaner ( sanitizer) on your machine?
That is it, must have forgotten to turn it off overnight. You do know how opposed I am to these don't you?
He might have missed your most excellent tirades against them ;)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.