Supplemental O2 ?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Supplemental O2 ?

Post by russjpaul » Fri Jan 10, 2020 5:52 pm

Could anyone give me a couple of pointers interpreting my Sleepyhead output. I live at 7200 feet and while my o2 stats are good at sea level, I'm noticing significant drops in spo2 while at home. Could anyone tell me how to arrange my Sleepyhead report so one of you knowledgeable folks might be able to give me some advice.

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Wed Jan 15, 2020 9:00 am

Sorry for not reading the introductory instructions. :( .

Attached is Sleepyhead with Spo2 reading. do you folks think a small amount of supplemental o2 will help ? I'm 56, 5'11 & 190 lbs and live in Park City, Utah. 7200 feet elevation.
Attachments
Screen Shot 2020-01-15 at 7.54.21 AM.png
Screen Shot 2020-01-15 at 7.54.21 AM.png (290.75 KiB) Viewed 8171 times

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Wed Jan 15, 2020 9:26 am

Have you tried raising your min pressure a bit to see if that helps? EPAP/CPAP helps keeps the airways open so that proper gas exchange can occur.

What kind of pulse oximeter do you use to record? It may be helpful to use their software to upload the screenshots on here because it will show the % of movement artifact and a bit more data than importing into SleepyHead/OSCAR will.

It does appear from the rough chart that you do have a few sustained clusters of desaturations below 88%, leading me to guess that they aren't artifacts, but it's just a guess. I don't see anything in your data that screams an obvious culprit.

The general rule/advice is to optimize PAP therapy to the best ability first, and then if the issue persists, consider taking further steps.

My amateur plan would be to raise your min pressure until the remaining events are knocked away, including any snoring and flow limitations. Get those charts looking beautiful, and then test the O2 levels overnight.

Do wait until smarter people than I comment :lol:
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Wed Jan 15, 2020 4:06 pm

Hi raisedfist !

Thanks so much for your insight.

I'm going to increase the low pressure setting by a couple of inches tonight and run the pulse ox too and see what we get tomorrow.

The pulse-ox I have is a CMS50FW and it uses Spo2 Assistant.

Attached is the graphic summary and oximetry it generated. If you can use a different report let me know and see if it will generate what you need.

It didnt dawn on me that the issue occurs at low pressure.

Thanks again!

Russ
Attachments
russspo2graphicsum.png
russspo2graphicsum.png (22.3 KiB) Viewed 8137 times
russspo2oximetry011520.png
russspo2oximetry011520.png (37.68 KiB) Viewed 8137 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Wed Jan 15, 2020 7:29 pm

russjpaul wrote: ↑
Wed Jan 15, 2020 9:00 am
Sorry for not reading the introductory instructions. :( .

Attached is Sleepyhead with Spo2 reading. do you folks think a small amount of supplemental o2 will help ? I'm 56, 5'11 & 190 lbs and live in Park City, Utah. 7200 feet elevation.
Please read the instructions on how to format charts. the AHI chart is *completely* useless.
http://cpaptalk.com/wiki/index.php/Oscar:organize

your SpO2 isn't bad for most of the night, a little more pressure would probably take care of it.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Thu Jan 16, 2020 9:14 pm

Basal SpO2 of 90ish is low but perhaps understandable due to the altitude. I think it’s similar to a pressurized cabin. Not harmful in itself but I would indeed try to improve that with EPAP. Below 88% can be harmful if repetitive or sustained but 12 mins below 88% should be able to be cleaned up with more pressure.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Fri Jan 17, 2020 1:08 pm

Increased Min Press.to 10.6. AHI index went down but spo2 actually dropped a tiny bit. Maybe increase min pressure to 11.5 ?

Thanks folks - appreciate the help.

R.
Attachments
OximetryGraphic011620.png
OximetryGraphic011620.png (22.48 KiB) Viewed 8071 times
Oximetry011620.png
Oximetry011620.png (38.04 KiB) Viewed 8071 times
OSCAR011620.png
OSCAR011620.png (112.16 KiB) Viewed 8071 times

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Fri Jan 17, 2020 3:03 pm

Hmm well your respiratory statistics look more or less normal...in terms of respiratory rate, minute ventilation and tidal volume. An adequate amount of air seems to be coming in and going out. Tidal volume *could be* a bit low depending on your height. No periodic breathing either from the altitude. I don't see any glaring issues. You could perhaps increase EPR to 3 full time just to see what effect it has if anything, to see what a bit of extra ventilation does, but I don't think that will yield much going from 2 to 3.

What are your oxygen levels while awake at altitude? Have you brought this issue up to your doctor?

Do you have an established life where it is necessary to stay in the town you live in? If there were nearby towns at lower elevation, and it wouldn't change your life/routine much, maybe consider moving depending on if you have any symptoms related to the lower oxygen levels.

An oxygen concentrator isn't a big deal to have and connect to a PAP device, but it does of course cost money, increase noise levels and electricity costs of course :mrgreen:
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Fri Jan 17, 2020 5:16 pm

I'll adjust the Min up to 11 and change the EPR to 3 just out of interest - watch it for a couple of days and see....my headgear was in a terrible state so I replaced that and the seal cushion too so leakage should go down.

My o2 stats are normal when up and about; 96 - 98 most of the time....admittedly I've never worn the PulseOx during the day and let it record.

I have a very established life in a great little ski town, we own and run a bed & breakfast so moving into the pollution of Salt Lake City wouldn't be my first choice.

I found a concentrator for $350 on Amazon, 1 liter/min at 93% o2, 2l/min at 73% and 5l/min at 35%. I've already got a climateline with an o2 port on it so its a breeze to connect the CPAP.

see how we do without o2......

R.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Fri Jan 17, 2020 5:50 pm

russjpaul wrote: ↑
Fri Jan 17, 2020 1:08 pm
Increased Min Press.to 10.6. AHI index went down but spo2 actually dropped a tiny bit. Maybe increase min pressure to 11.5 ?

Thanks folks - appreciate the help.

R.
I'd say min of 12.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Fri Jan 17, 2020 6:34 pm

No need to record o2 during the day unless you had obvious symptoms while awake.

More min pressure like Palerider said is worth a shot. It won't cause any harm.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Thumper68
Posts: 129
Joined: Sun Sep 16, 2018 9:48 am

Re: Supplemental O2 ?

Post by Thumper68 » Fri Jan 17, 2020 6:41 pm

Why are you restricting your max pressure to 14.8? I think that your machine wants to increase above that at times to fight something.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Supplemental O2 ?

Post by zonker » Fri Jan 17, 2020 7:13 pm

Thumper68 wrote: ↑
Fri Jan 17, 2020 6:41 pm
Why are you restricting your max pressure to 14.8? I think that your machine wants to increase above that at times to fight something.
i don't see it trying to raise above 14.8 on the charts he's shown.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Thumper68
Posts: 129
Joined: Sun Sep 16, 2018 9:48 am

Re: Supplemental O2 ?

Post by Thumper68 » Sat Jan 18, 2020 8:12 am

zonker wrote: ↑
Fri Jan 17, 2020 7:13 pm
Thumper68 wrote: ↑
Fri Jan 17, 2020 6:41 pm
Why are you restricting your max pressure to 14.8? I think that your machine wants to increase above that at times to fight something.
i don't see it trying to raise above 14.8 on the charts he's shown.
On 1-9 it went to 14.6 and on the next one 1-15 it did max at 14.8. He may have a valid reason to limit it but I prefer to set the max to max!

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Supplemental O2 ?

Post by zonker » Sat Jan 18, 2020 12:17 pm

Thumper68 wrote: ↑
Sat Jan 18, 2020 8:12 am
zonker wrote: ↑
Fri Jan 17, 2020 7:13 pm
Thumper68 wrote: ↑
Fri Jan 17, 2020 6:41 pm
Why are you restricting your max pressure to 14.8? I think that your machine wants to increase above that at times to fight something.
i don't see it trying to raise above 14.8 on the charts he's shown.
On 1-9 it went to 14.6 and on the next one 1-15 it did max at 14.8. He may have a valid reason to limit it but I prefer to set the max to max!
you've got better eyes than mine! good catch. and i agree with setting max to max. otherwise, it's not max! :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe